Menu

Question 4221

Topic: 3. Adult Reconstruction (Hip & Knee)

A 55-year-old male with long-standing psoriatic arthritis is planning to undergo a total knee arthroplasty (TKA). Compared to a patient with primary osteoarthritis, this patient is at a significantly higher risk for which of the following postoperative complications?

. Periprosthetic fracture
. Aseptic loosening
. Surgical site infection
. Symptomatic venous thromboembolism
. Instability requiring revision

Correct Answer & Explanation

. Surgical site infection


Explanation

Patients with psoriatic arthritis have a higher risk of periprosthetic joint infection following TKA compared to osteoarthritis patients. This is attributed to skin colonization from psoriatic plaques, underlying immunosuppression from biologic therapies, and chronic systemic inflammation.

Question 4222

Topic: 3. Adult Reconstruction (Hip & Knee)

A 42-year-old male presents with a chronic, slowly enlarging, painless knee effusion. An MRI is obtained.

Based on the most likely diagnosis, what is the definitive treatment of choice?

. Intra-articular corticosteroid injection
. Arthroscopic or open synovectomy
. Total knee arthroplasty
. Radiation synovectomy
. Observation and NSAIDs

Correct Answer & Explanation

. Arthroscopic or open synovectomy


Explanation

The image demonstrates lipoma arborescens, characterized by frond-like fat accumulations in the synovium. The definitive and curative treatment is an open or arthroscopic synovectomy.

Question 4223

Topic: 3. Adult Reconstruction (Hip & Knee)

A 45-year-old male presents with chronic, painless swelling of the right knee without prior trauma. A sagittal T1-weighted MRI is shown below.

Which of the following is the most definitive and appropriate treatment for this condition?

. Intra-articular corticosteroid injection
. Arthroscopic or open complete synovectomy
. Systemic administration of methotrexate
. Marginal excision of a localized extra-articular mass
. Total knee arthroplasty

Correct Answer & Explanation

. Arthroscopic or open complete synovectomy


Explanation

The MRI demonstrates lipoma arborescens, characterized by a frond-like, villous proliferation of hyperplastic synovium replaced by mature fat, classically in the suprapatellar pouch. The definitive treatment for symptomatic patients is arthroscopic or open complete synovectomy.

Question 4224

Topic: 3. Adult Reconstruction (Hip & Knee)

A 42-year-old female with severe psoriatic arthritis presents with incapacitating pain, instability, and a 40-degree flexion deformity of her dominant index finger distal interphalangeal (DIP) joint. Radiographs show a classic "pencil-in-cup" deformity. If surgery is planned, which of the following is the most reliable procedure?

. DIP joint silicone replacement arthroplasty
. DIP joint arthrodesis positioned in 0 to 10 degrees of flexion
. Resection arthroplasty with palmaris longus tendon interposition
. Extensor tendon central slip centralization with lateral band release
. Volar plate advancement combined with collateral ligament reconstruction

Correct Answer & Explanation

. DIP joint arthrodesis positioned in 0 to 10 degrees of flexion


Explanation

For advanced, symptomatic DIP joint destruction in psoriatic arthritis, such as a "pencil-in-cup" deformity, DIP joint arthrodesis is the gold standard treatment. It provides stable, pain-free pinch function and is typically positioned in 0 to 10 degrees of flexion for the index finger.

Question 4225

Topic: Total Hip Arthroplasty (THA)

An 8-year-old boy with OI Type IV underwent bilateral femoral rodding with Fassier-Duval telescoping rods 4 years ago. He now presents with a 3 cm leg length discrepancy and recurrent bowing of the left femur. Radiographs show the left rod has failed to elongate, and the distal femur has grown past the end of the rod. What is the most appropriate management?

. Observation until skeletal maturity.
. Revision surgery to exchange the jammed rod and correct the deformity.
. Epiphysiodesis of the contralateral distal femur.
. Application of a circular external fixator for lengthening.
. Initiation of high-dose bisphosphonate therapy to halt bowing.

Correct Answer & Explanation

. Revision surgery to exchange the jammed rod and correct the deformity.


Explanation

Correct Answer: BFassier-Duval rods are designed to telescope and accommodate longitudinal bone growth. A known complication is failure of the rod to expand (jamming). When this occurs, the bone continues to grow around and past the rod, leading to recurrent deformity (bowing), loss of internal splinting, and leg length discrepancy. The most appropriate management for a symptomatic, jammed rod with recurrent deformity is revision surgery to remove the failed hardware, perform corrective osteotomies if necessary, and insert a new telescoping rod.

Question 4226

Topic: 3. Adult Reconstruction (Hip & Knee)

A 6-year-old non-ambulatory boy with SMA Type II is noted to have a painless, unilateral, chronic hip dislocation on routine surveillance radiographs. He sits comfortably in his custom wheelchair and has no difficulties with perineal hygiene. What is the most appropriate management of his hip dislocation?

. Open reduction and femoral varus derotational osteotomy (VDRO).
. Closed reduction and spica casting.
. Pelvic support osteotomy.
. Observation and continued wheelchair seating modifications.
. Girdlestone resection arthroplasty.

Correct Answer & Explanation

. Observation and continued wheelchair seating modifications.


Explanation

Correct Answer: DIn non-ambulatory patients with Spinal Muscular Atrophy, unilateral or bilateral hip dislocations are common due to progressive muscle imbalance (stronger hip flexors/adductors overpowering extensors/abductors). If the hip dislocation is painless, does not interfere with sitting balance, and does not impede perineal hygiene, surgical intervention is generally not recommended. Surgery in this population carries high risks of recurrence, post-operative stiffness, and medical complications without significantly improving the patient's functional status. Observation and seating modifications are the preferred management.

Question 4227

Topic: 3. Adult Reconstruction (Hip & Knee)

A 6-year-old non-ambulatory male with Spinal Muscular Atrophy Type II is found to have a unilateral, chronic paralytic dislocation of the right hip during a routine radiographic screening. The patient is completely pain-free, has a supple range of motion, and sits comfortably in his custom-molded wheelchair. What is the most appropriate orthopedic management for this hip?

. Open reduction, femoral varus derotational osteotomy, and Dega pelvic osteotomy.
. Closed reduction under general anesthesia and application of a hip spica cast for 6 weeks.
. Observation and continued physical therapy focusing on seating and positioning.
. Adductor tenotomy and psoas release to prevent contralateral hip dislocation.
. Resection arthroplasty (Girdlestone procedure) to prevent future pain.

Correct Answer & Explanation

. Observation and continued physical therapy focusing on seating and positioning.


Explanation

Correct Answer: CHip subluxation and dislocation are extremely common in non-ambulatory patients with SMA due to profound muscle weakness and imbalance (typically strong hip flexors/adductors overpowering weak extensors/abductors). In non-ambulatory patients, a chronically dislocated hip is usually painless and does not significantly impair their primary function, which is sitting. Surgical reconstruction (open reduction, pelvic/femoral osteotomies) in this population carries a high complication rate (infection, recurrent dislocation, fracture due to osteopenia) and does not improve functional status or sitting balance. Therefore, the standard of care for a painless, chronic hip dislocation in a non-ambulatory SMA patient is observation and ensuring appropriate wheelchair seating modifications.

Question 4228

Topic: 3. Adult Reconstruction (Hip & Knee)

A 42-year-old female with a history of systemic sclerosis presents with severe pain and a non-healing ulcer on the tip of her index finger. Radiographs show resorption of the terminal phalangeal tuft. Which of the following autoantibodies is most strongly associated with this specific radiographic finding?

. Anti-dsDNA
. Anti-cyclic citrullinated peptide
. Anti-centromere
. Anti-Ro (SSA)
. Anti-Jo-1

Correct Answer & Explanation

. Anti-centromere


Explanation

Acro-osteolysis and calcinosis cutis are classic hand manifestations of limited cutaneous systemic sclerosis (CREST syndrome). This syndrome is most strongly associated with anti-centromere antibodies.

Question 4229

Topic: 3. Adult Reconstruction (Hip & Knee)

A 45-year-old female with systemic sclerosis (scleroderma) presents with chronic hand pain and fingertip ulcerations. Which of the following radiographic findings is most specifically associated with her condition?

. Ulnar drift with metacarpophalangeal joint subluxation
. Acro-osteolysis of the terminal phalangeal tufts
. Pencil-in-cup deformities of the distal interphalangeal joints
. Subchondral cysts with joint space preservation

Correct Answer & Explanation

. Acro-osteolysis of the terminal phalangeal tufts


Explanation

Acro-osteolysis, which is the resorption of the terminal phalangeal tufts, along with calcinosis cutis, are classic radiographic manifestations of scleroderma in the hand. These occur secondary to severe microvascular ischemia.

Question 4230

Topic: 3. Adult Reconstruction (Hip & Knee)

A 65-year-old male presents with chronic pain in his total knee arthroplasty, performed 3 years ago. Aspiration yields 3,500 WBC/uL with 75% PMNs. Alpha-defensin is positive. Synovial CRP is elevated. According to the 2018 ICM criteria, what is the next best step in management?

. Single-stage revision arthroplasty with cementless components
. Two-stage revision arthroplasty with an antibiotic spacer
. DAIR (Debridement, Antibiotics, and Implant Retention)
. Six weeks of intravenous antibiotics without surgery
. Observation and repeat aspiration in 3 months

Correct Answer & Explanation

. Two-stage revision arthroplasty with an antibiotic spacer


Explanation

The patient meets criteria for chronic periprosthetic joint infection (PJI). The gold standard treatment for chronic PJI in North America is a two-stage revision with an antibiotic spacer.

Question 4231

Topic: 3. Adult Reconstruction (Hip & Knee)
An orthopedic surgeon is planning a total joint arthroplasty in a 68-year-old patient with long-standing systemic sclerosis. Which of the following perioperative complications is this patient at the highest risk of developing compared to the general population?
. Periprosthetic joint infection due to immunosuppression alone
. Delayed wound healing and skin necrosis due to microvascular disease
. Aseptic loosening secondary to osteolysis
. Heterotopic ossification
. Acute implant fracture

Correct Answer & Explanation

. Delayed wound healing and skin necrosis due to microvascular disease


Explanation

Systemic sclerosis leads to significant microvascular obliteration and skin fibrosis. This places patients at an exceedingly high risk of wound dehiscence, delayed healing, and skin necrosis following surgical incisions.

Question 4232

Topic: 3. Adult Reconstruction (Hip & Knee)

A 45-year-old female with a history of systemic sclerosis presents with severe fingertip pain and recurrent ulceration. Radiographs of the hands are most likely to demonstrate which of the following pathognomonic findings?

. Subperiosteal bone resorption on the radial aspect of the middle phalanges
. Acro-osteolysis of the distal phalanges with soft tissue calcification
. Juxta-articular osteopenia with marginal erosions
. Central erosions creating a 'gull-wing' appearance
. Chondrocalcinosis of the triangular fibrocartilage complex

Correct Answer & Explanation

. Acro-osteolysis of the distal phalanges with soft tissue calcification


Explanation

Systemic sclerosis (scleroderma) frequently involves the hand, classically presenting with acro-osteolysis (resorption of the distal phalangeal tufts) and calcinosis cutis. These changes are secondary to severe microvascular ischemia and chronic Raynaud phenomenon.

Question 4233

Topic: 3. Adult Reconstruction (Hip & Knee)

A 65-year-old man presents with acute onset of right knee pain, swelling, and fever 3 weeks after a primary total knee arthroplasty. Aspiration yields a white blood cell count of 75,000 cells/uL with 92% neutrophils. The implant is well-fixed. What is the most appropriate surgical management?

. One-stage revision arthroplasty
. Two-stage revision arthroplasty with an antibiotic spacer
. Debridement, antibiotics, and implant retention (DAIR) with modular component exchange
. Suppressive chronic oral antibiotics without surgery
. Arthroscopic lavage of the knee

Correct Answer & Explanation

. Debridement, antibiotics, and implant retention (DAIR) with modular component exchange


Explanation

Debridement, antibiotics, and implant retention (DAIR) with modular polyethylene exchange is indicated for early postoperative prosthetic joint infections (typically within 4 weeks of surgery). It relies on a well-fixed prosthesis and a healthy soft tissue envelope.

Question 4234

Topic: 3. Adult Reconstruction (Hip & Knee)

A 45-year-old female presents with tightening of the skin on her fingers, dysphagia, and cold-induced digital vasospasm. Radiographs of her hands are most likely to demonstrate which of the following classic findings?

. Juxta-articular osteopenia and marginal erosions
. Resorption of the distal phalangeal tufts and soft tissue calcification
. Central erosions with a "gull-wing" appearance
. Subchondral sclerosis and osteophyte formation
. Punched-out periarticular erosions with overhanging edges

Correct Answer & Explanation

. Resorption of the distal phalangeal tufts and soft tissue calcification


Explanation

This patient has CREST syndrome (limited scleroderma). Classic hand radiographic findings include acro-osteolysis (resorption of the distal tufts) and calcinosis cutis.

Question 4235

Topic: 3. Adult Reconstruction (Hip & Knee)

A 70-year-old man underwent a total hip arthroplasty 3 years ago. He had a dental extraction 4 weeks ago and now presents with 3 days of severe hip pain, erythema, and a temperature of 38.8°C. Aspiration yields 85,000 WBC/uL (95% neutrophils). What is the most appropriate initial surgical management?

. Aspiration alone and 6 weeks of intravenous antibiotics
. One-stage component revision
. Two-stage component revision with an antibiotic spacer
. Debridement, antibiotics, and implant retention (DAIR) with modular exchange
. Suppressive oral antibiotic therapy indefinitely

Correct Answer & Explanation

. Debridement, antibiotics, and implant retention (DAIR) with modular exchange


Explanation

This is an acute hematogenous periprosthetic joint infection (symptoms < 3 weeks, well-fixed implants). The standard of care is DAIR (Debridement, Antibiotics, Implant Retention) with exchange of modular components.

Question 4236

Topic: 3. Adult Reconstruction (Hip & Knee)

Acro-osteolysis of the distal phalanges is frequently observed in patients with systemic sclerosis. Which of the following best describes the primary pathogenesis of this finding?

. Osteoclast hyperactivation by inflammatory cytokines
. Direct autoantibody attack on type I collagen
. Chronic ischemia secondary to microvascular disease
. Recurrent microtrauma due to sensory neuropathy
. Ectopic calcification within the distal tuft

Correct Answer & Explanation

. Chronic ischemia secondary to microvascular disease


Explanation

Acro-osteolysis (resorption of the terminal phalangeal tufts) in systemic sclerosis is primarily driven by chronic tissue ischemia resulting from severe microvascular disease and Raynaud's phenomenon.

Question 4237

Topic: 3. Adult Reconstruction (Hip & Knee)

A 60-year-old female presents with severe hand deformities. Radiographs reveal prominent resorption of the terminal phalangeal tufts, preservation of the joint spaces, and diffuse subcutaneous calcifications. Marginal erosions are absent. This radiographic pattern is most pathognomonic for which of the following diseases?

. Rheumatoid arthritis
. Systemic lupus erythematosus
. Systemic sclerosis
. Psoriatic arthritis
. Osteoarthritis

Correct Answer & Explanation

. Systemic sclerosis


Explanation

Acro-osteolysis (resorption of the terminal tufts) combined with soft tissue calcification (calcinosis cutis) and preservation of joint spaces is a hallmark radiographic presentation of systemic sclerosis (scleroderma).

Question 4238

Topic: 3. Adult Reconstruction (Hip & Knee)

A 54-year-old female with long-standing diffuse systemic sclerosis (scleroderma) presents with progressive, painful flexion contractures of her proximal and distal interphalangeal joints bilaterally. Radiographs show acro-osteolysis but no periarticular erosions. Which of the following is the most likely primary cause of her joint contractures?

. Erosive synovitis leading to joint destruction
. Avascular necrosis of the phalanges
. Fibrosis of the periarticular soft tissues and skin
. Crystalline deposition in the articular cartilage
. Neuropathic joint degeneration secondary to microvascular ischemia

Correct Answer & Explanation

. Fibrosis of the periarticular soft tissues and skin


Explanation

Joint contractures in systemic sclerosis are primarily caused by progressive fibrosis, induration, and tightening of the skin and periarticular soft tissues. Primary erosive arthropathy is rare in scleroderma unless there is an overlapping rheumatoid syndrome.

Question 4239

Topic: 3. Adult Reconstruction (Hip & Knee)

During a total hip arthroplasty in a young adult with advanced Juvenile Idiopathic Arthritis (JIA), the surgeon must be prepared for which of the following common anatomic challenges?

. Coxa vara and retroverted acetabulum
. Excessive femoral offset and coxa magna
. Protrusio acetabuli and excessive femoral anteversion
. Acetabular dysplasia with lateralized center of rotation
. Femoral canal widening (champagne-flute morphology)

Correct Answer & Explanation

. Protrusio acetabuli and excessive femoral anteversion


Explanation

The classic hip morphology in patients with severe JIA includes protrusio acetabuli, a narrow (stovepipe) medullary canal, excessive femoral anteversion, and a small overall skeleton. Surgeons often require specialized or pediatric-sized implants to address these features.

Question 4240

Topic: 3. Adult Reconstruction (Hip & Knee)

A 5-year-old boy has unilateral pauciarticular juvenile idiopathic arthritis predominantly affecting his right knee. If left untreated, what is the most likely long-term effect on his lower extremities?

. Shortening of the right leg
. Overgrowth (lengthening) of the right leg
. Severe valgus deformity of the left knee
. Development of infantile tibia vara
. Avascular necrosis of the right femoral head

Correct Answer & Explanation

. Overgrowth (lengthening) of the right leg


Explanation

In young children with active inflammatory arthritis of a single lower extremity joint (like the knee), chronic hyperemia of the physis stimulates bone growth. This classically results in limb overgrowth (leg-length discrepancy where the affected leg is longer).