Question 4801

Topic: Knee Sports

A 28-year-old male is evaluated for knee pain and instability after a motorcycle accident. Examination reveals a normal posterior drawer test but increased varus laxity at 30 degrees of flexion. The dial test shows 15 degrees of increased external rotation at 30 degrees of knee flexion compared to the contralateral side, but symmetric external rotation at 90 degrees. Which of the following structures is most likely injured?

. Isolated Posterior Cruciate Ligament (PCL)
. Isolated Posterolateral Corner (PLC)
. Combined PCL and PLC
. Combined ACL and PLC
. Medial Collateral Ligament (MCL)

Correct Answer & Explanation

. Isolated Posterolateral Corner (PLC)


Explanation

The dial test is used to evaluate the posterolateral corner (PLC) and posterior cruciate ligament (PCL). Increased external rotation (>10 degrees compared to the normal knee) strictly at 30 degrees of flexion with symmetric rotation at 90 degrees indicates an isolated PLC injury. If the dial test is positive at both 30 and 90 degrees, it strongly suggests a combined PCL and PLC injury.

Question 4802

Topic: Shoulder & Hip Sports

A 21-year-old male hockey player presents with deep anterior groin pain that worsens with prolonged sitting and deep flexion activities. Physical exam is remarkable for a positive flexion, adduction, internal rotation (FADIR) test. Radiographs reveal an alpha angle of 65 degrees and a positive crossover sign. What is the most accurate description of his pathology?

. Isolated Cam impingement
. Isolated Pincer impingement
. Combined Cam and Pincer impingement
. Ischiofemoral impingement
. Subspine impingement

Correct Answer & Explanation

. Combined Cam and Pincer impingement


Explanation

Femoroacetabular impingement (FAI) is typically evaluated radiographically. An elevated alpha angle (>50-55 degrees) is indicative of a Cam lesion, which is an aspherical deformity of the femoral head-neck junction. The crossover sign on an anteroposterior pelvis radiograph indicates focal cranial retroversion of the acetabulum, typical of Pincer impingement. The presence of both findings indicates combined (mixed) FAI, which is the most common clinical presentation.

Question 4803

Topic: Knee Sports

A 16-year-old female dancer experiences her third lateral patellar dislocation. Conservative management and physical therapy have failed. Radiographs show a Caton-Deschamps index of 1.1 and a sulcus angle of 135 degrees. A CT scan reveals a tibial tubercle-trochlear groove (TT-TG) distance of 22 mm. Which of the following surgical interventions is most appropriate to restore stability?

. Isolated medial patellofemoral ligament (MPFL) reconstruction
. Isolated lateral retinacular release
. MPFL reconstruction combined with medializing tibial tubercle osteotomy
. Distal femoral derotational osteotomy
. Trochleoplasty

Correct Answer & Explanation

. MPFL reconstruction combined with medializing tibial tubercle osteotomy


Explanation

The patient has recurrent patellar instability with a significantly elevated TT-TG distance (normal is <15 mm; >20 mm is widely considered a biomechanical indication for a medializing tibial tubercle osteotomy). MPFL reconstruction performed in isolation in the presence of an elevated TT-TG distance places excessive tension on the graft, leading to a high risk of graft failure and recurrent instability. Combined MPFL reconstruction and medializing TTO is the most appropriate management.

Question 4804

Topic: 5. Sports Medicine

A 10-year-old skeletally immature male sustains a mid-substance complete ACL tear. He has wide-open physes with an estimated 5 years of growth remaining. Nonoperative management is attempted, but he experiences recurrent instability episodes, prompting surgical intervention. Which of the following surgical techniques poses the highest risk for iatrogenic angular limb deformity or growth arrest in this patient?

. Iliotibial band extra-articular tenodesis (MacIntosh procedure)
. Physeal-sparing all-epiphyseal reconstruction
. Transphyseal reconstruction using a 10 mm bone-patellar tendon-bone (BTB) autograft
. Transphyseal reconstruction using a soft-tissue hamstring autograft
. Partial transphyseal reconstruction (epiphyseal femoral and transphyseal tibial tunnel)

Correct Answer & Explanation

. Transphyseal reconstruction using a soft-tissue hamstring autograft


Explanation

Using a bone block (such as a bone-patellar tendon-bone autograft) across an open physis significantly increases the risk of premature physeal closure, growth arrest, or angular deformity due to the formation of a rigid bony bridge across the growth plate. Transphyseal reconstruction using soft-tissue grafts (like hamstrings) carries a much lower risk, provided the tunnels are strictly vertical, appropriately sized, and fixation hardware does not cross the physis.

Question 4805

Topic: 5. Sports Medicine

A 15-year-old female high school soccer player sustains an anterior cruciate ligament (ACL) tear. Examination reveals a grade 3 Lachman test, a positive pivot shift, and generalized ligamentous laxity (Beighton score 6/9). Radiographs show closed physes. She wishes to return to competitive soccer. Which of the following graft choices is most strongly associated with the lowest risk of revision surgery in this patient profile?

. Quadrupled hamstring tendon autograft
. Bone-patellar tendon-bone autograft
. Tibialis anterior allograft
. Achilles tendon allograft
. Quadriceps tendon autograft with suspensory fixation

Correct Answer & Explanation

. Bone-patellar tendon-bone autograft


Explanation

In young, highly active female athletes with generalized ligamentous laxity, hamstring autografts have been shown to have a significantly higher failure rate and risk of revision compared to bone-patellar tendon-bone (BTB) autografts. Allografts have an unacceptably high failure rate in young, active patients and are contraindicated in this demographic. BTB autograft provides rigid bone-to-bone fixation and has historically demonstrated the lowest revision rates in high-risk groups, including young females with hyperlaxity.

Question 4806

Topic: Shoulder & Hip Sports

A 25-year-old male undergoes a Latarjet procedure for recurrent anterior shoulder instability with 25% glenoid bone loss. Postoperatively, he presents with profound weakness in elbow flexion and decreased sensation over the lateral aspect of his forearm. Which of the following intraoperative maneuvers most likely caused this neurologic injury?

. Excessive medial retraction of the conjoint tendon
. Placement of screws penetrating the posterior glenoid cortex
. Inferior capsular release at the 6 o'clock position
. Osteotomy of the coracoid process at its base
. Splitting of the subscapularis muscle in line with its fibers

Correct Answer & Explanation

. Excessive medial retraction of the conjoint tendon


Explanation

The patient's presentation of elbow flexion weakness and lateral forearm sensory deficit is classic for a musculocutaneous nerve injury. The musculocutaneous nerve is the most frequently injured nerve during a Latarjet procedure. It typically enters the conjoint tendon 3 to 8 cm distal to the tip of the coracoid. Excessive medial and distal retraction of the conjoint tendon places significant traction on the musculocutaneous nerve, leading to neuropraxia or structural injury. Inferior capsular release endangers the axillary nerve, while posterior screw penetration puts the suprascapular nerve at risk.

Question 4807

Topic: Shoulder & Hip Sports

A 26-year-old male ice hockey player presents with insidious onset right groin pain, worsened by deep flexion and internal rotation. Examination demonstrates a positive FADIR test. Radiographs reveal a prominent bony bump at the anterolateral femoral head-neck junction with an alpha angle of 65 degrees. He undergoes arthroscopic osteochondroplasty for a cam deformity. During the resection of the femoral neck deformity, over-resection of the head-neck junction poses the greatest risk for which of the following complications?

. Avascular necrosis of the femoral head
. Iatrogenic femoral neck fracture
. Heterotopic ossification
. Injury to the lateral femoral cutaneous nerve
. Ischiofemoral impingement

Correct Answer & Explanation

. Iatrogenic femoral neck fracture


Explanation

Arthroscopic osteochondroplasty is indicated for symptomatic cam femoroacetabular impingement (FAI). Over-resection of the cam deformity significantly increases the risk of a postoperative iatrogenic femoral neck fracture. Biomechanical studies recommend resecting no more than 30% of the anterolateral femoral neck diameter to maintain structural integrity. Avascular necrosis is primarily a risk if the retinacular vessels (branches of the medial femoral circumflex artery) are damaged, which are located more posterosuperiorly, not typically at the primary site of anterolateral cam resection.

Question 4808

Topic: Knee Sports

A 10-year-old boy complains of a clunking sensation and pain in his lateral right knee. He has no history of trauma. MRI confirms the diagnosis of a symptomatic complete discoid lateral meniscus. There is no meniscal tear. What is the most appropriate surgical treatment?

. Total meniscectomy
. Nonoperative management with physical therapy
. Arthroscopic saucerization with preservation of a peripheral rim
. Arthroscopic meniscal repair
. Anterior cruciate ligament reconstruction

Correct Answer & Explanation

. Arthroscopic saucerization with preservation of a peripheral rim


Explanation

The standard surgical treatment for a symptomatic complete discoid lateral meniscus without an unstable tear or peripheral detachment is arthroscopic saucerization. The goal is to reshape the meniscus to a more normal, crescentic configuration while preserving a stable peripheral rim (about 6-8 mm) to maintain its shock-absorbing function. Total meniscectomy is avoided due to the high risk of early osteoarthritis.

Question 4809

Topic: 5. Sports Medicine

A 6-year-old boy presents with a 6-month history of a painless snapping sound in his right knee. Examination reveals a palpable clunk at 20 degrees of flexion during extension of the knee. MRI confirms a complete discoid lateral meniscus with no evidence of a meniscal tear. What is the most appropriate management?

. Total meniscectomy
. Partial meniscectomy (saucerization)
. Partial meniscectomy and meniscal repair
. Observation
. Diagnostic arthroscopy

Correct Answer & Explanation

. Observation


Explanation

An incidental or completely asymptomatic discoid meniscus, or one that presents solely with a painless snap ('snapping knee syndrome'), requires observation. Surgical intervention (such as saucerization and/or repair) is reserved for patients who are symptomatic with pain, locking, mechanical symptoms, or MRI evidence of a tear.

Question 4810

Topic: Knee Sports
A 12-year-old boy presents with a painful, swollen knee after falling from a bicycle. Radiographs reveal a completely displaced (Meyers-McKeever Type III) tibial eminence fracture. Attempts at closed reduction in full extension fail to anatomically reduce the fragment. Which structure is most commonly entrapped beneath the fragment, blocking reduction?
. Anterior horn of the medial meniscus
. Posterior horn of the medial meniscus
. Anterior cruciate ligament fibers
. Posterior cruciate ligament
. Medial collateral ligament

Correct Answer & Explanation

. Anterior horn of the medial meniscus


Explanation

Tibial eminence (spine) fractures represent an avulsion of the anterior cruciate ligament insertion in children. When a Type III (completely displaced) fracture cannot be reduced closed, the most common anatomic block to reduction is the entrapment of the anterior horn of the medial meniscus (or the transverse intermeniscal ligament) under the bony fragment. Operative intervention (arthroscopic or open) is required to free the entrapped tissue and fix the fragment.

Question 4811

Topic: 5. Sports Medicine

A 14-year-old boy presents with chronic anteromedial knee pain. An MRI is obtained which demonstrates a 2.5 x 2.5 cm osteochondritis dissecans (OCD) lesion on the lateral aspect of the medial femoral condyle. T2-weighted sequences show high signal intensity fluid completely encircling the bony lesion.

Diagnostic arthroscopy reveals a ballottable but macroscopically intact articular cartilage surface. What is the most appropriate surgical treatment?

. Arthroscopic transarticular drilling of the lesion alone
. In situ fixation of the lesion with bioabsorbable or headless metallic screws
. Excision of the fragment with microfracture of the bed
. Osteochondral autograft transfer (OATS)
. Matrix-induced autologous chondrocyte implantation (MACI)

Correct Answer & Explanation

. In situ fixation of the lesion with bioabsorbable or headless metallic screws


Explanation

This patient has an unstable but intact osteochondritis dissecans (OCD) lesion. MRI evidence of fluid tracking behind the lesion indicates instability. Because the articular cartilage is intact and the patient is near skeletal maturity (where healing potential decreases compared to younger children), surgical stabilization is indicated. In situ fixation (using bioabsorbable pins/screws or headless metallic screws) promotes healing by compressing the unstable fragment into its bed while preserving the native articular cartilage.

Question 4812

Topic: Knee Sports

A 12-year-old boy presents with vague, poorly localized knee pain and occasional catching. Radiographs demonstrate an osteochondritis dissecans (OCD) lesion.

In which of the following anatomic locations is an OCD lesion of the knee most classically found?

. Lateral aspect of the medial femoral condyle
. Medial aspect of the lateral femoral condyle
. Central portion of the medial femoral condyle
. Patellar articular surface
. Tibial plateau

Correct Answer & Explanation

. Lateral aspect of the medial femoral condyle


Explanation

Osteochondritis dissecans (OCD) of the knee is most commonly found on the lateral aspect of the medial femoral condyle. This classic location accounts for approximately 70-80% of all knee OCD lesions. A helpful mnemonic is 'LAME': Lateral Aspect of the Medial Epicondyle/condyle. The lateral femoral condyle is the second most common site, usually on the inferocentral aspect.

Question 4813

Topic: Shoulder & Hip Sports

A 24-year-old hockey player presents with persistent anterior groin pain exacerbated by hip flexion.

An AP pelvis radiograph reveals a prominent 'crossover sign'. What is the primary pathomorphology associated with this radiographic finding?

. Decreased femoral head-neck offset
. Acetabular retroversion
. Coxa profunda
. Acetabular protrusion
. Coxa vara

Correct Answer & Explanation

. Acetabular retroversion


Explanation

The crossover sign on an AP pelvis radiograph indicates acetabular retroversion, where the anterior wall projects more laterally than the posterior wall in the cranial aspect of the joint. This finding is a hallmark of pincer-type femoroacetabular impingement (FAI).

Question 4814

Topic: Shoulder & Hip Sports

A 22-year-old elite hockey player presents with chronic, activity-limiting groin pain. An AP pelvis radiograph demonstrates a "crossover sign" and projection of the ischial spine into the pelvic basin. These radiographic findings are most indicative of which pathology?

. Cam-type femoroacetabular impingement
. Acetabular retroversion causing pincer-type impingement
. Developmental dysplasia of the hip
. Excessive femoral anteversion
. Avascular necrosis of the femoral head

Correct Answer & Explanation

. Acetabular retroversion causing pincer-type impingement


Explanation

The crossover sign (anterior wall crossing lateral to the posterior wall) and the ischial spine sign (visibility of the ischial spine inside the pelvic ring on an AP radiograph) are classic radiographic indicators of focal or global acetabular retroversion, causing pincer-type femoroacetabular impingement.

Question 4815

Topic: Knee Sports

A 22-year-old woman presents with recurrent lateral patellar dislocations after failing 6 months of targeted physical therapy.

Advanced imaging demonstrates a tibial tubercle-trochlear groove (TT-TG) distance of 23 mm and a Caton-Deschamps index of 1.35. Which of the following is the most appropriate surgical treatment?

. Medial patellofemoral ligament (MPFL) reconstruction alone
. MPFL reconstruction combined with distalization and medialization of the tibial tubercle
. MPFL reconstruction combined with anteriorization of the tibial tubercle
. Lateral retinacular release alone
. Sulcus-deepening trochleoplasty alone

Correct Answer & Explanation

. MPFL reconstruction combined with distalization and medialization of the tibial tubercle


Explanation

This patient has significant patella alta (Caton-Deschamps index > 1.2) and an abnormally elevated TT-TG distance (> 20 mm) contributing to instability. Management requires MPFL reconstruction to restore the primary medial restraint, combined with a tibial tubercle osteotomy (distalization and medialization) to correct the anatomic risk factors.

Question 4816

Topic: Shoulder & Hip Sports

A 24-year-old male athlete presents with deep anterior groin pain exacerbated by hip flexion and internal rotation. An AP pelvis radiograph demonstrates a crossover sign and a prominent ischial spine sign. The alpha angle on the lateral view is 45 degrees. These radiographic findings are most consistent with which of the following pathomorphologies?

. Cam femoroacetabular impingement
. Pincer impingement due to acetabular retroversion
. Pincer impingement due to coxa profunda
. Developmental dysplasia of the hip
. Slipped capital femoral epiphysis

Correct Answer & Explanation

. Pincer impingement due to acetabular retroversion


Explanation

The crossover sign (where the anterior wall of the acetabulum crosses the posterior wall) and a prominent ischial spine sign are classic radiographic features of acetabular retroversion. Acetabular retroversion leads to focal anterior overcoverage of the femoral head, causing Pincer-type femoroacetabular impingement (FAI). An alpha angle of 45 degrees is normal (<50-55 degrees), making Cam impingement unlikely. Coxa profunda is characterized by the acetabular fossa medial to the ilioischial line.

Question 4817

Topic: Shoulder & Hip Sports

A 28-year-old professional hockey player reports deep anterior groin pain that is exacerbated by hip flexion and internal rotation. An anteroposterior radiograph of the pelvis demonstrates a 'crossover sign'.

What is the primary pathophysiologic mechanism responsible for this patient's condition?

. Aspherical contour of the femoral head-neck junction causing cam impingement
. Focal retroversion of the acetabulum causing pincer impingement
. Global excessive anteversion of the acetabulum
. Coxa profunda with medialization of the joint center
. Excessive femoral anteversion

Correct Answer & Explanation

. Focal retroversion of the acetabulum causing pincer impingement


Explanation

A 'crossover sign' on an AP pelvis radiograph indicates that the anterior wall of the acetabulum crosses over the posterior wall before reaching the lateral edge of the acetabular roof. This radiographic finding is pathognomonic for focal or global acetabular retroversion, which leads to pincer-type femoroacetabular impingement (FAI) due to anterior overcoverage of the femoral head.

Question 4818

Topic: Shoulder & Hip Sports

A 24-year-old collegiate hockey player presents with chronic anterior groin pain exacerbated by hip flexion and internal rotation. Anteroposterior pelvis radiographs reveal a prominent crossover sign and an ischial spine sign. Which of the following best describes the pathomorphology contributing to this patient's impingement?

. Decreased femoral head-neck offset
. Acetabular retroversion
. Coxa vara
. Global acetabular overcoverage (Coxa profunda)
. An alpha angle greater than 55 degrees

Correct Answer & Explanation

. Acetabular retroversion


Explanation

The crossover sign (where the anterior rim of the acetabulum crosses the posterior rim on an AP pelvis radiograph) and the ischial spine sign (visibility of the ischial spines medial to the pelvic brim) are classic radiographic indicators of acetabular retroversion. This structural abnormality causes pincer-type femoroacetabular impingement (FAI). An alpha angle >55 degrees and decreased head-neck offset are indicative of cam-type impingement.

Question 4819

Topic: Knee Sports

A 28-year-old male sustains a direct blow to the anteromedial aspect of his proximal tibia while his knee is flexed. Physical examination reveals increased external rotation of the tibia compared to the contralateral side when tested at 30 degrees of knee flexion, but symmetric external rotation when tested at 90 degrees of knee flexion. Which of the following structures is most likely injured?

. Anterior cruciate ligament
. Posterior cruciate ligament
. Posterolateral corner
. Posteromedial corner
. Superficial medial collateral ligament

Correct Answer & Explanation

. Posterolateral corner


Explanation

The dial test is used to evaluate the posterolateral corner (PLC) and the posterior cruciate ligament (PCL). Increased external rotation of more than 10 degrees compared to the uninjured knee at 30 degrees of flexion, but symmetric external rotation at 90 degrees of flexion, is indicative of an isolated PLC injury. If the external rotation was increased at both 30 and 90 degrees, it would suggest a combined PLC and PCL injury.

Question 4820

Topic: Knee Sports

In a posterior-stabilized (PS) total knee arthroplasty, the cam-post mechanism is designed to mechanically substitute for the resected posterior cruciate ligament (PCL). What is the primary kinematic function of this mechanism during deep knee flexion?

. To induce femoral rollback and prevent anterior translation of the femur on the tibia
. To limit excessive internal rotation of the tibia
. To provide varus-valgus stability in deep flexion
. To prevent posterior subluxation of the tibia in extension
. To force paradoxical anterior sliding of the femur on the tibia

Correct Answer & Explanation

. To induce femoral rollback and prevent anterior translation of the femur on the tibia


Explanation

In a native knee, the PCL causes the femur to roll back posteriorly on the tibia during flexion, which prevents anterior translation of the femur and maximizes the lever arm of the extensor mechanism, aiding in deep flexion. In a posterior-stabilized (PS) TKA, the PCL is excised. To replicate this kinematic function, a cam on the femoral component engages a post on the tibial polyethylene insert during mid-to-deep flexion. This engagement forces mandatory femoral rollback and prevents the femur from translating anteriorly (paradoxical anterior sliding), a phenomenon that can occur in cruciate-retaining knees with a non-functional PCL.