Question 3521

Topic: 3. Adult Reconstruction (Hip & Knee)

A 60-year-old man presents with a painful right hip 6 years following a primary total hip arthroplasty. He has a metal-on-polyethylene bearing with a titanium femoral stem and a large diameter (36 mm) cobalt-chromium femoral head. Serum cobalt levels are markedly elevated at 12 ppb, while chromium levels are normal. An MRI reveals a solid tissue mass adjacent to the hip joint. What is the most likely etiology of this patient's presentation?

. Accelerated wear of the highly cross-linked polyethylene liner
. Mechanically assisted crevice corrosion at the head-neck junction
. Periprosthetic joint infection with an atypical mycobacterium
. Aseptic loosening of the femoral stem
. Galvanic corrosion at the stem-cement interface

Correct Answer & Explanation

. Mechanically assisted crevice corrosion at the head-neck junction


Explanation

The patient's clinical presentation, featuring an Adverse Local Tissue Reaction (ALTR) mass, elevated cobalt, and normal chromium in the setting of a metal-on-polyethylene THA, is characteristic of trunnionosis. This condition is driven by mechanically assisted crevice corrosion (MACC) and fretting at the modular head-neck junction (the trunnion) between a titanium stem and a cobalt-chromium head. The use of large-diameter cobalt-chromium heads increases torque at the trunnion, exacerbating this specific mode of failure.

Question 3522

Topic: 3. Adult Reconstruction (Hip & Knee)

A 64-year-old man presents with progressive left groin pain 6 years after a primary total hip arthroplasty. The implant utilizes a titanium cementless stem, a cobalt-chromium femoral head, and a highly cross-linked polyethylene liner in a titanium shell. Radiographs show no evidence of component loosening. Laboratory workup reveals an erythrocyte sedimentation rate of 12 mm/hr, a C-reactive protein of 0.4 mg/L, a serum cobalt level of 16.5 mcg/L, and a serum chromium level of 1.2 mcg/L. An MRI demonstrates a solid-cystic pseudotumor in the joint space. What is the most likely etiology of his presentation?

. Polyethylene wear debris causing osteolysis
. Mechanically assisted crevice corrosion at the head-neck junction
. Periprosthetic joint infection with a low-virulence organism
. Bearing surface wear of a metal-on-metal articulation
. Impingement of the iliopsoas tendon

Correct Answer & Explanation

. Mechanically assisted crevice corrosion at the head-neck junction


Explanation

The patient has elevated metal ions with a classic disproportionate elevation of serum cobalt compared to chromium (Co >> Cr). In the presence of a metal head on a polyethylene liner, this indicates mechanically assisted crevice corrosion (MACC), also known as trunnionosis, at the modular head-neck junction. A metal-on-metal bearing surface wear pattern typically presents with more equally elevated cobalt and chromium levels.

Question 3523

Topic: Total Knee Arthroplasty (TKA)

A 68-year-old woman complains of recurrent knee swelling and a sensation of her knee 'giving way' particularly when descending stairs, 1 year after a primary posterior-stabilized total knee arthroplasty. On examination, the knee is completely stable to varus and valgus stress in full extension. At 90 degrees of flexion, there is 12 mm of joint opening with both varus and valgus stress, and a positive anterior drawer test. Which intraoperative technical error most likely caused this specific complication?

. Excessive distal femoral resection
. Undersizing the femoral component
. Undersizing the tibial component
. Excessive proximal tibial resection
. Using a tibial insert that is too thick

Correct Answer & Explanation

. Undersizing the femoral component


Explanation

The patient presents with isolated flexion instability, characterized by a stable extension gap and a loose flexion gap. During a measured resection TKA, undersizing the femoral component in the anteroposterior (AP) dimension increases the flexion gap without altering the extension gap, leading to flexion instability. Excessive distal femoral resection would affect only the extension gap, causing extension instability. Excessive proximal tibial resection affects both gaps equally.

Question 3524

Topic: 3. Adult Reconstruction (Hip & Knee)

A 79-year-old woman sustains a fall and presents with severe thigh pain. She underwent a total hip arthroplasty 12 years ago with a polished taper-slip cemented stem. Radiographs demonstrate a periprosthetic spiral fracture of the femur located around the tip of the stem. The stem has subsided 3 cm compared to prior films, and the cement mantle is extensively fractured. Proximal bone stock is adequate.

According to the Vancouver classification, what is the most appropriate management?

. Open reduction and internal fixation with a locking plate
. Open reduction and internal fixation with cables and cortical strut allografts
. Revision total hip arthroplasty utilizing a long cementless diaphyseal-engaging stem
. Revision total hip arthroplasty with a standard-length cemented stem
. Nonoperative management with a hinged fracture brace

Correct Answer & Explanation

. Revision total hip arthroplasty utilizing a long cementless diaphyseal-engaging stem


Explanation

This is a Vancouver B2 periprosthetic fracture. The fracture is located around or just below the tip of the stem (Type B), the implant is radiographically loose as evidenced by subsidence and a fractured cement mantle (B2), and the proximal bone stock is adequate. The standard of care for a Vancouver B2 fracture is revision of the femoral component to a long cementless diaphyseal-engaging stem that bypasses the fracture by at least two cortical diameters, often accompanied by cerclage wiring of the fracture.

Question 3525

Topic: 3. Adult Reconstruction (Hip & Knee)

A 62-year-old woman is 18 months post-op from a posterior-stabilized total knee arthroplasty. She reports a painful popping and catching sensation at the anterior aspect of her knee when extending her leg from a flexed seated position. Physical exam reveals a palpable, painful clunk at approximately 35-40 degrees of flexion as the knee extends. Which of the following implant design features or surgical factors is most strongly associated with this complication?

. Increased femoral component flexion
. Patella baja
. A high intercondylar box design of the femoral component
. Over-resection of the patellar bone
. Internal rotation of the tibial component

Correct Answer & Explanation

. A high intercondylar box design of the femoral component


Explanation

The patient is presenting with patellar clunk syndrome, a well-recognized complication associated primarily with posterior-stabilized (PS) knee designs. It occurs when a fibrous nodule forms on the deep surface of the quadriceps tendon just proximal to the superior pole of the patella. During extension, this nodule gets caught in the intercondylar notch (box) of the femoral component and then 'clunks' out as the knee extends further. It is most strongly associated with older PS implant designs that feature a sharp, high, or anteriorly positioned intercondylar box.

Question 3526

Topic: 3. Adult Reconstruction (Hip & Knee)

A 55-year-old active man complains of a high-pitched squeaking sound originating from his right hip when walking or bending, 3 years after receiving a primary total hip arthroplasty with a ceramic-on-ceramic bearing. He reports no pain, and radiographs show well-fixed components. Which of the following factors is most strongly associated with the development of this phenomenon?

. Occult periprosthetic joint infection
. Micro-fracture of the ceramic femoral head
. Trunnionosis at the head-neck junction
. Edge loading secondary to cup malposition
. Femoral stem subsidence

Correct Answer & Explanation

. Edge loading secondary to cup malposition


Explanation

Squeaking in ceramic-on-ceramic total hip arthroplasty is a well-documented phenomenon. While its exact etiology can be multifactorial, it is most strongly associated with edge loading caused by component malposition (such as excessive cup inclination or version). Edge loading leads to stripe wear, loss of fluid film lubrication, and altered tribology, resulting in the high-pitched audible squeak.

Question 3527

Topic: 3. Adult Reconstruction (Hip & Knee)

A 72-year-old woman sustains a complete spontaneous rupture of her patellar tendon 4 years after a primary total knee arthroplasty. The implant components are clinically and radiographically well-fixed, and infection has been ruled out. She is scheduled for an extensor mechanism reconstruction using a whole extensor mechanism allograft. To optimize the functional outcome and minimize postoperative extensor lag, how should the allograft be tensioned during the reconstruction?

. In 90 degrees of flexion with 1 cm of laxity
. In 30 degrees of flexion under maximum tension
. In full extension under maximum tension
. In full extension with 2 cm of laxity
. In 45 degrees of flexion under minimal tension

Correct Answer & Explanation

. In full extension under maximum tension


Explanation

Extensor mechanism allografts used for post-TKA ruptures are notorious for stretching out over time, leading to significant and debilitating extensor lags. To counteract this expected biological creep, the current standard of care dictates that the allograft must be tensioned 'bar-string' tight with the knee in full, absolute extension (0 degrees) during the surgical reconstruction.

Question 3528

Topic: 3. Adult Reconstruction (Hip & Knee)
A 38-year-old man presents with an 8-month history of debilitating right groin pain. He has a history of severe asthma managed with frequent bursts of oral corticosteroids. Anteroposterior and lateral radiographs of the hip demonstrate a dense sclerotic rim and a subchondral radiolucent line (crescent sign) in the anterosuperior aspect of the femoral head. The joint space is well-preserved, and there is no flattening of the articular surface. According to the Ficat and Arlet classification, what is the most appropriate definitive management?
. Core decompression with bone marrow aspirate concentrate
. Non-weight bearing for 6 weeks and bisphosphonate therapy
. Vascularized free fibular grafting
. Total hip arthroplasty
. Proximal femoral derotational osteotomy

Correct Answer & Explanation

. Total hip arthroplasty


Explanation

The presence of a 'crescent sign' on radiographs indicates a subchondral fracture, which classifies the osteonecrosis as Ficat Stage III. While joint-preserving procedures like core decompression are often indicated for pre-collapse stages (Ficat I and II), they have unacceptably high failure rates once subchondral fracture (Stage III) or articular collapse/arthritis (Stage IV) has occurred. Therefore, Total Hip Arthroplasty (THA) is the most reliable and definitive treatment for Ficat Stage III osteonecrosis.

Question 3529

Topic: 3. Adult Reconstruction (Hip & Knee)
A 66-year-old male with a painful total knee arthroplasty 3 years post-operatively undergoes a joint aspiration. The synovial fluid analysis reveals a white blood cell count of 4,200 cells/μL with 88% polymorphonuclear leukocytes. Gram stain is negative. Which of the following synovial fluid biomarkers is known to be an antimicrobial peptide released by neutrophils and is highly specific for diagnosing a periprosthetic joint infection (PJI)?
. Interleukin-6 (IL-6)
. Alpha-defensin
. C-reactive protein (CRP)
. Procalcitonin
. Leukocyte esterase

Correct Answer & Explanation

. Alpha-defensin


Explanation

Alpha-defensin is an antimicrobial peptide released by neutrophils in response to pathogens. It has been incorporated into the newer MSIS/ICM criteria for diagnosing periprosthetic joint infection (PJI) due to its high sensitivity and specificity. Unlike systemic markers, synovial alpha-defensin is a direct local marker of infection in the joint fluid and remains highly accurate even in the presence of systemic inflammatory conditions.

Question 3530

Topic: 3. Adult Reconstruction (Hip & Knee)

A surgeon is performing a crucial step in a posterior-stabilized total knee arthroplasty (TKA). After making the initial bone cuts, the trial components are placed. The knee is symmetric and balanced in extension but is too tight in flexion, preventing full range of motion. What is the most appropriate next step?

. Decrease the anteroposterior size of the femoral component
. Resect more proximal tibia
. Resect more distal femur
. Release the posterior capsule
. Use a thinner polyethylene insert

Correct Answer & Explanation

. Decrease the anteroposterior size of the femoral component


Explanation

A tight flexion gap with a balanced extension gap requires reducing the posterior condylar offset to loosen the flexion space. Downsizing the femoral component achieves this without altering the extension gap. Resecting more proximal tibia or using a thinner insert would loosen both gaps symmetrically. Resecting more distal femur would loosen the extension gap only.

Question 3531

Topic: 3. Adult Reconstruction (Hip & Knee)

A 65-year-old woman undergoes primary total hip arthroplasty via a posterior approach. Six weeks postoperatively, she sustains a posterior dislocation while picking an item off the floor. CT scan demonstrates the acetabular component is placed in 45 degrees of inclination and 5 degrees of retroversion. The femoral stem is anteverted 15 degrees. What is the most appropriate definitive management if recurrent instability occurs?

. Revision of the acetabular component to increase anteversion
. Revision of the femoral component to increase anteversion
. Conversion to a bipolar hemiarthroplasty
. Trochanteric advancement
. Placement of a constrained acetabular liner without changing cup position

Correct Answer & Explanation

. Revision of the acetabular component to increase anteversion


Explanation

The target for safe acetabular anteversion is typically 15-20 degrees. Retroversion of the cup strongly predisposes to posterior dislocation. The definitive management for recurrent instability in the setting of an excessively retroverted cup is revision of the acetabular component to correct the version.

Question 3532

Topic: 3. Adult Reconstruction (Hip & Knee)

A 72-year-old man presents with acute onset of severe right knee pain and swelling. He underwent primary TKA 5 years ago and had excellent function. Three days ago, he developed fever and chills following a routine dental cleaning without prophylactic antibiotics. Synovial fluid aspiration yields a WBC count of 85,000 cells/µL with 95% neutrophils. What is the most appropriate surgical management?

. One-stage revision arthroplasty
. Two-stage revision arthroplasty with a static antibiotic spacer
. Debridement, antibiotics, and implant retention (DAIR) with modular polyethylene exchange
. Arthroscopic lavage and placement of an intra-articular drain
. Suppressive long-term oral antibiotics without surgical intervention

Correct Answer & Explanation

. Debridement, antibiotics, and implant retention (DAIR) with modular polyethylene exchange


Explanation

Acute hematogenous periprosthetic joint infections (characterized by symptom onset < 3 weeks in a previously well-functioning joint) are best managed with Debridement, Antibiotics, and Implant Retention (DAIR), which must include exchange of the modular polyethylene insert to access and debride the posterior joint spaces adequately.

Question 3533

Topic: 3. Adult Reconstruction (Hip & Knee)

An 80-year-old man sustains a periprosthetic femur fracture around a cemented polished taper-slip femoral stem placed 10 years ago. Radiographs demonstrate a spiral fracture at the tip of the stem. The stem is radiographically loose with a fractured cement mantle, but there is adequate cortical bone stock both proximally and distally. According to the Vancouver classification, what is the most appropriate treatment?

. Open reduction and internal fixation with cerclage cables only
. Open reduction and internal fixation with a lateral locking plate
. Revision to a cementless, long diaphyseal-engaging stem
. Revision to a standard length cemented stem
. Proximal femoral replacement

Correct Answer & Explanation

. Revision to a cementless, long diaphyseal-engaging stem


Explanation

This describes a Vancouver B2 fracture (fracture around the stem, loose stem, good bone stock). The standard of care is revision arthroplasty to bypass the fracture and achieve stability distal to the fracture site, typically utilizing a cementless long diaphyseal-engaging stem (e.g., fluted, tapered design).

Question 3534

Topic: 3. Adult Reconstruction (Hip & Knee)

A 58-year-old man presents with progressive groin pain 4 years after a primary metal-on-polyethylene THA utilizing a 36mm femoral head and a titanium alloy stem. Inflammatory markers are normal, and joint aspiration is negative for infection. Serum cobalt levels are markedly elevated (15 ppb) while chromium levels are mildly elevated (3 ppb). MARS MRI demonstrates a cystic mass communicating with the joint space. What is the most likely source of the elevated metal ions?

. Bearing surface wear
. Mechanically assisted crevice corrosion at the head-neck junction
. Impingement of the femoral neck on the acetabular rim
. Galvanic corrosion at the stem-cement interface
. Accelerated polyethylene wear debris

Correct Answer & Explanation

. Mechanically assisted crevice corrosion at the head-neck junction


Explanation

Trunnionosis (mechanically assisted crevice corrosion at the modular head-neck junction) typically presents with an elevated Cobalt-to-Chromium ratio and local adverse tissue reactions (ALTR/pseudotumor). It is associated with large femoral heads on titanium stems, which increases torque and micro-motion at the taper junction, even with metal-on-polyethylene bearing surfaces.

Question 3535

Topic: 3. Adult Reconstruction (Hip & Knee)

A 68-year-old woman complains of a painful 'catching' sensation in her knee when rising from a chair, 1 year after a posterior-stabilized TKA. Physical exam reveals a palpable pop at the superior pole of the patella as the knee actively extends from 40 degrees of flexion to full extension. What is the most likely pathogenesis of this condition?

. Impingement of the patella on the anterior aspect of the tibial post
. Fibrous nodule formation at the superior pole of the patella engaging the intercondylar notch
. Asymmetric polyethylene wear from component malalignment
. Patella baja secondary to patellar tendon contracture
. Subluxation of the extensor mechanism due to lateral retinacular tightness

Correct Answer & Explanation

. Fibrous nodule formation at the superior pole of the patella engaging the intercondylar notch


Explanation

Patellar clunk syndrome is caused by a proliferative fibrous nodule forming at the superior pole of the patella (under the quadriceps tendon). During flexion, the nodule enters the intercondylar box of a posterior-stabilized femoral component; during active extension, it catches and 'clunks' as it pops out. Management is typically arthroscopic excision of the nodule.

Question 3536

Topic: 3. Adult Reconstruction (Hip & Knee)

A 62-year-old woman presents with acute onset of severe medial knee pain that began abruptly while walking. Radiographs show minimal joint space narrowing and no obvious fractures. MRI of the knee demonstrates localized bone marrow edema in the medial femoral condyle with a subchondral crescent sign, but no cortical collapse. What is the most appropriate initial management?

. Core decompression of the medial femoral condyle
. Unicompartmental knee arthroplasty
. Total knee arthroplasty
. Protected weight-bearing and analgesia
. Arthroscopic medial meniscectomy

Correct Answer & Explanation

. Protected weight-bearing and analgesia


Explanation

This presentation is characteristic of spontaneous osteonecrosis of the knee (SONK), increasingly referred to as a subchondral insufficiency fracture of the knee (SIFK). Because there is no subchondral collapse, the initial treatment should be nonoperative with protected weight-bearing and analgesics, as a significant portion will resolve without surgical intervention.

Question 3537

Topic: 3. Adult Reconstruction (Hip & Knee)

During preoperative templating for a total hip arthroplasty, a surgeon notes that the planned femoral component will increase the femoral neck offset by 8 mm compared to the contralateral native hip, without altering the leg length. Which of the following biomechanical effects will this change have?

. Decreases the abductor moment arm
. Increases the resultant joint reactive force across the hip
. Decreases the required abductor muscle force during single-leg stance
. Decreases the bending moment on the femoral stem
. Increases the risk of greater trochanteric impingement against the ilium

Correct Answer & Explanation

. Decreases the required abductor muscle force during single-leg stance


Explanation

Increasing femoral offset extends the greater trochanter further laterally, thereby increasing the moment arm of the abductor muscles. This mechanically advantageous position decreases the required abductor muscle force needed to maintain a level pelvis during single-leg stance. Consequently, this also decreases the overall joint reactive force across the hip, though it increases the bending moment at the stem-neck junction.

Question 3538

Topic: Total Knee Arthroplasty (TKA)

A 70-year-old man complains of persistent anterior knee pain and a feeling of instability 2 years after a primary TKA. CT scan evaluation demonstrates that the femoral component is internally rotated 6 degrees relative to the surgical transepicondylar axis, and the tibial component is internally rotated 9 degrees relative to the medial third of the tibial tubercle. What is the most likely clinical consequence of this combined component positioning?

. Lateral patellar maltracking and anterior knee pain
. Medial patellar subluxation
. Excessive tightness of the flexion gap
. Isolated mid-flexion instability
. Posterior cruciate ligament rupture

Correct Answer & Explanation

. Lateral patellar maltracking and anterior knee pain


Explanation

Internal rotation of the femoral component medially translates the trochlear groove, and internal rotation of the tibial component externalizes the tibial tubercle. Both errors functionally increase the Q-angle effect, pulling the extensor mechanism laterally. This combined malrotation predictably leads to lateral patellar maltracking, subluxation, and severe anterior knee pain.

Question 3539

Topic: Total Hip Arthroplasty (THA)

A 55-year-old active man underwent THA with a ceramic-on-ceramic bearing surface. Three years postoperatively, he complains of an audible squeaking sound from his hip during ambulation, though he denies any pain. Radiographs show well-fixed components with the acetabular cup placed in 65 degrees of inclination and 35 degrees of anteversion. What is the most likely underlying cause of the squeaking?

. Unrecognized fracture of the ceramic liner
. Edge loading due to component malposition
. Trunnionosis at the femoral head-neck junction
. Third-body wear from retained bone cement
. Galvanic corrosion at the modular interfaces

Correct Answer & Explanation

. Edge loading due to component malposition


Explanation

Squeaking in ceramic-on-ceramic THA is heavily associated with component malposition, specifically excessive cup inclination and anteversion. This abnormal biomechanics leads to edge loading, which disrupts fluid film lubrication and causes localized 'stripe wear' on the ceramic head, generating the characteristic squeaking sound.

Question 3540

Topic: 3. Adult Reconstruction (Hip & Knee)

A 68-year-old woman who underwent a posterior-stabilized total knee arthroplasty 18 months ago presents with a painful catching sensation and an audible 'pop' when extending her knee from a flexed position. The range of motion is 0 to 120 degrees. Radiographs show well-fixed components with no evidence of loosening. Which of the following is the most likely cause of her symptoms?

. Impingement of the patella on the anterior tibial post
. A fibrous nodule at the superior pole of the patella catching in the intercondylar box of the femoral component
. Excessive internal rotation of the tibial component
. Global overstuffing of the patellofemoral joint
. Aseptic loosening of the patellar button

Correct Answer & Explanation

. A fibrous nodule at the superior pole of the patella catching in the intercondylar box of the femoral component


Explanation

This presentation is highly characteristic of 'patellar clunk syndrome,' a complication seen most frequently following posterior-stabilized (PS) total knee arthroplasty. It is caused by the formation of a fibrous nodule at the superior pole of the patella. As the knee extends from a flexed position, this nodule catches in the intercondylar box of the PS femoral component and then pops out, producing a painful clunk. Treatment typically involves arthroscopic or open debridement of the nodule.