Question 2721

Topic: 3. Adult Reconstruction (Hip & Knee)

During a posterior-stabilized total knee arthroplasty, the surgeon uses spacer blocks to assess gap kinematics. With the trial components in place, the extension gap is perfectly symmetric and stable. However, when the knee is flexed to 90 degrees, the joint is grossly unstable both medially and laterally. Which of the following technical errors is the most likely cause of this isolated flexion gap looseness?

. Excessive distal femoral resection
. Undersized femoral component
. Oversized femoral component
. Inadequate distal femoral resection
. Excessive proximal tibial resection

Correct Answer & Explanation

. Undersized femoral component


Explanation

In total knee arthroplasty, the size of the femoral component dictates the anteroposterior (AP) dimension of the femur, which directly influences the flexion gap. An undersized femoral component decreases the AP dimension, leading to a loose flexion gap. Because the distal femoral resection is unchanged, the extension gap remains stable.

Question 2722

Topic: 3. Adult Reconstruction (Hip & Knee)

During a posterior-stabilized total knee arthroplasty using an anterior referencing system, trial components are inserted. The knee is stable and symmetric in extension, but the flexion gap is unacceptably tight, preventing flexion beyond 90 degrees. What is the most appropriate next step to balance the knee?

. Resect additional distal femur
. Downsize the femoral component
. Decrease the posterior tibial slope
. Release the superficial medial collateral ligament
. Upsize the femoral component

Correct Answer & Explanation

. Downsize the femoral component


Explanation

In a TKA, a tight flexion gap with a perfectly balanced extension gap can be addressed by downsizing the femoral component (which decreases the anteroposterior dimension of the femur without affecting the distal cut/extension gap) or by increasing the posterior tibial slope. Resecting more distal femur would loosen the extension gap.

Question 2723

Topic: 3. Adult Reconstruction (Hip & Knee)

A 68-year-old male with a metal-on-polyethylene total hip arthroplasty implanted 5 years ago presents with new-onset groin pain. MRI reveals a large cystic pseudotumor. Aspiration yields cloudy fluid with negative cultures but markedly elevated cobalt and chromium levels. What is the most likely source of the metal debris?

. Polyethylene wear debris
. Trunnionosis (mechanically assisted crevice corrosion)
. Third-body wear from retained cement
. Galvanic corrosion between the cup and screws
. Femoral stem fretting against the cortical bone

Correct Answer & Explanation

. Trunnionosis (mechanically assisted crevice corrosion)


Explanation

In a metal-on-polyethylene (MoP) total hip arthroplasty presenting with elevated metal ions and adverse local tissue reaction (ALTR/pseudotumor), the primary source of metal debris is wear and corrosion at the head-neck modular junction (taper), known as trunnionosis.

Question 2724

Topic: 3. Adult Reconstruction (Hip & Knee)

During a primary total knee arthroplasty, the surgeon assesses the gaps and notes the knee is tight in flexion but well-balanced and stable in extension. Which of the following intraoperative maneuvers is most appropriate to balance the knee?

. Recut the distal femur to remove more bone
. Release the posterior capsule
. Downsize the femoral component
. Decrease the tibial polyethylene thickness
. Release the superficial medial collateral ligament

Correct Answer & Explanation

. Downsize the femoral component


Explanation

A tight flexion gap with a balanced extension gap requires a reduction in the posterior femoral offset. Downsizing the femoral component (utilizing anterior referencing) removes more posterior femoral condylar bone, thereby increasing the flexion gap without affecting the extension gap.

Question 2725

Topic: 3. Adult Reconstruction (Hip & Knee)

A 65-year-old woman undergoes primary total hip arthroplasty via a posterior approach. Postoperatively, she experiences recurrent posterior dislocations. Radiographic evaluation shows the acetabular component in 45 degrees of abduction and 0 degrees of anteversion. Which of the following revision strategies is most appropriate?

. Revision of the femoral component to increase anteversion
. Revision of the acetabular component to increase anteversion
. Application of a constrained liner in the current shell
. Exchange to a larger femoral head
. Advancement of the greater trochanter

Correct Answer & Explanation

. Revision of the acetabular component to increase anteversion


Explanation

The acetabular component is relatively retroverted (0 degrees of anteversion instead of the normal 15-20 degrees), strongly predisposing the patient to posterior dislocation. The primary architectural cause of instability must be corrected; thus, revising the acetabular component to increase anteversion is required.

Question 2726

Topic: 3. Adult Reconstruction (Hip & Knee)

A 62-year-old female with a metal-on-metal total hip arthroplasty presents with groin pain and swelling. Laboratory tests show elevated serum cobalt and chromium levels. MRI reveals a large cystic mass communicating with the joint space. A revision surgery is performed. What is the classic histologic hallmark found in the periprosthetic tissue of this condition?

. Extensive sheets of neutrophils and intracellular bacteria
. Birefringent particulate wear debris surrounded by foreign body giant cells
. Perivascular lymphocytic infiltrate with plasma cells and macrophages
. Woven bone formation with prominent osteoblastic rimming
. Acellular hyaline cartilage with subchondral cysts

Correct Answer & Explanation

. Perivascular lymphocytic infiltrate with plasma cells and macrophages


Explanation

The patient has Aseptic Lymphocyte-Dominated Vasculitis-Associated Lesion (ALVAL), representing a Type IV delayed hypersensitivity reaction to metal ions (cobalt and chromium). The classic histologic hallmark is a diffuse or perivascular infiltrate of T-lymphocytes, plasma cells, and macrophages containing metal debris.

Question 2727

Topic: 3. Adult Reconstruction (Hip & Knee)

During a primary total knee arthroplasty in a patient with a severe fixed varus deformity, the surgeon must perform sequential medial releases to balance the knee in extension. After releasing the deep medial collateral ligament (MCL) and removing medial tibial osteophytes, the medial compartment remains tight in extension. What is the next most appropriate anatomic structure to release?

. Superficial medial collateral ligament (MCL)
. Pes anserinus tendons
. Posteromedial capsule and semimembranosus expansion
. Medial head of the gastrocnemius
. Posterior cruciate ligament (PCL)

Correct Answer & Explanation

. Posteromedial capsule and semimembranosus expansion


Explanation

The classic sequence for correcting a fixed varus deformity in TKA involves releasing structures from anterior to posterior and proximal to distal. The sequence generally is: 1) Deep MCL and removal of osteophytes. 2) Posteromedial capsule and semimembranosus insertions (to release tightness in extension). 3) Superficial MCL (subperiosteal peeling off the tibia). 4) Pes anserinus (if still tight, though rarely needed). Releasing the posteromedial corner specifically helps balance the knee in extension.

Question 2728

Topic: 3. Adult Reconstruction (Hip & Knee)

In total joint arthroplasty, the use of highly cross-linked ultra-high-molecular-weight polyethylene (HXLPE) has significantly reduced volumetric wear rates. However, the radiation used to create cross-links also generates free radicals that can cause oxidative degradation. To eliminate these free radicals without compromising the fatigue strength of the polyethylene, manufacturers currently utilize which of the following techniques?

. Remelting the polyethylene above its melting point (approx 150°C)
. Annealing the polyethylene just below its melting point
. Doping or blending the polyethylene with Vitamin E (alpha-tocopherol)
. Sterilization via gamma irradiation in an oxygen-rich environment
. Increasing the thickness of the polyethylene liner above 8 mm

Correct Answer & Explanation

. Doping or blending the polyethylene with Vitamin E (alpha-tocopherol)


Explanation

Irradiation creates cross-links (reducing wear) but leaves free radicals (causing oxidation). Remelting eliminates free radicals but decreases the mechanical fatigue strength and fracture toughness of the polyethylene. Annealing preserves strength but leaves some free radicals. Doping the HXLPE with Vitamin E (a potent antioxidant) neutralizes the free radicals without the need for post-irradiation thermal treatment (remelting), thereby preserving the mechanical strength and fatigue resistance of the material while maintaining excellent wear properties.

Question 2729

Topic: 3. Adult Reconstruction (Hip & Knee)

A 68-year-old man who underwent a total hip arthroplasty 5 years ago with a cobalt-chromium head and highly cross-linked polyethylene liner on a titanium stem presents with insidious onset of groin pain. Inflammatory markers are within normal limits. Aspiration yields fluid with 500 WBCs/uL (mostly mononuclear), but serum metal ions show isolated elevated cobalt levels. What is the most likely source of the cobalt?

. Bearing surface wear
. Trunnionosis (head-neck junction wear)
. Acetabular shell fretting
. Stem-cement interface micromotion
. Proximal stem coating delamination

Correct Answer & Explanation

. Trunnionosis (head-neck junction wear)


Explanation

In a metal-on-polyethylene total hip arthroplasty, elevated cobalt levels in the presence of a titanium stem and cobalt-chromium head are indicative of mechanically assisted crevice corrosion (trunnionosis) at the modular head-neck junction. The bearing surface is polyethylene, so it would not produce metal ions. A titanium stem would produce titanium ions, not cobalt. Thus, the modular cobalt-chromium head articulating with the titanium trunnion is the source.

Question 2730

Topic: 3. Adult Reconstruction (Hip & Knee)

Ceramic-on-ceramic (CoC) bearing surfaces in total hip arthroplasty are known for excellent wear characteristics. Which of the following is the most commonly cited cause of squeaking in a CoC THA?

. Edge loading due to component malposition
. Periprosthetic joint infection
. Metallosis from the trunnion
. Excessive femoral offset
. Use of a 36-mm femoral head

Correct Answer & Explanation

. Edge loading due to component malposition


Explanation

Squeaking is a specific complication of ceramic-on-ceramic bearings. The most common cause is edge loading, which typically results from component malposition (such as excessive acetabular cup anteversion or inclination). This malposition leads to loss of fluid-film lubrication, causing direct ceramic-on-ceramic contact at the rim, stripe wear, and the characteristic squeaking sound.

Question 2731

Topic: 3. Adult Reconstruction (Hip & Knee)

A 68-year-old woman is evaluated for a stiff and painful total knee arthroplasty (TKA) 2 years postoperatively. Her range of motion is 10 to 80 degrees. Radiographs show a well-fixed TKA, but the joint line is elevated, and the patella appears significantly inferior to the joint line. Which of the following intraoperative technical errors is the most likely cause of this presentation?

. Elevating the joint line by over-resecting the distal femur and using a thicker polyethylene insert
. Undersizing the femoral component
. Internal rotation of the tibial component
. Over-resecting the proximal tibia
. Placing the femoral component in excessive flexion

Correct Answer & Explanation

. Elevating the joint line by over-resecting the distal femur and using a thicker polyethylene insert


Explanation

Elevating the joint line often occurs when too much distal femur is resected, requiring a thicker polyethylene insert to balance the extension gap. This effectively moves the joint line proximally relative to the tibial tubercle, leading to relative (or pseudo-) patella baja. The patellar tendon is relatively shortened, causing the patella to impinge on the tibial polyethylene, which alters patellofemoral kinematics and results in anterior knee pain and decreased range of motion.

Question 2732

Topic: 3. Adult Reconstruction (Hip & Knee)

A 60-year-old female presents with progressive groin pain 7 years after undergoing an uncomplicated primary total hip arthroplasty. Her implants consist of a large-diameter cobalt-chromium head on a titanium stem with a highly cross-linked polyethylene liner. Radiographs show well-fixed components with no osteolysis. Laboratory analysis reveals significantly elevated serum cobalt levels with normal serum chromium levels. What is the most likely source of the elevated metal ions?

. Bearing surface wear
. Acetabular shell loosening
. Head-neck junction fretting and corrosion
. Stem-cement interface debonding
. Acetabular liner impingement

Correct Answer & Explanation

. Head-neck junction fretting and corrosion


Explanation

This patient is exhibiting signs of an Adverse Local Tissue Reaction (ALTR) due to mechanically assisted crevice corrosion (trunnionosis). In a metal-on-polyethylene THA, high serum cobalt combined with normal or low chromium strongly indicates fretting and corrosion at the modular head-neck junction. Bearing surface wear in this construct (metal on poly) does not release clinically significant systemic metal ions compared to metal-on-metal bearings.

Question 2733

Topic: 3. Adult Reconstruction (Hip & Knee)

A 65-year-old man presents with persistent right groin pain 5 years after undergoing a primary metal-on-polyethylene total hip arthroplasty. Aspiration of the hip yields clear fluid with a negative Gram stain and negative cultures. Laboratory workup reveals substantially elevated serum cobalt levels with normal chromium levels. Which of the following is the most likely source of the metal ions?

. Bearing surface wear
. Mechanically assisted crevice corrosion at the head-neck junction
. Acetabular shell micromotion
. Third-body wear from retained cement
. Femoral stem distal toggle

Correct Answer & Explanation

. Mechanically assisted crevice corrosion at the head-neck junction


Explanation

The scenario describes trunnionosis, which is mechanically assisted crevice corrosion occurring at the modular head-neck junction (the trunnion). In a metal-on-polyethylene total hip arthroplasty, the bearing surface does not contain metal-on-metal articulation, so elevated metal ions point to a non-bearing surface source. A disproportionately elevated cobalt level relative to chromium is a classic hallmark of corrosion at the cobalt-chromium femoral head and titanium stem junction.

Question 2734

Topic: 3. Adult Reconstruction (Hip & Knee)

A 62-year-old man presents with progressive right groin pain and a palpable anterior thigh mass 7 years after undergoing a primary total hip arthroplasty. His implants include a highly cross-linked polyethylene liner, a titanium alloy femoral stem, and a cobalt-chromium femoral head. Serum cobalt levels are 8.5 ppb and chromium is 1.2 ppb. Aspiration of the hip yields synovial fluid with 1,200 WBC/µL and 60% polymorphonuclear leukocytes. Cultures are negative at 7 days. Metal artifact reduction sequence (MARS) MRI demonstrates a large, thick-walled solid and cystic mass compressing the femoral vein. During revision surgery, severe mechanically assisted crevice corrosion (MACC) is noted at the modular head-neck junction. The femoral stem is found to be solidly fixed with no evidence of loosening. Which of the following is the most appropriate definitive management of the femoral side?

. Retain the stem, meticulously clean the trunnion, and place a new cobalt-chromium head
. Retain the stem, meticulously clean the trunnion, and place a ceramic head directly on the existing trunnion
. Retain the stem and place a ceramic head with a titanium taper sleeve
. Explant the stem via an extended trochanteric osteotomy and place a cemented cobalt-chromium stem
. Explant the stem and place a modular fluted tapered titanium stem

Correct Answer & Explanation

. Retain the stem and place a ceramic head with a titanium taper sleeve


Explanation

The patient presents with an adverse local tissue reaction (ALTR) secondary to mechanically assisted crevice corrosion (trunnionosis) at the head-neck junction, characterized by elevated cobalt levels disproportionate to chromium. When addressing trunnionosis during revision surgery, if the femoral stem is well-fixed, it should generally be retained to minimize surgical morbidity and bone loss. However, placing a new metal head on a damaged trunnion will invariably lead to recurrent corrosion. Placing a ceramic head directly onto a damaged trunnion is contraindicated due to the high risk of stress risers leading to catastrophic ceramic head fracture. The gold standard for a well-fixed stem with trunnion damage is the use of a ceramic head paired with a titanium taper sleeve. The titanium sleeve bypasses the damaged portion of the trunnion and provides a pristine, geometrically matched surface for the ceramic head, preventing further MACC and avoiding ceramic fracture.

Question 2735

Topic: 3. Adult Reconstruction (Hip & Knee)

A 75-year-old woman who fell on her right knee now reports pain and is unable to bear weight. History reveals that she underwent total knee arthroplasty on the right knee 6 years ago. Radiographs are shown in Figure 5. Management should now consist of

Orthopedic Surgery Board Review 2026 | High-Yield MCQs - Set 9 - Figure 50

. closed reduction and casting for 6 weeks.
. open reduction and internal fixation, using a locked intramedullary rod.
. open reduction and internal fixation, using two cancellous screws.
. open reduction and internal fixation, using a locked plate and screws.
. open reduction and internal fixation and revision of the femoral component.

Correct Answer & Explanation

. open reduction and internal fixation and revision of the femoral component.


Explanation

The radiographs show a loose femoral component with an associated medial condyle distal femoral fracture. The treatment of choice is open reduction and internal fixation with revision of the femoral component because of the femoral component loosening. Moran MC, Brick GW, Sledge CB, et al: Supracondylar femoral fracture following total knee arthroplasty. Clin Orthop 1996;324:196-209. McLaren AC, DuPont JA, Schroeber DC: Open reduction internal fixation of supracondylar fractures above total knee arthroplasties using the intramedullary supracondylar rod. Clin Orthop 1994;302:194-198.

Question 2736

Topic: 3. Adult Reconstruction (Hip & Knee)

Figures 3a and 3b show the current radiographs of a 58-year-old man who underwent total knee arthroplasty with a cruciate ligament sparing prosthesis 7 years ago. Examination reveals boggy synovitis and moderate pain, particularly anteriorly. Management should consist of

. follow-up radiographs.
. alendronate, with follow-up examinations every 6 months.
. revision to a posterior stabilized prosthesis.
. exchange of the tibial insert through a limited incision.
. surgical exploration with revision or exchange based on the findings.

Correct Answer & Explanation

. surgical exploration with revision or exchange based on the findings.


Explanation

The patient has symptoms of synovitis that are most likely the result of the release of particles from the tibial polyethylene. While observation may be warranted in a completely asymtomatic knee, some intervention is indicated for this patient as there is clear radiographic evidence of lysis in both the tibia and femur. The decision about the extent of the revision should be made at the time of surgery. A limited incision technique is not indicated. Grafting (or using graft substitute) the defect is the most appropriate approach for treating the osteolytic lesions. While a posterior stabilized prosthesis might be the solution, surgical findings might dictate otherwise.

Question 2737

Topic: 3. Adult Reconstruction (Hip & Knee)

Which of the following is considered the most predictive factor in determining whether a patient will need a blood transfusion after total knee arthroplasty?

. Patient's weight
. Gender
. Preoperative hemoglobin level
. Duration of surgery
. Number of medical comorbidities

Correct Answer & Explanation

. Preoperative hemoglobin level


Explanation

Multiple studies have shown that the preoperative hemoglobin level is the most predictive factor in determining whether a transfusion will be necessary after total hip or total knee arthroplasty. Studies have not shown any correlation with the other options. Salido JA, Marin LA, Gomez LA, Zorrilla P, Martinez C: Preoperative hemoglobin levels and the need for transfusion after prosthetic hip and knee surgery: Analysis of predictive factors. J Bone Joint Surg Am 2002;84:216-220. Sculco TP, Gallina J: Blood management experience: Relationship between autologous blood donation and transfusion in orthopedic surgery. Orthopedics 1999;22:S129-S134.

Question 2738

Topic: 3. Adult Reconstruction (Hip & Knee)

With the increasing availability of total hip arthroplasty (THA) to younger patients with hip osteoarthritis, there has been increased use of alternative bearing surfaces. Compared to a ceramic-on-ceramic articulation, which of the following is a specific advantage of a metal-on-metal bearing surface?

. Increased wettability
. Increased hardness
. Increased fracture toughness
. Decreased surface roughness
. Lower coefficient of friction

Correct Answer & Explanation

. Increased fracture toughness


Explanation

Alternative bearing surfaces in THA have received much attention in recent years as more and more hip arthroplasties are being performed on younger patients with hip arthritis. The two most popular nonmetal-on-polyethylene bearing surfaces are metal-on-metal and ceramic-on-ceramic. There are arguments supporting the use of either, but ceramic bearings have been shown to have a theoretic increased risk of fracture compared with cobalt-chromium. This has been shown to be clinically relevant with zirconium ceramics. Newer alumina ceramics are being produced with lower porosity and grain size and with higher density and purity, resulting in lower fracture risk but still greater than that of cobalt-chromium. Heisel C, Silva M, Schmalzried TP: Bearing surface options for total hip replacement in young patients. Instr Course Lect 2004;53:49-65.

Question 2739

Topic: 3. Adult Reconstruction (Hip & Knee)

Figure 7 shows the AP radiograph of a 60-year-old man who has had pain in the thigh for past 6 months. History reveals that he underwent hip replacement 1 year ago. The radiographic changes are most likely the result of what process?

Orthopedic Surgery Board Review 2026 | High-Yield MCQs - Set 4 - Figure 13

. Microtrauma
. Mechanical loosening
. Septic loosening
. Neoplasia
. Congenital anomaly

Correct Answer & Explanation

. Septic loosening


Explanation

The arrows in the radiograph point to circumferential radiolucencies that strongly support the diagnosis of septic loosening. Radiolucent lines that occur in such a short time are also typical of an infection. Garvin KL, Hanssen AD: Infection after total hip arthroplasty: Past, present, and future. J Bone Joint Surg Am 1995;77:1576-1588.

Question 2740

Topic: 3. Adult Reconstruction (Hip & Knee)

Figure 37 reveals a periprosthetic fracture around a cemented femoral stem in an 81-year-old patient with Paget's disease and mild coagulopathy. What is the most appropriate reconstructive management on the femoral side?

Orthopedic Surgery Board Review 2026 | High-Yield MCQs - Set 2 - Figure 69

. Open reduction and internal fixation
. Impaction allografting
. Proximally coated femoral stem
. Allograft prosthetic composite (APC)
. Proximal femoral replacement (PFR)

Correct Answer & Explanation

. Proximal femoral replacement (PFR)


Explanation

This is an example of a Vancouver B3 periprosthetic fracture that consists of a fracture around a loose femoral stem with poor proximal bone support. Therefore, open reduction and internal fixation is not an option. PFR is an excellent choice for elderly inactive patients with poor femoral bone stock. The surgery can be performed in an expeditious manner, which is very important in a patient with mild coagulopathy. Impaction allografting and APC are both options for younger patients who have bone stock that needs to be restored. The results of revision arthroplasty using proximally coated stems, especially under these circumstances, are poor. Duncan CP, Masri BA: Fractures of the femur after hip replacement. Instr Course Lect 1995;44:293-304. Parvizi J, Sim FH: Proximal femoral replacements with megaprostheses. Clin Orthop 2004;420:169-175.