Question 5201

Topic: 3. Adult Reconstruction (Hip & Knee)

A 55-year-old active male underwent a total hip arthroplasty using a ceramic-on-ceramic bearing. Two years postoperatively, he complains of a reproducible squeaking noise during walking, though he is pain-free. Which of the following factors is most strongly associated with this phenomenon?

. Use of a highly cross-linked polyethylene liner
. Femoral stem undersizing
. Malpositioning of the acetabular component
. Aseptic loosening of the femoral stem
. History of developmental dysplasia of the hip

Correct Answer & Explanation

. Use of a highly cross-linked polyethylene liner


Explanation

Squeaking in ceramic-on-ceramic THA is heavily associated with component malposition, specifically edge loading from severe acetabular component anteversion or inclination. Other factors include microseparation and specific implant designs.

Question 5202

Topic: 3. Adult Reconstruction (Hip & Knee)

During a primary total knee arthroplasty, the surgeon assesses the gaps after the initial bone cuts. The knee is perfectly balanced in full extension, but the joint is significantly tight in 90 degrees of flexion. Which of the following adjustments is the most appropriate next step?

. Recut the proximal tibia with more distal resection
. Downsize the femoral component
. Release the posterior capsule
. Upsize the femoral component
. Use a thicker polyethylene insert

Correct Answer & Explanation

. Recut the proximal tibia with more distal resection


Explanation

A tight flexion gap with a balanced extension gap is best addressed by downsizing the femoral component, which translates the posterior condyles anteriorly, or by increasing the posterior tibial slope. Re-cutting the proximal tibia would inappropriately widen the already balanced extension gap.

Question 5203

Topic: 3. Adult Reconstruction (Hip & Knee)

A 58-year-old male presents with new-onset groin pain 5 years after an uncomplicated primary metal-on-polyethylene total hip arthroplasty. He has a 36-mm cobalt-chromium head on a titanium stem. Serum cobalt levels are significantly elevated out of proportion to chromium levels. Inflammatory markers are normal. What is the most likely diagnosis?

. Polyethylene wear-induced osteolysis
. Mechanically assisted crevice corrosion (Trunnionosis)
. Low-grade periprosthetic joint infection
. Aseptic loosening of the femoral stem
. Iliopsoas impingement

Correct Answer & Explanation

. Polyethylene wear-induced osteolysis


Explanation

High serum cobalt levels out of proportion to chromium in a metal-on-polyethylene THA indicates mechanically assisted crevice corrosion (trunnionosis) at the head-neck junction. This complication is associated with larger head sizes and higher offset necks.

Question 5204

Topic: 3. Adult Reconstruction (Hip & Knee)
A 65-year-old female undergoes a primary total knee arthroplasty. Four weeks postoperatively, she presents with a swollen, erythematous, and painful knee. A diagnostic arthrocentesis is performed. According to the modified Musculoskeletal Infection Society (MSIS) criteria, what is the synovial fluid white blood cell (WBC) count threshold suggestive of an acute periprosthetic joint infection (<90 days post-op)?
. 1,500 cells/uL
. 3,000 cells/uL
. 10,000 cells/uL
. 27,800 cells/uL
. 50,000 cells/uL

Correct Answer & Explanation

. 10,000 cells/uL


Explanation

For acute periprosthetic joint infections (less than 90 days postoperative), the diagnostic threshold for synovial fluid WBC is typically >10,000 cells/uL. For chronic infections (>90 days), the threshold is lower, typically >3,000 cells/uL.

Question 5205

Topic: Total Hip Arthroplasty (THA)

A 45-year-old active male underwent a total hip arthroplasty utilizing a ceramic-on-ceramic bearing. Two years later, he complains of an audible squeaking sound from the hip during deep flexion activities. Which of the following factors is most strongly associated with the development of this phenomenon?

. Use of a 28mm ceramic head
. Improper cup positioning leading to edge loading
. Use of a cemented femoral stem
. High level of activity in a low BMI patient
. Use of highly cross-linked polyethylene

Correct Answer & Explanation

. Use of a 28mm ceramic head


Explanation

Squeaking in ceramic-on-ceramic THA is highly associated with edge loading, which often results from component malposition (such as excessive cup anteversion or inclination), microseparation, and higher patient BMI.

Question 5206

Topic: 3. Adult Reconstruction (Hip & Knee)

A 62-year-old female reports a painful catching and popping sensation in her knee when rising from a chair, 14 months after receiving a posterior-stabilized total knee arthroplasty. The catch is consistently felt as the knee extends from 45 to 30 degrees of flexion. What is the most likely etiology of this patient's symptoms?

. Undersized femoral component
. Patella baja
. Fibrous nodule catching in the intercondylar box
. Loose tibial baseplate
. Posterior cruciate ligament tightness

Correct Answer & Explanation

. Undersized femoral component


Explanation

Patellar clunk syndrome occurs primarily in posterior-stabilized TKAs when a fibrous nodule forms at the superior pole of the patella and mechanically catches in the intercondylar box of the femoral component during active extension.

Question 5207

Topic: 3. Adult Reconstruction (Hip & Knee)

A 71-year-old male presents with an inability to perform a straight leg raise two years following a primary total knee arthroplasty. Imaging confirms a chronic, retracted rupture of the patellar tendon. The components are well-fixed and appropriately positioned. What is the most reliable surgical salvage option?

. Primary repair with non-absorbable sutures
. Autologous hamstring graft reconstruction
. Extensor mechanism allograft or synthetic mesh reconstruction
. Revision to a hinged knee replacement
. Medial gastrocnemius rotational flap

Correct Answer & Explanation

. Primary repair with non-absorbable sutures


Explanation

Chronic extensor mechanism ruptures post-TKA have extremely high failure rates with primary repair. Reconstruction utilizing a whole extensor mechanism allograft or synthetic Marlex mesh provides the most reliable salvage outcome.

Question 5208

Topic: 3. Adult Reconstruction (Hip & Knee)

A surgeon is performing a primary total hip arthroplasty via the direct anterior approach utilizing the Hueter interval. Postoperatively, the patient complains of dysesthesia and numbness over the anterolateral aspect of the operative thigh. Which nerve was most likely injured during the superficial dissection?

. Femoral nerve
. Sciatic nerve
. Lateral femoral cutaneous nerve
. Obturator nerve
. Superior gluteal nerve

Correct Answer & Explanation

. Femoral nerve


Explanation

The lateral femoral cutaneous nerve (LFCN) crosses the surgical field during the direct anterior approach to the hip and is at high risk of stretch or transection when developing the interval between the tensor fasciae latae and the sartorius.

Question 5209

Topic: Total Knee Arthroplasty (TKA)

A 75-year-old female with severe rheumatoid arthritis presents for a primary total knee arthroplasty. Clinical and radiographic evaluation demonstrates a severe valgus deformity with a completely incompetent, non-functional medial collateral ligament (MCL). Which of the following implant constraints is indicated?

. Cruciate retaining prosthesis
. Posterior stabilized prosthesis
. Constrained condylar knee (CCK)
. Rotating hinge prosthesis
. Unicompartmental knee replacement

Correct Answer & Explanation

. Cruciate retaining prosthesis


Explanation

In the setting of a completely incompetent or absent MCL, a standard varus-valgus constrained (CCK) implant will not provide sufficient stability and may fail. A rotating hinge prosthesis is strictly indicated to provide coronal stability.

Question 5210

Topic: 3. Adult Reconstruction (Hip & Knee)

A 68-year-old male with a prior long lumbar fusion (L2-pelvis) for degenerative scoliosis is planned for a primary total hip arthroplasty. He exhibits a stiff spinopelvic junction with a fixed pelvic retroversion (flatback deformity). How should the acetabular component positioning be modified to minimize the risk of anterior dislocation in extension?

. Target the standard Lewinnek safe zone
. Increase cup anteversion
. Decrease cup anteversion
. Increase cup inclination
. Decrease cup inclination

Correct Answer & Explanation

. Target the standard Lewinnek safe zone


Explanation

A fused spine with a flatback deformity results in fixed pelvic retroversion, causing relative functional over-anteversion of the acetabulum when standing. To prevent anterior impingement and dislocation in extension, the cup should be placed in less anteversion.

Question 5211

Topic: Total Knee Arthroplasty (TKA)

Which of the following best describes the primary goal of the kinematic alignment philosophy in total knee arthroplasty compared to traditional mechanical alignment?

. Setting the joint line exactly perpendicular to the mechanical axis of the limb
. Restoration of the pre-arthritic constitutional alignment and joint line obliquity
. Intentional placement of the tibial component in 3 degrees of varus for all patients
. Routine release of the medial collateral ligament to achieve a rectangular gap
. Cutting the tibia strictly perpendicular to its anatomic axis

Correct Answer & Explanation

. Setting the joint line exactly perpendicular to the mechanical axis of the limb


Explanation

Kinematic alignment aims to restore the patient's pre-arthritic constitutional alignment and natural joint line obliquity by resurfacing the articular wear, minimizing the need for routine ligamentous releases.

Question 5212

Topic: 3. Adult Reconstruction (Hip & Knee)
Highly cross-linked ultra-high-molecular-weight polyethylene (UHMWPE) is widely used in total hip arthroplasty to reduce wear-induced osteolysis. By what primary mechanism does cross-linking improve the tribological properties of the polyethylene?
. Increasing oxidative degradation capacity
. Increasing the number of free radicals
. Decreasing linear wear rate by reducing molecular chain mobility
. Increasing the elastic modulus significantly
. Promoting the formation of larger wear debris particles

Correct Answer & Explanation

. Decreasing linear wear rate by reducing molecular chain mobility


Explanation

Irradiation causes cross-linking of UHMWPE, which significantly decreases the linear wear rate by restricting molecular chain mobility. Subsequent melting or annealing is required to eliminate residual free radicals and prevent long-term oxidation.

Question 5213

Topic: 3. Adult Reconstruction (Hip & Knee)

An 82-year-old female sustains a fall 8 years after an uncemented total hip arthroplasty. Radiographs reveal a periprosthetic femur fracture extending around the stem. The stem appears subsided and clinically loose, but there is adequate cortical bone stock in the proximal femur. According to the Vancouver classification, what is the most appropriate management?

. Open reduction internal fixation with a locking plate
. Impaction grafting with a cemented stem
. Revision to a long fluted tapered uncemented stem
. Cortical strut allograft only
. Proximal femoral replacement

Correct Answer & Explanation

. Open reduction internal fixation with a locking plate


Explanation

This is a Vancouver B2 fracture (fracture around the stem, loose implant, adequate bone stock). The gold standard treatment is revision to a long bypass stem (fluted, tapered, uncemented) that spans the fracture by at least two cortical diameters.

Question 5214

Topic: 3. Adult Reconstruction (Hip & Knee)

A 59-year-old male is evaluated 8 weeks after a primary total knee arthroplasty. He complains of a stiff knee despite compliant physical therapy. His range of motion is 10 to 75 degrees. Inflammatory markers are normal, and radiographs show well-positioned components. What is the most appropriate next step in management?

. Revision total knee arthroplasty
. Manipulation under anesthesia
. Continuation of physical therapy alone for 6 more months
. Arthroscopic lysis of adhesions
. V-Y quadricepsplasty

Correct Answer & Explanation

. Revision total knee arthroplasty


Explanation

Manipulation under anesthesia (MUA) is highly successful when performed between 6 and 12 weeks postoperatively for persistent knee stiffness (less than 90 degrees of flexion) after ensuring there is no underlying infection or component malposition.

Question 5215

Topic: Total Knee Arthroplasty (TKA)

A 65-year-old male complains of knee instability when descending stairs 1 year after a primary TKA. Examination reveals a knee that is fully stable in full extension but exhibits marked anteroposterior laxity in 90 degrees of flexion. Which of the following intraoperative technical errors most likely occurred?

. Excessive distal femoral resection
. Undersizing the femoral component in the anteroposterior (AP) dimension
. Oversizing the femoral component in the AP dimension
. Inadequate proximal tibial resection
. Placing the femoral component in excessive external rotation

Correct Answer & Explanation

. Excessive distal femoral resection


Explanation

Undersizing the femoral component in the AP dimension reduces the posterior condylar offset, thereby increasing the flexion gap while leaving the extension gap unaffected. This creates isolated flexion instability often noticed during activities like stair descent.

Question 5216

Topic: 3. Adult Reconstruction (Hip & Knee)

A 55-year-old male presents with deep groin pain 6 years after a THA utilizing a metal-on-polyethylene bearing with a large diameter metal head. Labs demonstrate elevated serum cobalt and chromium levels. Aspiration yields cloudy fluid with a normal neutrophil count. What is the most likely cause of his symptoms?

. Adverse local tissue reaction secondary to trunnionosis
. Polyethylene wear-induced osteolysis
. Low-grade periprosthetic joint infection
. Aseptic loosening of the acetabular component
. Iliopsoas impingement

Correct Answer & Explanation

. Adverse local tissue reaction secondary to trunnionosis


Explanation

Elevated metal ions in a metal-on-polyethylene THA with a large metal head indicate mechanically assisted crevice corrosion (trunnionosis) at the head-neck junction. This can lead to an adverse local tissue reaction (ALTR) mimicking infection or loosening.

Question 5217

Topic: Total Knee Arthroplasty (TKA)

During a primary TKA for a fixed valgus deformity of 20 degrees, the surgeon notes a tight lateral compartment in full extension, but the knee is well-balanced in 90 degrees of flexion. Which of the following structures is the primary tether and should be released to balance the gaps?

. Lateral collateral ligament
. Popliteus tendon
. Iliotibial band
. Posterior cruciate ligament
. Lateral head of the gastrocnemius

Correct Answer & Explanation

. Lateral collateral ligament


Explanation

The iliotibial band (ITB) acts as a primary lateral tether in extension but has minimal effect in flexion. Releasing or pie-crusting the ITB will correct the tight extension gap without disrupting the already balanced flexion gap.

Question 5218

Topic: Total Hip Arthroplasty (THA)

During templating for a primary THA, the surgeon notes a 15 mm leg length discrepancy (affected leg is short) and a highly offset native femur. If a standard offset stem is used instead of a high offset stem, what compensatory intraoperative error is most likely to occur in an attempt to restore soft tissue tension?

. Excessive medialization of the cup
. Increasing acetabular anteversion
. Shortening the leg excessively
. Excessive lengthening of the operative leg
. Decreasing acetabular inclination

Correct Answer & Explanation

. Excessive medialization of the cup


Explanation

If a standard offset stem is used in a patient requiring high offset, the abductors will remain lax. To achieve stability and proper soft tissue tension, the surgeon may inadvertently increase the neck length, resulting in an excessively lengthened leg.

Question 5219

Topic: 3. Adult Reconstruction (Hip & Knee)

Which of the following statements best describes the primary biomechanical goal of kinematic alignment in total knee arthroplasty?

. Restoring the overall limb mechanical axis to neutral (0 degrees)
. Aligning both components strictly perpendicular to the mechanical axes of the femur and tibia
. Recreating the patient's pre-arthritic native joint lines and kinematic axes
. Placing the femoral component in 3 degrees of external rotation relative to the posterior condylar axis universally
. Over-resecting the distal medial femur to definitively correct varus deformity

Correct Answer & Explanation

. Restoring the overall limb mechanical axis to neutral (0 degrees)


Explanation

Kinematic alignment aims to restore the patient's pre-arthritic constitutional alignment and joint line obliquity. It achieves this by co-aligning the components with the three specific kinematic axes of the native knee.

Question 5220

Topic: Total Hip Arthroplasty (THA)

A 42-year-old highly active female undergoes THA using a ceramic-on-ceramic bearing. One year later, she complains of a loud, audible squeaking noise from the hip during specific movements, though she is pain-free. What is the most significant risk factor for this phenomenon?

. Use of a large diameter femoral head
. Femoral stem retroversion
. Acetabular component malposition leading to edge loading
. Body mass index less than 25
. High levels of daily physical activity

Correct Answer & Explanation

. Use of a large diameter femoral head


Explanation

Squeaking in ceramic-on-ceramic THA is most strongly associated with component malposition (such as a steeply inclined or excessively anteverted/retroverted cup). This malposition leads to edge loading, loss of fluid-film lubrication, and stripe wear.