Menu

Question 4581

Topic: 6. Spine

A 65-year-old man presents with progressive gait instability and loss of fine motor skills. Examination demonstrates a positive Hoffman sign, hyperreflexia, and intrinsic hand muscle atrophy. Lateral radiographs reveal a flexible kyphotic deformity of the cervical spine. MRI demonstrates multi-level cervical stenosis (C3-C6) with spinal cord compression anteriorly. What is the most appropriate surgical intervention?

. C3-C6 laminectomy alone
. C3-C6 posterior laminoplasty
. Anterior cervical discectomy and fusion (ACDF) or corpectomy
. Posterior cervical fusion without decompression
. Intensive physical therapy and bracing

Correct Answer & Explanation

. C3-C6 laminectomy alone


Explanation

In patients with cervical myelopathy and a flexible kyphotic deformity, an anterior approach is preferred to restore lordosis and directly decompress the anterior pathology. Posterior decompression alone in a kyphotic spine is contraindicated as it can lead to progressive deformity and fails to allow the spinal cord to drift posteriorly away from the compression.

Question 4582

Topic: 6. Spine

A 68-year-old woman presents with severe neurogenic claudication and low back pain. Radiographs reveal a grade 1 degenerative spondylolisthesis at L4-L5 that increases to grade 2 on dynamic flexion views. MRI confirms severe central canal stenosis at L4-L5. She has failed 6 months of comprehensive nonoperative management. What is the most appropriate surgical treatment?

. L4-L5 laminectomy alone
. L4-L5 laminectomy and instrumented posterolateral fusion
. Placement of an interspinous process spacer
. Repeat lumbar epidural steroid injections
. Anterior lumbar interbody fusion without posterior decompression

Correct Answer & Explanation

. L4-L5 laminectomy alone


Explanation

Decompression with instrumented fusion is indicated for lumbar spinal stenosis associated with dynamic instability or mobile degenerative spondylolisthesis. Performing a decompression alone in the setting of instability carries a high risk of iatrogenic progressive deformity and recurrent symptoms.

Question 4583

Topic: 6. Spine

A 55-year-old man is newly diagnosed with cervical spondylotic myelopathy. Which of the following MRI findings is the most reliable independent predictor of poor neurologic recovery following surgical decompression?

. Focal hyperintensity on T2-weighted images
. Spinal cord hypointensity on T1-weighted images
. Multi-level effacement of the subarachnoid space
. Decreased T2 signal intensity within the adjacent disc spaces
. Hypertrophy of the ligamentum flavum

Correct Answer & Explanation

. Focal hyperintensity on T2-weighted images


Explanation

Spinal cord hypointensity on T1-weighted MRI indicates permanent cord damage, such as myelomalacia, necrosis, or cyst formation, and is a strong predictor of poor postoperative neurologic recovery. Hyperintensity on T2 alone is less specific and may represent reversible edema.

Question 4584

Topic: 6. Spine

A 72-year-old man complains of bilateral leg pain that worsens significantly with walking. Which of the following historical factors or examination findings is most specific for neurogenic claudication secondary to lumbar spinal stenosis, as opposed to vascular claudication?

. Pain relief upon standing still in an upright position
. Absence of bilateral dorsalis pedis pulses
. Pallor of the skin with elevation of the legs
. Pain relief strictly associated with lumbar flexion (e.g., leaning on a shopping cart)
. Pain typically originating in the calves and radiating distally

Correct Answer & Explanation

. Pain relief upon standing still in an upright position


Explanation

Neurogenic claudication is classically relieved by leaning forward or sitting, which flexes the lumbar spine and transiently increases the cross-sectional area of the spinal canal and foramina. Vascular claudication is typically relieved by simply resting or standing still, without a specific postural requirement.

Question 4585

Topic: 6. Spine

Three days following a C3-C6 posterior cervical laminectomy and fusion for cervical myelopathy, a 60-year-old patient develops isolated, profound weakness in unilateral shoulder abduction and elbow flexion. Sensation is completely intact, and his preoperative myelopathic symptoms have improved. What is the most likely etiology?

. Postoperative epidural hematoma
. Intraoperative iatrogenic spinal cord contusion
. C5 nerve root palsy
. Inadequate central decompression at C4-C5
. Acute subarachnoid hemorrhage

Correct Answer & Explanation

. Postoperative epidural hematoma


Explanation

C5 palsy is a known complication following cervical decompression (especially posterior laminectomy or laminoplasty), resulting in isolated deltoid and biceps weakness. It is often attributed to posterior drift of the spinal cord resulting in tethering or traction on the short, horizontally oriented C5 nerve roots.

Question 4586

Topic: Cervical Spine

A 50-year-old man of East Asian descent presents with progressive hand clumsiness and broad-based gait. CT demonstrates a continuous, dense bony mass along the posterior aspect of the C3 to C6 vertebral bodies. MRI shows significant anterior spinal cord compression, but sagittal alignment is lordotic. What is the most appropriate surgical option?

. Laminectomy without instrumented fusion
. Anterior cervical corpectomies and fusion
. Posterior cervical laminoplasty
. Multi-level anterior cervical discectomy and fusion (ACDF)
. Continuous cervical traction and observation

Correct Answer & Explanation

. Laminectomy without instrumented fusion


Explanation

The presentation is classic for multi-level Ossification of the Posterior Longitudinal Ligament (OPLL). For multi-level OPLL with maintained cervical lordosis, a posterior approach (laminoplasty or laminectomy with fusion) is preferred to avoid the high risk of dural tears and massive bleeding associated with anterior resection of the ossified mass.

Question 4587

Topic: 6. Spine

Look at the provided image:

A 67-year-old male presents with progressive lower extremity weakness and neurogenic claudication exacerbated by prolonged standing. Based on standard surgical principles for degenerative lumbar spinal stenosis, if this patient fails non-operative management, what is the primary surgical objective?

. Interlaminar stabilization without decompression
. Direct neural repair via microscopic grafting
. Enlargement of the spinal canal and lateral recesses to relieve neural compression
. Total disc replacement to restore foraminal height
. Kyphoplasty to stabilize adjacent vertebral bodies

Correct Answer & Explanation

. Interlaminar stabilization without decompression


Explanation

The cornerstone of surgical treatment for lumbar spinal stenosis is adequate decompression of the neural elements by enlarging the central canal and lateral recesses. Fusion may be added if dynamic instability or significant deformity is present, but definitive decompression remains the primary objective.

Question 4588

Topic: 6. Spine

In the pathophysiology of central degenerative lumbar spinal stenosis, which anatomical structure primarily contributes to the narrowing of the central spinal canal from a posterior direction?

. Herniated nucleus pulposus
. Hypertrophied ligamentum flavum
. Ossified posterior longitudinal ligament
. Facet joint synovial cysts
. Anterior marginal osteophytes

Correct Answer & Explanation

. Herniated nucleus pulposus


Explanation

Central canal stenosis in the degenerative lumbar spine is a multifactorial process. It is most commonly caused by disc bulging anteriorly, facet arthropathy laterally, and hypertrophy or inward buckling of the ligamentum flavum posteriorly.

Question 4589

Topic: 6. Spine

A 70-year-old woman with suspected cervical spondylotic myelopathy is asked to open and close her hands rapidly. She demonstrates spontaneous abduction of her small finger during the exam. What does the abduction of the small finger represent?

. Hoffman sign
. Wartenberg sign
. Lhermitte sign
. Babinski sign
. Inverted radial reflex

Correct Answer & Explanation

. Hoffman sign


Explanation

The Wartenberg sign is the spontaneous abduction of the small finger due to weakness of the intrinsic hand muscles (ulnar nerve-innervated interossei) unopposed by the extensor digiti minimi. It is a classic exam finding in cervical myelopathy, representing upper motor neuron dysfunction.

Question 4590

Topic: 6. Spine

During an L4-L5 laminectomy for severe lumbar spinal stenosis, a 4-mm incidental dural tear occurs dorsally. Cerebrospinal fluid leakage is noted, and the tear is primarily repaired watertight with 4-0 Nurolon sutures. What is the most appropriate next step in management?

. Convert the procedure to an instrumented fusion to immobilize the segment
. Place a lumbar subarachnoid drain for 7 days to divert fluid
. Apply a fascial graft or fibrin glue over the repair followed by primary wound closure
. Abort the procedure immediately without completing the decompression
. Leave the tear open, close the fascia loosely, and place a superficial drain

Correct Answer & Explanation

. Convert the procedure to an instrumented fusion to immobilize the segment


Explanation

For small incidental dural tears that are primarily repaired watertight, reinforcing the repair with fibrin glue, Gelfoam, or fascial patches, followed by tight closure of the overlying fascia, is the standard of care. Routine use of subarachnoid drains is not indicated for uncomplicated, primarily repaired tears.

Question 4591

Topic: 6. Spine

When comparing minimally invasive (MIS) tubular decompression (e.g., unilateral laminotomy for bilateral decompression) to traditional open laminectomy for lumbar spinal stenosis, the MIS approach has been most consistently shown in the literature to result in which of the following?

. Lower rates of incidental dural tears
. Higher long-term success rates for axial back pain
. Decreased intraoperative blood loss and shorter hospital stays
. Significantly lower rates of re-operation at 5 years
. Vastly improved Oswestry Disability Index scores at 2 years compared to open techniques

Correct Answer & Explanation

. Lower rates of incidental dural tears


Explanation

Studies comparing MIS decompression to open laminectomy for lumbar stenosis consistently demonstrate decreased intraoperative blood loss, less tissue disruption, and shorter hospital stays for the MIS group. Long-term clinical outcomes and re-operation rates, however, remain statistically equivalent between the two techniques.

Question 4592

Topic: 6. Spine

A 32-year-old man presents with severe neck pain and right arm weakness (deltoid and biceps) after a mountain biking accident. He is awake, alert, and cooperative. Plain radiographs demonstrate a right-sided C4-C5 unilateral facet dislocation. What is the most appropriate next step in his management?

. Immediate MRI of the cervical spine
. Closed reduction using cranial tongs and traction
. Anterior cervical discectomy and fusion (ACDF)
. Posterior cervical fusion
. Open reduction without preoperative imaging

Correct Answer & Explanation

. Immediate MRI of the cervical spine


Explanation

In an awake and cooperative patient with a cervical facet dislocation and an acute neurologic deficit, immediate closed reduction via cranial traction is indicated. MRI can be delayed until after reduction is achieved to prevent prolonged spinal cord or nerve root compression.

Question 4593

Topic: 6. Spine

A 68-year-old man with a known history of severe cervical spondylosis falls forward and strikes his chin, sustaining a hyperextension injury. He presents with profound bilateral upper extremity weakness (hands greater than shoulders) and relatively preserved lower extremity strength. What is the most likely neurologic diagnosis?

. Anterior cord syndrome
. Brown-Sequard syndrome
. Central cord syndrome
. Posterior cord syndrome
. Complete spinal cord injury

Correct Answer & Explanation

. Anterior cord syndrome


Explanation

Central cord syndrome typically occurs following a hyperextension injury in older adults with pre-existing cervical spondylosis. It characteristically presents with disproportionately greater motor impairment in the upper extremities compared to the lower extremities.

Question 4594

Topic: Cervical Spine

An 82-year-old woman falls from a standing height. CT imaging reveals a Type II odontoid fracture with 2 mm of posterior displacement. She has no neurologic deficits but has a history of congestive heart failure and severe COPD. What is the most appropriate management?

. Halo vest immobilization
. Rigid cervical collar
. Anterior odontoid screw fixation
. Posterior C1-C2 fusion
. Occipitocervical fusion

Correct Answer & Explanation

. Halo vest immobilization


Explanation

In elderly patients with Type II odontoid fractures and significant medical comorbidities, rigid cervical collar immobilization is generally preferred. Both surgical intervention and halo vest immobilization carry disproportionately high morbidity and mortality in this specific patient population.

Question 4595

Topic: 6. Spine

A 60-year-old man undergoes an uncomplicated C3-C6 laminectomy and posterior fusion for cervical myelopathy. Two days postoperatively, he develops new-onset isolated right deltoid weakness (2/5) with intact sensation and no lower extremity changes. An urgent MRI shows a well-decompressed cord with no hematoma. What is the most likely diagnosis?

. Epidural hematoma
. Hardware failure
. Inadequate decompression
. C5 nerve root palsy
. Spinal cord infarction

Correct Answer & Explanation

. Epidural hematoma


Explanation

C5 nerve root palsy is a known complication following posterior cervical decompression, believed to be caused by posterior spinal cord shift and subsequent tethering of the short C5 nerve root. Treatment is typically supportive, and many patients recover spontaneously over time.

Question 4596

Topic: 6. Spine

A 65-year-old woman presents with severe neurogenic claudication and an L4-L5 degenerative spondylolisthesis. She has failed 6 months of conservative management. According to classical literature, which surgical treatment provides the lowest rate of reoperation and best clinical outcome?

. Laminectomy alone
. Laminectomy with posterolateral instrumented fusion
. Interspinous process spacer
. Total disc replacement
. Microdiscectomy

Correct Answer & Explanation

. Laminectomy alone


Explanation

For degenerative spondylolisthesis accompanied by symptomatic stenosis, a decompressive laminectomy combined with instrumented fusion has traditionally demonstrated better long-term clinical outcomes and lower reoperation rates compared to laminectomy alone.

Question 4597

Topic: Cervical Spine

A 30-year-old man dives into a shallow pool and sustains a burst fracture of C1 (Jefferson fracture). On the open-mouth odontoid radiograph, the sum of the lateral mass displacement of C1 over C2 is measured at 8 mm. What does this finding indicate?

. Intact transverse ligament
. Ruptured transverse ligament
. Ruptured alar ligament
. Associated Type II odontoid fracture
. Atlanto-occipital dissociation

Correct Answer & Explanation

. Intact transverse ligament


Explanation

According to Spence's rule, a combined lateral mass overhang of C1 on C2 greater than 6.9 mm on an AP radiograph strongly suggests a rupture of the transverse ligament. This renders the C1 ring fracture highly unstable.

Question 4598

Topic: 6. Spine

A 72-year-old man presents with deteriorating handwriting, difficulty buttoning his shirt, and a wide-based, unsteady gait. Which of the following physical exam findings is most specific for the likely diagnosis?

. Decreased ankle reflexes
. Positive straight leg raise
. Hoffmann sign
. Waddell signs
. Spurling's test

Correct Answer & Explanation

. Decreased ankle reflexes


Explanation

The patient's symptoms are classic for cervical spondylotic myelopathy. A positive Hoffmann sign indicates an upper motor neuron lesion and is highly suggestive of cervical myelopathy in this clinical context.

Question 4599

Topic: Thoracolumbar Spine & Deformity

A 25-year-old man is involved in a motor vehicle accident. CT reveals a traumatic spondylolisthesis of the axis with C2 pars fractures. There is 4 mm of translation and 15 degrees of angulation of C2 on C3. What is the Levine-Edwards classification and the most appropriate initial management?

. Type I; rigid collar
. Type II; reduction with traction and halo vest
. Type IIA; reduction with traction and halo vest
. Type IIA; halo vest in slight extension (traction contraindicated)
. Type III; rigid collar

Correct Answer & Explanation

. Type I; rigid collar


Explanation

A Levine-Edwards Type IIA fracture is characterized by severe angulation with minimal translation. Cervical traction is strictly contraindicated as it can cause over-distraction and neurologic injury; management involves gentle extension and halo vest immobilization.

Question 4600

Topic: 6. Spine

A 55-year-old man complains of severe right leg radicular pain. MRI demonstrates a synovial cyst arising from the L4-L5 facet joint, compressing the right L5 nerve root. If non-operative management fails, what is the most appropriate definitive surgical intervention?

. Cyst aspiration alone
. Laminectomy and cyst excision without fusion
. Laminectomy, cyst excision, and fusion
. Endoscopic rhizotomy
. Total disc replacement

Correct Answer & Explanation

. Cyst aspiration alone


Explanation

Lumbar facet cysts strongly indicate underlying segmental instability. Therefore, decompression and cyst excision combined with a spinal fusion provide the most definitive treatment and carry the lowest risk of cyst recurrence.