Menu

Question 4101

Topic: 5. Sports Medicine

A 22-year-old collegiate soccer player undergoes anterior cruciate ligament (ACL) reconstruction using a bone-patellar tendon-bone autograft. Compared to a quadruple-stranded hamstring autograft, what is the most likely long-term complication associated with this graft choice?

. Decreased deep knee flexion strength
. Anterior knee pain
. Graft rupture
. Increased anterior tibial translation
. Saphenous nerve injury

Correct Answer & Explanation

. Decreased deep knee flexion strength


Explanation

Bone-patellar tendon-bone autografts are historically associated with a higher incidence of anterior knee pain and pain with kneeling compared to hamstring autografts. Hamstring grafts, on the other hand, are associated with decreased deep knee flexion strength.

Question 4102

Topic: Knee Sports

A 25-year-old man presents with knee pain after a twisting injury. Physical examination reveals an asymmetric increased external rotation of the tibia of 15 degrees at 30 degrees of knee flexion, but no difference compared to the contralateral knee at 90 degrees of flexion. Which of the following structures is most likely injured?

. Posterior cruciate ligament (PCL) alone
. Posterolateral corner (PLC) alone
. Combined PCL and PLC
. Anterior cruciate ligament (ACL)
. Medial collateral ligament (MCL)

Correct Answer & Explanation

. Posterior cruciate ligament (PCL) alone


Explanation

The dial test evaluates the PCL and PLC. Greater than 10 degrees of asymmetric external rotation at 30 degrees but symmetric at 90 degrees indicates an isolated PLC injury. Asymmetry at both 30 and 90 degrees indicates a combined PLC and PCL injury.

Question 4103

Topic: Shoulder & Hip Sports

A 20-year-old rugby player undergoes a Latarjet procedure for recurrent anterior shoulder instability with significant glenoid bone loss. Postoperatively, he presents with weakness in elbow flexion and decreased sensation over the lateral forearm. Which nerve was most likely injured during the procedure?

. Axillary nerve
. Median nerve
. Musculocutaneous nerve
. Radial nerve
. Suprascapular nerve

Correct Answer & Explanation

. Axillary nerve


Explanation

The musculocutaneous nerve is the most commonly injured nerve during the Latarjet procedure due to its proximity to the coracoid process. It innervates the biceps brachii and brachialis, and provides sensation to the lateral forearm via the lateral antebrachial cutaneous nerve.

Question 4104

Topic: 5. Sports Medicine

A 45-year-old male manual laborer complains of deep shoulder pain and mechanical clicking. MRI demonstrates a Type II Superior Labrum Anterior to Posterior (SLAP) tear. Nonoperative management has failed. What is the most appropriate surgical treatment in this patient demographic?

. Arthroscopic SLAP repair
. Arthroscopic debridement of the labrum
. Biceps tenodesis
. Biceps tenotomy
. Diagnostic arthroscopy only

Correct Answer & Explanation

. Arthroscopic SLAP repair


Explanation

In patients older than 40 years, especially non-overhead athletes or manual laborers, biceps tenodesis provides superior outcomes and lower revision rates compared to SLAP repair for Type II SLAP lesions. SLAP repair in this age group has a higher rate of postoperative stiffness and persistent pain.

Question 4105

Topic: Shoulder & Hip Sports

A 35-year-old man presents to the emergency department complaining of shoulder pain after a first-time generalized tonic-clonic seizure. On examination, his arm is locked in adduction and internal rotation. Radiographs reveal a posterior shoulder dislocation with an anteromedial humeral head defect involving 20% of the articular surface. What is this defect termed?

. Bankart lesion
. Hill-Sachs lesion
. Reverse Hill-Sachs lesion
. ALPSA lesion
. GLAD lesion

Correct Answer & Explanation

. Bankart lesion


Explanation

A reverse Hill-Sachs lesion (McLaughlin lesion) is an impaction fracture of the anteromedial humeral head commonly associated with posterior shoulder dislocations. A classic Hill-Sachs lesion is a posterolateral defect associated with anterior dislocations.

Question 4106

Topic: Knee Sports

During an anterior cruciate ligament (ACL) reconstruction, the surgeon evaluates the anatomy of the native ACL footprint. Which of the following statements correctly describes the biomechanics of the two main bundles of the ACL?

. Both bundles are maximally tight at 45 degrees of flexion
. The anteromedial bundle is tight in flexion and the posterolateral bundle is tight in extension
. The anteromedial bundle is tight in extension and the posterolateral bundle is tight in flexion
. Both bundles are equally tight in extension
. The anteromedial bundle provides primarily rotatory stability

Correct Answer & Explanation

. Both bundles are maximally tight at 45 degrees of flexion


Explanation

The ACL consists of the anteromedial (AM) and posterolateral (PL) bundles. The AM bundle tightens in flexion to resist anterior tibial translation, while the PL bundle tightens in extension and is the primary restraint to rotatory loads.

Question 4107

Topic: Knee Sports

A 28-year-old dashboard-injury victim presents with a positive posterior drawer test and a posterior sag sign. MRI confirms an isolated Grade II posterior cruciate ligament (PCL) tear. What is the most appropriate initial management?

. Immediate PCL reconstruction
. Primary repair of the PCL
. Immobilization in a cylinder cast in full flexion for 6 weeks
. Physical therapy and bracing in extension
. Arthroscopic debridement

Correct Answer & Explanation

. Immediate PCL reconstruction


Explanation

Isolated Grade I and II PCL tears have a high intrinsic healing capacity. They are best managed nonoperatively with relative immobilization in extension (to prevent posterior tibial sag) and physical therapy focusing on quadriceps strengthening.

Question 4108

Topic: Shoulder & Hip Sports

A 35-year-old man presents with a locked shoulder after a generalized seizure. Radiographs demonstrate a posterior shoulder dislocation with an anterior humeral head defect involving 25% of the articular surface as shown in Figure 1.

What is the most appropriate surgical management?

. Closed reduction and sling immobilization
. Arthroscopic Bankart repair
. Open reduction and lesser tuberosity transfer
. Coracoid transfer (Latarjet procedure)
. Total shoulder arthroplasty

Correct Answer & Explanation

. Closed reduction and sling immobilization


Explanation

Posterior shoulder dislocations with an associated reverse Hill-Sachs lesion involving 20% to 40% of the articular surface are best treated with a lesser tuberosity transfer (McLaughlin procedure) into the defect. Defects >40% typically require arthroplasty.

Question 4109

Topic: 5. Sports Medicine

A 22-year-old football player undergoes an anterior cruciate ligament (ACL) reconstruction using a bone-patellar tendon-bone autograft. Six months postoperatively, he complains of anterior knee pain and a persistent 10-degree extension lag. A sagittal MRI reveals graft impingement against the intercondylar notch roof. Which of the following technical errors most likely caused this complication?

. Tibial tunnel placed too anterior
. Tibial tunnel placed too posterior
. Femoral tunnel placed too anterior
. Femoral tunnel placed too posterior
. Inadequate femoral notchplasty alone

Correct Answer & Explanation

. Tibial tunnel placed too anterior


Explanation

A tibial tunnel placed too far anteriorly causes the ACL graft to impinge on the roof of the intercondylar notch in terminal extension. This leads to an extension deficit, anterior knee pain, and potentially a cyclops lesion.

Question 4110

Topic: Knee Sports

An 8-year-old girl (Tanner stage 1) sustains a midsubstance anterior cruciate ligament (ACL) tear. She experiences recurrent instability despite bracing and physical therapy. What is the most appropriate surgical technique to minimize the risk of physeal growth arrest?

. Transphyseal bone-patellar tendon-bone reconstruction
. Transphyseal hamstring reconstruction
. Primary repair with synthetic augmentation
. All-epiphyseal ACL reconstruction
. Over-the-top extra-articular tenodesis alone

Correct Answer & Explanation

. Transphyseal bone-patellar tendon-bone reconstruction


Explanation

In prepubescent children with significant remaining growth (Tanner stage 1 or 2), physeal-sparing techniques like an all-epiphyseal reconstruction or extra-articular ITB tenodesis are recommended. Transphyseal drilling, especially with bone blocks, carries an unacceptably high risk of growth arrest.

Question 4111

Topic: 5. Sports Medicine

A 22-year-old collegiate soccer player undergoes anterior cruciate ligament (ACL) reconstruction using a bone-patellar tendon-bone (BPTB) autograft. Which of the following is the most common complication uniquely associated with this specific graft choice compared to hamstring autograft?

. Increased postoperative AP laxity
. Anterior knee pain and difficulty kneeling
. Higher rate of deep joint infection
. Increased risk of postoperative arthrofibrosis
. Premature osteoarthritis at 5 years

Correct Answer & Explanation

. Increased postoperative AP laxity


Explanation

Bone-patellar tendon-bone (BPTB) autografts are historically associated with a higher incidence of donor site morbidity, specifically anterior knee pain and pain with kneeling, compared to hamstring autografts. Rates of graft rupture and postoperative laxity are generally comparable between the two.

Question 4112

Topic: Shoulder & Hip Sports

A 19-year-old female gymnast presents with bilateral shoulder pain and a sensation of the shoulders "slipping" out of joint. Examination reveals positive sulcus signs bilaterally and generalized ligamentous laxity (Beighton score 7/9). Initial management should consist of:

. Arthroscopic capsular plication
. Open inferior capsular shift
. Thermal capsulorrhaphy
. Physical therapy focusing on periscapular and rotator cuff strengthening
. Arthroscopic Bankart repair

Correct Answer & Explanation

. Arthroscopic capsular plication


Explanation

Multidirectional instability (MDI) often presents with generalized ligamentous laxity and bilateral symptoms. First-line treatment is strictly nonoperative, focusing on an extensive rehabilitation program to strengthen the dynamic stabilizers of the shoulder.

Question 4113

Topic: Knee Sports

A 26-year-old rugby player sustains a direct blow to the proximal tibia with the knee flexed at 90 degrees. Examination demonstrates a posterior sag sign and 8 mm of posterior translation of the tibia on the femur at 90 degrees of flexion, but normal translation at 30 degrees. The dial test is negative. What is the most appropriate initial management?

. Immediate surgical reconstruction of the PCL
. Primary repair of the torn ligament
. Nonoperative management with early rehabilitation emphasizing quadriceps strengthening
. Cylinder cast immobilization in full extension for 6 weeks
. Aggressive hamstring strengthening program

Correct Answer & Explanation

. Immediate surgical reconstruction of the PCL


Explanation

The scenario describes an isolated Grade II posterior cruciate ligament (PCL) injury (5-10 mm translation). The standard of care for isolated Grade I and II PCL injuries is nonoperative management, emphasizing early range of motion and quadriceps strengthening to counteract posterior tibial translation.

Question 4114

Topic: 5. Sports Medicine

A 24-year-old throwing athlete undergoes an arthroscopic Type II SLAP repair. What is the most common postoperative complication that prevents a return to the previous level of overhead competition?

. Recurrent anterior instability
. Postoperative stiffness with loss of external rotation
. Suprascapular nerve entrapment
. Biceps tendon rupture
. Axillary nerve palsy

Correct Answer & Explanation

. Recurrent anterior instability


Explanation

The most common complication following SLAP repair in overhead athletes is postoperative stiffness, specifically a loss of external rotation. This loss of motion can severely impact throwing mechanics and prevent a successful return to elite play.

Question 4115

Topic: Knee Sports

A 31-year-old soccer player sustains a hyperextension and varus injury to his knee. On examination, he has 15 degrees of increased external tibial rotation at 30 degrees of knee flexion compared to the contralateral side. This asymmetry completely resolves when tested at 90 degrees of flexion. Which structure is most likely injured?

. Anterior cruciate ligament (ACL)
. Posterior cruciate ligament (PCL)
. Posterolateral corner (PLC)
. Medial collateral ligament (MCL)
. Posteromedial corner (PMC)

Correct Answer & Explanation

. Anterior cruciate ligament (ACL)


Explanation

Increased external rotation of the tibia at 30 degrees of flexion that reduces to normal at 90 degrees indicates an isolated injury to the posterolateral corner (PLC). If the rotational asymmetry persisted at 90 degrees, it would suggest a combined PLC and PCL injury.

Question 4116

Topic: Shoulder & Hip Sports

A 22-year-old football player sustains a recurrent anterior shoulder dislocation.

Advanced imaging demonstrates an engaging Hill-Sachs lesion and anterior glenoid bone loss of 28%. What is the most appropriate surgical management to minimize the risk of recurrence?

. Arthroscopic Bankart repair with superior labral repair
. Arthroscopic Remplissage and standard Bankart repair
. Coracoid transfer (Latarjet procedure)
. Open inferior capsular shift
. Arthroscopic thermal capsulorrhaphy

Correct Answer & Explanation

. Arthroscopic Bankart repair with superior labral repair


Explanation

In the setting of significant anterior glenoid bone loss (subcritical typically >13.5%, critical >20-25%) and an engaging Hill-Sachs lesion, soft tissue procedures have unacceptably high failure rates. A bony augmentation procedure, such as the Latarjet procedure, is indicated to restore stability.

Question 4117

Topic: 5. Sports Medicine

A 40-year-old recreational skier presents with medial knee pain after catching an inside edge. Examination reveals 4 mm of medial opening to valgus stress at 30 degrees of flexion with a firm endpoint, but no opening at 0 degrees. MRI confirms a partial tear of the superficial MCL at its femoral attachment. What is the recommended treatment?

. Surgical repair of the MCL using suture anchors
. Nonoperative management with a hinged knee brace and early functional rehabilitation
. MCL reconstruction using semitendinosus allograft
. Immobilization in a cylinder cast for 6 weeks
. Diagnostic arthroscopy to confirm healing potential

Correct Answer & Explanation

. Surgical repair of the MCL using suture anchors


Explanation

An isolated Grade I or II medial collateral ligament (MCL) injury is optimally managed nonoperatively. A hinged knee brace allows early range of motion while protecting against valgus stress, leading to excellent ligamentous healing and return to sport.

Question 4118

Topic: Shoulder & Hip Sports

A 48-year-old ski instructor dislocates his nondominant shoulder in a fall. Management consisting of application of a sling for 1 week results in improvement in his pain. Follow-up examination 6 weeks after the injury reveals that the patient continues to have difficulty with shoulder elevation. Management should now include

. use of the sling for an additional 3 weeks.
. physical therapy.
. a corticosteroid injection.
. an MRI scan of the rotator cuff.
. arthroscopic labral repair.

Correct Answer & Explanation

. use of the sling for an additional 3 weeks.


Explanation

Patients who are older than age 45 years and have initial dislocations are at greater risk for tearing the rotator cuff. Patients who are unable to lift the upper extremity or who have continued pain should undergo further evaluation for potential rotator cuff tears; early diagnosis is preferred. Physical therapy or continued use of a sling will be of little benefit. A corticosteroid injection might delay the diagnosis and compromise subsequent rotator cuff repair. Repairing the labrum generally is not necessary in a patient of this age who has an initial dislocation. Hawkins RJ, Bell RH, Hawkins RH, Koppert GJ: Anterior dislocation of the shoulder in the older patient. Clin Orthop 1986;206:192-195.

Question 4119

Topic: Knee Sports

Anterior cruciate ligament (ACL) reconstruction is planned. When evaluating graft placement, a femoral tunnel positioned too anteriorly (shallow) in the intercondylar notch will result in a graft that is biomechanically:

. Tight in flexion, loose in extension
. Tight in extension, loose in flexion
. Tight in both flexion and extension
. Loose in both flexion and extension
. Isometrically tensioned throughout range of motion

Correct Answer & Explanation

. Tight in flexion, loose in extension


Explanation

An anteriorly placed femoral tunnel causes the distance between the femoral and tibial attachments to increase during knee flexion. This results in a graft that is tight in flexion and loose in extension.

Question 4120

Topic: Knee Sports

A 22-year-old collegiate football player undergoes anterior cruciate ligament (ACL) reconstruction using a bone-patellar tendon-bone autograft. During the procedure, the femoral tunnel is drilled too anteriorly. Which of the following complications is most likely to occur?

. Knee stiffness in flexion (loss of extension)
. Knee stiffness in extension (loss of flexion)
. Lateral patellar tracking
. Laxity of the graft in extension
. Anterior knee pain

Correct Answer & Explanation

. Knee stiffness in flexion (loss of extension)


Explanation

An anterior femoral tunnel placement in ACL reconstruction results in increased graft tension during flexion, limiting knee flexion. Conversely, an excessively posterior femoral tunnel limits extension.