This practice set contains high-yield board review questions covering key concepts in Knee Sports. Each clinical scenario is designed to test your diagnostic and management skills relevant to this subspecialty.
Question 321
Topic: Knee Sports
A 29-year-old male is undergoing arthroscopic-assisted ACL reconstruction following a knee dislocation. The surgeon is preparing to pass the graft through the femoral and tibial tunnels. The image below depicts a common technique for graft fixation during this procedure. Based on the image and the case description, what is the primary purpose of the structure being fixed in the femoral tunnel?
Correct Answer & Explanation
. To resist anterior translation of the tibia relative to the femur.
Explanation
Correct Answer: CThe image provided shows an ACL reconstruction with a graft being fixed in the femoral tunnel, likely with an Endobutton or similar cortical suspension device. The question context specifies ACL reconstruction. The case explicitly states that the 'Anterior Cruciate Ligament (ACL): Resists anterior translation of the tibia relative to the femur and secondary valgus/varus and internal/external rotation.' Therefore, the primary purpose of the reconstructed ACL, and thus the graft being fixed, is to resist anterior translation of the tibia relative to the femur.Option A (To provide primary resistance to posterior tibial translation) is the primary function of the PCL, not the ACL.Option B (To reconstruct the posterolateral corner (PLC) of the knee) is incorrect. PLC reconstruction addresses varus and external rotation instability, distinct from ACL function.Option D (To stabilize the medial collateral ligament (MCL) complex) is incorrect. The MCL is the primary restraint to valgus stress.Option E (To repair a meniscal bucket-handle tear) is incorrect. Meniscal repair addresses meniscal pathology, not primary ligamentous stability.
Question 322
Topic: Knee Sports
A 33-year-old active duty military personnel sustains a multi-ligamentous knee injury (ACL, PCL, and MCL tears) during a training exercise. After successful reduction and confirmation of vascular integrity, the orthopedic surgeon discusses the timing of definitive ligament reconstruction. According to current literature and guidelines, what is the most common approach regarding the timing of multi-ligament knee reconstruction?
Correct Answer & Explanation
. Acute reconstruction (within 1-3 weeks) after resolution of swelling and inflammation.
Explanation
Correct Answer: CThe case discusses the timing of ligament reconstruction: 'Timing of Surgery: ...Acute (within 1-3 weeks): Many surgeons prefer early ligament repair/reconstruction to capitalize on tissue healing potential and facilitate rehabilitation, provided soft tissue swelling has subsided.' It further states under 'Summary of Key Literature and Guidelines': 'Current Consensus: The trend leans towards early surgical intervention for ligamentous reconstruction (within 1-3 weeks) once swelling has subsided and the knee has achieved a relatively quiescent state.'Option A (Immediate reconstruction) is generally reserved for vascular repair, open dislocations, or irreducible dislocations. For ligamentous reconstruction, waiting for soft tissue swelling to subside (1-3 weeks) is often preferred to reduce the risk of arthrofibrosis and improve surgical conditions.Option B (Delayed reconstruction 6-12 months) is typically too late and can lead to chronic instability, muscle atrophy, and increased difficulty in reconstruction due to scar tissue and retraction.Option D (Non-operative management) is rarely indicated for true multi-ligament knee dislocations, as it almost universally leads to significant long-term instability and functional deficits in active individuals.Option E (Staged reconstruction at 3-month intervals) is a possibility for very complex cases or specific patient factors, but the 'acute' (1-3 weeks) approach for all ligaments in a single setting (or closely staged) is more common once the acute inflammatory phase has passed.
Question 323
Topic: Knee Sports
A 30-year-old professional dancer presents with a 5-year history of progressive pain and stiffness in her right great toe, significantly impacting her career. She has failed all conservative measures. Radiographs show moderate dorsal osteophytes, mild joint space narrowing, and a focal osteochondral lesion on the metatarsal head, consistent with Coughlin and Shurnas Grade 2-3. She is highly motivated to preserve motion. Based on the case, which surgical approach, potentially combined with cheilectomy, would be most appropriate for this patient?
Correct Answer & Explanation
. D. Cartilage restoration technique (e.g., microfracture) for the focal defect.
Explanation
Correct Answer: DThe patient is a young, active professional dancer with a focal osteochondral lesion and a strong desire to preserve motion. The case, under 'Cartilage Restoration Techniques (Microfracture, OATS, MACI),' states: 'Indication: For younger patients with focal cartilage defects (e.g., osteochondral lesions) associated with early hallux rigidus, where the surrounding cartilage is relatively healthy. Not suitable for diffuse, end-stage arthritis.' Given her age, activity level, and the focal nature of the lesion, a cartilage restoration technique combined with cheilectomy (to address the osteophytes) is the most appropriate motion-preserving option.A. Arthrodesis sacrifices motion and is the gold standard for end-stage (Grade 4) HR, which would be detrimental to a professional dancer.B. Hemiarthroplasty is for Grade 3-4 HR, typically in older, less active individuals, and has concerns regarding implant loosening and contralateral cartilage wear, making it less ideal for a young, high-demand patient with a focal defect.C. Interposition arthroplasty is for Grade 3-4 HR, with variable outcomes and concerns about graft durability, and is not specifically indicated for focal lesions in young patients.E. Resection arthroplasty (Keller procedure) is not mentioned in the case as a recommended procedure for hallux rigidus. It is an older procedure associated with significant shortening and instability, generally considered a salvage procedure and not suitable for a high-demand patient seeking motion preservation.
Question 324
Topic: Knee Sports
A 2-year-old child with fibular hemimelia is being evaluated for lower extremity reconstruction versus amputation. Along with the fibular absence and limb length discrepancy, which of the following knee ligament anomalies is most commonly associated with this condition?
Correct Answer & Explanation
. Absent anterior cruciate ligament
Explanation
Fibular deficiency is highly associated with structural abnormalities of the entire lower limb, including the knee. The most common ligamentous anomaly in the knee is an absent or hypoplastic anterior cruciate ligament (ACL).
Question 325
Topic: Knee Sports
A 2-year-old child presents with a foreshortened leg, absent lateral digits, and an anteromedial bow of the tibia. Which of the following knee anomalies is most universally associated with this patient's underlying diagnosis?
Correct Answer & Explanation
. ACL deficiency
Explanation
The patient has fibular hemimelia (longitudinal fibular deficiency). Anterior cruciate ligament (ACL) deficiency or complete absence is consistently observed in patients with fibular hemimelia, though they are often clinically asymptomatic due to compensatory joint geometry.
Question 326
Topic: Knee Sports
Which of the following describes the anatomical structure primarily responsible for providing a 'rotary stability' to the knee, especially against external tibial rotation in a valgus stress?
Correct Answer & Explanation
. Posteromedial corner (PMC) structures
Explanation
Correct Answer: EThe posteromedial corner (PMC) of the knee, which includes the superficial MCL, posterior oblique ligament, and semimembranosus attachments, is crucial for controlling valgus and external rotational stability, particularly at 30 degrees of knee flexion. While the MCL resists valgus stress, the complete 'rotary stability' against combined valgus and external rotation (often seen in complex knee injuries) is attributed to the integrity of the PMC. The ACL primarily resists anterior translation and internal rotation. The LCL resists varus stress. The PCL resists posterior translation.
Question 327
Topic: Knee Sports
A 25-year-old male undergoes an anterior cruciate ligament (ACL) reconstruction using a bone-patellar tendon-bone autograft. Postoperatively, he lacks full knee flexion but easily achieves full extension. The surgeon suspects malpositioning of the femoral tunnel. Where was the femoral tunnel most likely placed?
Correct Answer & Explanation
. Too anterior
Explanation
An anteriorly placed femoral tunnel during ACL reconstruction results in a graft that is tight in flexion and loose in extension, leading to a loss of knee flexion. Conversely, a posterior femoral tunnel leads to a graft that is tight in extension.
Question 328
Topic: Knee Sports
A 27-year-old soccer player sustains a PCL injury after falling onto a flexed knee. On examination, there is a slight sag and posterior translation of the tibia less than one centimeter. The initial management of this isolated PCL injury should primarily emphasize:
Correct Answer & Explanation
. Quadriceps strengthening
Explanation
Correct Answer: DFor isolated PCL injuries, especially those with minimal posterior translation, non-operative management is often successful. The primary focus of rehabilitation is quadriceps strengthening. The quadriceps muscle provides an anteriorly directed force on the tibia, which can dynamically stabilize the knee and counteract the posterior laxity caused by the PCL injury. Hamstring strengthening, while important for overall knee health, can exacerbate posterior translation. Strict immobilization can lead to stiffness and atrophy. Open chain exercises may place undue stress on the PCL. Surgical reconstruction is typically reserved for higher-grade PCL injuries or combined ligamentous injuries.
Question 329
Topic: Knee Sports
A 28-year-old male presents with persistent loss of terminal extension six months after an anterior cruciate ligament (ACL) reconstruction. Lateral radiographs reveal that the entire tibial tunnel is positioned entirely anterior to Blumensaat's line when the knee is in full extension. Which of the following complications is most likely causing his loss of motion?
Correct Answer & Explanation
. Roof impingement
Explanation
Anterior placement of the tibial tunnel leads to graft impingement against the intercondylar roof (Blumensaat's line) during terminal extension. This biomechanical conflict typically presents as a significant loss of extension and potential graft attrition.
Question 330
Topic: Knee Sports
A patient complains of severe knee stiffness and an inability to flex the knee past 80 degrees following an ACL reconstruction. Intraoperative notes and postoperative imaging confirm that the femoral tunnel was drilled significantly anterior (high) to the anatomic footprint. What is the kinematic consequence of this technical error?
Correct Answer & Explanation
. The graft tightens excessively in deep flexion
Explanation
A femoral tunnel placed too anteriorly (non-anatomic) results in the graft tightening significantly as the knee flexes. This "captures" the joint and severely limits knee flexion, often requiring revision surgery if conservative management fails.
Question 331
Topic: Knee Sports
The posterior cruciate ligament (PCL) is the primary restraint to posterior tibial translation and is composed of two distinct functional bundles. Which of the following best describes their biomechanical behavior during knee motion?
Correct Answer & Explanation
. The anterolateral bundle is tightest in flexion and lax in extension
Explanation
The PCL consists of the larger anterolateral (AL) bundle and the smaller posteromedial (PM) bundle. The AL bundle is tight in flexion, whereas the PM bundle is tight in extension.
Question 332
Topic: Knee Sports
A 25-year-old athlete sustains a twisting injury to the knee. On physical examination, the dial test demonstrates a 15-degree increase in external rotation compared to the uninjured contralateral knee at 30 degrees of flexion. At 90 degrees of flexion, the external rotation is symmetric bilaterally. What is the most likely injury pattern?
Correct Answer & Explanation
. Isolated Posterolateral Corner (PLC) injury
Explanation
An increase in external rotation greater than 10 degrees at 30 degrees of flexion, with symmetric rotation at 90 degrees, indicates an isolated posterolateral corner (PLC) injury. If rotation is increased at both 30 and 90 degrees, a combined PLC and PCL injury is present.
Question 333
Topic: Knee Sports
Which bundle of the posterior cruciate ligament (PCL) is tightest in deep knee flexion?
Correct Answer & Explanation
. Anterolateral bundle
Explanation
The PCL consists of the anterolateral (AL) and posteromedial (PM) bundles. The larger AL bundle is tightest in knee flexion, whereas the smaller PM bundle is tightest in knee extension.
Question 334
Topic: Knee Sports
A patient presents with a knee injury. The dial test shows 15 degrees of increased external rotation at 30 degrees of knee flexion compared to the contralateral side, but symmetric rotation at 90 degrees. What is the most likely diagnosis?
Correct Answer & Explanation
. Isolated posterolateral corner injury
Explanation
An increase in external rotation of more than 10 degrees at 30 degrees of flexion, but not at 90 degrees, indicates an isolated posterolateral corner (PLC) injury. Increased external rotation at both 30 and 90 degrees indicates a combined PLC and PCL injury.
Question 335
Topic: Knee Sports
A 45-year-old female sustains a medial meniscus posterior root tear. Biomechanically, the loss of hoop stresses in this condition is most equivalent to which of the following?
Correct Answer & Explanation
. Total medial meniscectomy
Explanation
A complete meniscal root tear disrupts the circumferential hoop stresses that normally distribute axial loads. Biomechanically, this derangement results in increased contact pressures equivalent to a total meniscectomy.
Question 336
Topic: Knee Sports
During reconstruction of the medial patellofemoral ligament (MPFL), identifying the anatomic femoral attachment is critical. According to Schöttle, where is this point located radiographically?
Correct Answer & Explanation
. Anterior to the posterior femoral cortex line, between the adductor tubercle and medial epicondyle
Explanation
Schöttle's point is the radiographic landmark for the femoral origin of the MPFL. It is located 1 mm anterior to the posterior cortex line, 2.5 mm distal to the posterior origin of the medial femoral condyle, and proximal to the medial epicondyle.
Question 337
Topic: Knee Sports
A patient presents with a painful lack of terminal knee extension 4 months following an ACL reconstruction. MRI demonstrates a nodular mass in the anterior intercondylar notch. What is the histologic composition of this mass?
Correct Answer & Explanation
. Fibrovascular and granulation tissue
Explanation
The patient has a "Cyclops lesion," which is a localized form of anterior arthrofibrosis that blocks terminal extension. Histologically, it is composed of a dense nodule of fibrovascular and granulation tissue.
Question 338
Topic: Knee Sports
The classic, most common anatomic location for an osteochondritis dissecans (OCD) lesion of the knee is the:
Correct Answer & Explanation
. Lateral aspect of the medial femoral condyle
Explanation
Approximately 70% to 80% of knee OCD lesions occur on the lateral aspect of the medial femoral condyle. This classic location is often remembered by the mnemonic "LAME" (Lateral Aspect Medial Epicondyle/Condyle).
Question 339
Topic: Knee Sports
During an ACL reconstruction, placing the tibial tunnel too far anteriorly most commonly leads to which of the following complications?
Correct Answer & Explanation
. Roof impingement and loss of terminal extension
Explanation
An anteriorly placed tibial tunnel causes the graft to impinge against the roof of the intercondylar notch (Blumensaat's line) during knee extension. This results in a mechanical block to terminal extension and potential early graft failure.
Question 340
Topic: Knee Sports
If the femoral tunnel during an ACL reconstruction is placed too far anteriorly (shallow) in the intercondylar notch, what resulting kinematic mismatch will occur?
Correct Answer & Explanation
. The graft will be tight in flexion and loose in extension
Explanation
A femoral tunnel placed too far anteriorly (shallow) is non-anatomic. This positioning causes the graft to artificially tighten as the knee goes into flexion and become abnormally loose in full extension.