Menu

Question 1201

Topic: Knee Sports

An AP radiograph of a 22-year-old skier's acutely injured knee demonstrates an elliptic bony avulsion fragment just distal to the lateral tibial plateau. This finding is highly pathognomonic for an injury to which structure, and what associated major ligamentous tear is likely present?

. Biceps femoris; PCL tear
. Popliteus tendon; PLC injury
. Anterolateral ligament; ACL tear
. Iliotibial band; LCL tear
. Medial patellofemoral ligament; Patellar dislocation

Correct Answer & Explanation

. Biceps femoris; PCL tear


Explanation

The Segond fracture is a cortical avulsion of the anterolateral capsule (specifically the anterolateral ligament) from the proximal lateral tibia. It is highly pathognomonic (up to 75-100% association) for an underlying anterior cruciate ligament (ACL) tear.

Question 1202

Topic: Knee Sports

During a medial patellofemoral ligament (MPFL) reconstruction for recurrent patellar instability, the surgeon uses fluoroscopy to identify the femoral attachment point (Schöttle's point). This radiographic landmark is located:

. Anterior to the posterior femoral cortical line and proximal to the posterior aspect of the Blumensaat line
. Posterior to the posterior femoral cortical line and distal to the Blumensaat line
. Anterior to the posterior femoral cortical line, proximal to the Blumensaat line, and distal to the medial epicondyle
. Between the medial epicondyle and adductor tubercle, anterior to the posterior femoral cortical line
. At the distal pole of the patella

Correct Answer & Explanation

. Anterior to the posterior femoral cortical line and proximal to the posterior aspect of the Blumensaat line


Explanation

Schöttle's point is an essential radiographic landmark for anatomical femoral MPFL graft placement. It lies just anterior to the posterior femoral cortical line and proximal to the most posterior point of Blumensaat's line.

Question 1203

Topic: Knee Sports

A surgeon is performing a PCL reconstruction and chooses an open tibial inlay technique over a transtibial tunnel technique. The primary biomechanical advantage of the tibial inlay technique is:

. Reduced risk of popliteal artery injury
. Avoidance of the acute 'killer turn' at the proximal posterior tibia
. Shorter overall graft length requirement
. Ability to perform the procedure entirely arthroscopically
. Lower incidence of postoperative deep vein thrombosis

Correct Answer & Explanation

. Reduced risk of popliteal artery injury


Explanation

The transtibial PCL reconstruction creates an acute angle (the "killer turn") at the posterior aperture of the tibial tunnel, which can cause graft abrasion and attenuation. The open tibial inlay technique secures the graft directly to the posterior tibia, bypassing this sharp angle.

Question 1204

Topic: Knee Sports

A 24-year-old football player sustains a direct blow to the anteromedial aspect of his knee. Examination shows a positive dial test with 15 degrees of increased external rotation at 30 degrees of knee flexion, but symmetric external rotation compared to the contralateral knee at 90 degrees. Which structure is most likely injured?

. Posterior cruciate ligament
. Anterior cruciate ligament
. Posterolateral corner (isolated)
. Posterolateral corner and posterior cruciate ligament
. Medial collateral ligament

Correct Answer & Explanation

. Posterior cruciate ligament


Explanation

Increased external rotation at 30 degrees of flexion with normal rotation at 90 degrees indicates an isolated posterolateral corner (PLC) injury. Combined PLC and PCL injuries demonstrate increased external rotation at both 30 and 90 degrees of knee flexion.

Question 1205

Topic: Knee Sports

Six months following an uncomplicated bone-patellar tendon-bone ACL reconstruction, a patient complains of a painful "clunk" and an inability to achieve terminal knee extension. MRI shows a nodular mass anterior to the ACL graft. What surgical error is the most common cause of this complication?

. Oversized graft diameter
. Excessively posterior femoral tunnel placement
. Anterior placement of the tibial tunnel
. Inadequate notchplasty
. Vertical placement of the femoral tunnel

Correct Answer & Explanation

. Oversized graft diameter


Explanation

A cyclops lesion (localized anterior arthrofibrosis) causes loss of terminal extension and a painful clunk. It is strongly associated with an anteriorly placed tibial tunnel, leading to graft impingement in the intercondylar notch during extension.

Question 1206

Topic: Knee Sports

A 9-year-old boy (Tanner stage I) with widely open physes sustains a mid-substance ACL tear. His parents wish to proceed with surgical management due to recurrent instability episodes. Which surgical technique is most appropriate?

. Transphyseal reconstruction with bone-patellar tendon-bone graft
. Iliotibial band extra-articular tenodesis (MacIntosh procedure)
. Standard adult transphyseal hamstring reconstruction
. Primary repair of the ACL with suture augmentation
. All-inside transphyseal reconstruction

Correct Answer & Explanation

. Transphyseal reconstruction with bone-patellar tendon-bone graft


Explanation

In prepubescent patients with widely open physes (Tanner stage I), a physeal-sparing technique like the IT band extra-articular tenodesis or an all-epiphyseal reconstruction is recommended. Transphyseal techniques carry an unacceptably high risk of growth arrest or angular deformity in this age group.

Question 1207

Topic: Knee Sports

During an ACL reconstruction, a systematic arthroscopic evaluation is performed. Viewing from the anterolateral portal through the intercondylar notch reveals a tear at the meniscocapsular junction of the posterior horn of the medial meniscus. What is this specific lesion called?

. Root tear
. Wrisberg rip
. Ramp lesion
. Radial tear
. Parrot-beak tear

Correct Answer & Explanation

. Root tear


Explanation

A meniscal "ramp" lesion is a tear at the peripheral meniscocapsular junction of the posterior horn of the medial meniscus. It is frequently associated with acute ACL tears and is best visualized through a posteromedial portal or via an intercondylar trans-notch view.

Question 1208

Topic: Knee Sports



To address residual anterolateral rotatory instability during an ACL reconstruction, an anterolateral ligament (ALL) reconstruction is planned. Which of the following describes the correct anatomic origin and insertion of the ALL?

. Origin: Anterior to FCL; Insertion: Gerdy's tubercle
. Origin: Posterior and proximal to FCL; Insertion: Midway between Gerdy's tubercle and fibular head
. Origin: Anterior to popliteus; Insertion: Fibular head
. Origin: Posterior to lateral epicondyle; Insertion: Fibular head
. Origin: Proximal to FCL; Insertion: Lateral joint line

Correct Answer & Explanation

. Origin: Anterior to FCL; Insertion: Gerdy's tubercle


Explanation

The anterolateral ligament (ALL) originates slightly posterior and proximal to the fibular collateral ligament (FCL) on the lateral femoral epicondyle. It inserts on the proximal anterolateral tibia, approximately midway between Gerdy's tubercle and the fibular head.

Question 1209

Topic: Knee Sports

When comparing the tibial inlay technique to the transtibial tunnel technique for posterior cruciate ligament (PCL) reconstruction, the tibial inlay technique specifically avoids which of the following biomechanical issues?

. Graft divergence at the femoral tunnel
. The "killer turn" and subsequent graft abrasion
. Damage to the popliteal artery during drilling
. Insufficient graft length
. Posterior translation in deep flexion

Correct Answer & Explanation

. Graft divergence at the femoral tunnel


Explanation

The tibial inlay technique secures the bone block directly to the posterior tibial facet, avoiding the acute angle or "killer turn" at the proximal tibial aperture. This reduces the risk of graft abrasion and attenuation frequently seen in transtibial PCL reconstructions.

Question 1210

Topic: Knee Sports

A 48-year-old woman experiences a sudden "pop" in the posterior aspect of her knee while squatting. MRI reveals a posterior medial meniscal root tear with 4 mm of meniscal extrusion. Which of the following best describes the biomechanical consequence of this injury if left untreated?

. Increased tension on the anterior cruciate ligament
. Decreased peak contact pressures in the medial compartment
. Biomechanical equivalence to a total medial meniscectomy
. Increased external rotation of the tibia
. Shift of the mechanical axis into valgus

Correct Answer & Explanation

. Increased tension on the anterior cruciate ligament


Explanation

A posterior root tear of the medial meniscus disrupts the hoop stresses, rendering the meniscus functionally incompetent. This results in altered knee kinematics and contact pressures that are biomechanically equivalent to a total medial meniscectomy.

Question 1211

Topic: Knee Sports

A 26-year-old soccer player sustains a twisting injury to his knee. On examination, the dial test reveals 15 degrees of increased external rotation of the tibia at 30 degrees of knee flexion compared to the contralateral side. At 90 degrees of flexion, the side-to-side difference in external rotation is less than 5 degrees. Which of the following injuries is most likely present?

. Isolated posterior cruciate ligament (PCL) injury
. Combined PCL and posterolateral corner (PLC) injury
. Isolated posterolateral corner (PLC) injury
. Anterior cruciate ligament (ACL) and medial collateral ligament (MCL) injury
. Isolated posteromedial corner injury

Correct Answer & Explanation

. Isolated posterior cruciate ligament (PCL) injury


Explanation

The dial test evaluates for PLC and PCL injuries. Increased external rotation (>10 degrees compared to the normal knee) at 30 degrees of flexion only indicates an isolated PLC injury, whereas increased rotation at both 30 and 90 degrees indicates combined PLC and PCL injuries.

Question 1212

Topic: Knee Sports

A 25-year-old rugby player sustains a direct blow to the anteromedial aspect of his knee. Examination reveals increased external rotation of the tibia at 30 degrees of knee flexion, but symmetric external rotation at 90 degrees compared to the contralateral side. Which structure is the primary deficient restraint responsible for this examination finding?

. Posterior cruciate ligament
. Anterior cruciate ligament
. Fibular collateral ligament
. Biceps femoris
. Iliotibial band

Correct Answer & Explanation

. Posterior cruciate ligament


Explanation

An isolated posterolateral corner (PLC) injury presents with increased external rotation at 30 degrees of flexion but not at 90 degrees. The fibular collateral ligament (FCL) is the primary static constraint to varus and external rotation at 30 degrees.

Question 1213

Topic: Knee Sports

A 25-year-old rugby player presents with a twisting knee injury. On examination, the dial test demonstrates 15 degrees of increased external rotation compared to the contralateral normal knee at 30 degrees of knee flexion, but symmetrical external rotation at 90 degrees of flexion. Which structure is most likely injured?

. Combined posterolateral corner (PLC) and posterior cruciate ligament (PCL)
. Isolated posterior cruciate ligament (PCL)
. Isolated posterolateral corner (PLC)
. Anterior cruciate ligament (ACL) and PLC
. Medial collateral ligament (MCL) and posterior oblique ligament (POL)

Correct Answer & Explanation

. Combined posterolateral corner (PLC) and posterior cruciate ligament (PCL)


Explanation

An isolated posterolateral corner (PLC) injury presents with increased external rotation at 30 degrees but not 90 degrees of knee flexion. If increased external rotation is present at both 30 and 90 degrees, a combined PLC and PCL injury is present.

Question 1214

Topic: Knee Sports

During a medial patellofemoral ligament (MPFL) reconstruction for recurrent patellar instability, the surgeon uses fluoroscopy to identify the femoral attachment site. According to Schöttle's method, where is the anatomic femoral origin of the MPFL located?

. Anterior to a line extending the posterior cortex of the femoral diaphysis
. Just proximal to the medial epicondyle and distal to the adductor tubercle
. Distal to the medial epicondyle
. Anterior to the Blumensaat line on a true lateral radiograph
. Directly on the adductor tubercle

Correct Answer & Explanation

. Anterior to a line extending the posterior cortex of the femoral diaphysis


Explanation

Schöttle's point is located 1 mm anterior to the posterior femoral cortical line and 2.5 mm distal to the posterior origin of the medial femoral condyle. Clinically, this lies in the saddle just proximal to the medial epicondyle and distal to the adductor tubercle.

Question 1215

Topic: Knee Sports

A 28-year-old man sustains a twisting injury to his knee. Examination reveals 15 degrees of increased external rotation on the dial test at 30 degrees of knee flexion compared to the contralateral side. At 90 degrees of knee flexion, the external rotation is equal bilaterally. Which structure is most likely injured?

. Isolated posterior cruciate ligament (PCL)
. Isolated posterolateral corner (PLC)
. Combined PCL and PLC
. Anterior cruciate ligament (ACL) and PLC
. Isolated medial collateral ligament (MCL)

Correct Answer & Explanation

. Isolated posterior cruciate ligament (PCL)


Explanation

The dial test evaluates PLC and PCL integrity. Greater than 10 degrees of external rotation asymmetry at 30 degrees only indicates an isolated PLC injury. Asymmetry at both 30 and 90 degrees indicates a combined PLC and PCL injury.

Question 1216

Topic: Knee Sports

A 28-year-old skier sustains an acute knee injury and presents with a positive dial test at 30 degrees of flexion, which normalizes to the contralateral side at 90 degrees of flexion. Which of the following structures is most likely injured?

. Isolated Posterior cruciate ligament (PCL)
. Isolated Anterior cruciate ligament (ACL)
. Isolated Posterolateral corner (PLC)
. Isolated Medial collateral ligament (MCL)
. Combined PCL and PLC

Correct Answer & Explanation

. Isolated Posterior cruciate ligament (PCL)


Explanation

A positive dial test (increased external rotation of >10 degrees compared to the normal side) at 30 degrees of flexion that reduces at 90 degrees indicates an isolated injury to the posterolateral corner (PLC). If the test is positive at both 30 and 90 degrees, it suggests a combined PCL and PLC injury.

Question 1217

Topic: Knee Sports

A 16-year-old female experiences recurrent patellar dislocations. Physical exam reveals apprehension with lateral patellar translation. What is the primary soft-tissue restraint to lateral patellar translation at 20 degrees of knee flexion?

. Medial patellofemoral ligament (MPFL)
. Medial patellotibial ligament
. Medial patellomeniscal ligament
. Vastus medialis obliquus
. Lateral retinaculum

Correct Answer & Explanation

. Medial patellofemoral ligament (MPFL)


Explanation

The medial patellofemoral ligament (MPFL) provides approximately 50-60% of the restraining force against lateral patellar displacement, acting primarily in early flexion (0 to 30 degrees).

Question 1218

Topic: Knee Sports

Which bundle of the anterior cruciate ligament (ACL) is the primary restraint to anterior tibial translation when the knee is in 90 degrees of flexion?

. Anteromedial (AM) bundle
. Posterolateral (PL) bundle
. Anterolateral (AL) bundle
. Posteromedial (PM) bundle
. Intermediate bundle

Correct Answer & Explanation

. Anteromedial (AM) bundle


Explanation

The anteromedial (AM) bundle of the ACL is tightest in knee flexion and acts as the primary restraint to anterior tibial translation at 90 degrees. Conversely, the posterolateral (PL) bundle is tight in extension and primarily controls rotational stability.

Question 1219

Topic: Knee Sports

A 19-year-old football player sustains a valgus blow to his knee. Exam reveals a grade III MCL injury and an ACL tear. What is the recommended treatment strategy?

. Simultaneous ACL and MCL reconstruction acutely
. Acute MCL repair followed by delayed ACL reconstruction
. Nonoperative management of the MCL in a hinged brace, followed by delayed ACL reconstruction
. Acute ACL reconstruction and nonoperative management of the MCL
. Nonoperative management of both injuries

Correct Answer & Explanation

. Simultaneous ACL and MCL reconstruction acutely


Explanation

Combined ACL and grade III MCL injuries are typically managed with a period of bracing to allow the MCL to heal, followed by delayed ACL reconstruction. This approach minimizes the significant risk of postoperative arthrofibrosis seen with acute simultaneous surgeries.

Question 1220

Topic: Knee Sports

A patient presents with a chronic posterolateral corner (PLC) deficient knee resulting in a varus thrust during gait. Radiographs show mechanical axis falling medial to the knee center and no advanced osteoarthritis. What is the most appropriate initial surgical management?

. Total knee arthroplasty
. High tibial osteotomy (HTO)
. PLC reconstruction only
. Unicompartmental knee arthroplasty
. ACL and PLC reconstruction

Correct Answer & Explanation

. Total knee arthroplasty


Explanation

In the setting of chronic posterolateral corner deficiency with varus malalignment, correcting the bony alignment with a valgus-producing High Tibial Osteotomy (HTO) is essential. Soft-tissue reconstruction alone will likely fail due to the constant varus overload.