Question 581

Topic: Total Hip Arthroplasty (THA)

Trunnionosis, or wear at the head-neck junction of a modular femoral stem, has become increasingly recognized in total hip arthroplasty. Which of the following implant combinations represents the highest risk factor for developing symptomatic trunnionosis?

. A 28-mm titanium head on a titanium stem
. A 32-mm ceramic head on a cobalt-chrome stem
. A 28-mm cobalt-chrome head on a cobalt-chrome stem
. A 36-mm cobalt-chrome head on a titanium stem
. A 32-mm ceramic head on a titanium stem

Correct Answer & Explanation

. A 28-mm titanium head on a titanium stem


Explanation

Trunnionosis is caused by mechanically assisted crevice corrosion (fretting and galvanic corrosion). Risk factors include dissimilar metals (galvanic corrosion), larger femoral heads (which increase the toggle and frictional torque at the trunnion), and high-offset necks. A large (e.g., 36-mm) cobalt-chrome head on a titanium stem presents both mixed metals and increased torque, maximizing the risk for trunnionosis.

Question 582

Topic: Total Hip Arthroplasty (THA)

During the posterior approach to the hip, protecting the blood supply to the femoral head in the setting of a femoral neck fracture is not required; however, understanding the vascular anatomy is critical to limit bleeding. The medial femoral circumflex artery (MFCA) provides the primary blood supply to the head. Where is the deep branch of the MFCA anatomically located during this approach?

. Superficial to the piriformis tendon
. Between the superior gemellus and the obturator internus
. Between the quadratus femoris and the obturator externus
. Deep to the gluteus minimus muscle belly
. Within the ligamentum teres

Correct Answer & Explanation

. Superficial to the piriformis tendon


Explanation

The deep branch of the medial femoral circumflex artery passes consistently between the pectineus and iliopsoas, then courses posteriorly between the quadratus femoris and obturator externus muscles. Surgeons must be cautious when releasing the quadratus femoris to avoid brisk bleeding.

Question 583

Topic: Total Hip Arthroplasty (THA)

When templating for a total hip arthroplasty, the surgeon notes an anticipated leg length discrepancy. Lowering the center of rotation of the acetabular component without changing the femoral neck cut or stem size will result in which of the following?

. Increased leg length and increased femoral offset
. Increased leg length without changing femoral offset
. Decreased leg length and decreased femoral offset
. Decreased leg length without changing femoral offset
. No change in leg length or offset

Correct Answer & Explanation

. Increased leg length and increased femoral offset


Explanation

Lowering (inferiorizing) the acetabular center of rotation pushes the entire femur distally, thereby increasing leg length. It does not alter the horizontal distance (femoral offset) between the center of rotation and the femoral axis.

Question 584

Topic: Total Hip Arthroplasty (THA)

A 55-year-old male with a history of severe slipped capital femoral epiphysis (SCFE) undergoes a total hip arthroplasty. During the operation, the surgeon finds it difficult to restore femoral offset without excessively lengthening the limb. Sinking the femoral stem deeper is not possible due to a narrow diaphysis. What is the most appropriate next step?

. Use a standard offset stem with a longer head
. Use a high-offset stem with the appropriate neck length
. Perform a subtrochanteric shortening osteotomy
. Use a constrained acetabular liner to allow laxity
. Accept the limb lengthening to ensure stability

Correct Answer & Explanation

. Use a standard offset stem with a longer head


Explanation

A high-offset femoral stem increases the abductor lever arm (offset) without increasing vertical height (leg length). This is ideal when offset must be increased, but leg length is already equal or slightly long.

Question 585

Topic: Total Hip Arthroplasty (THA)

While performing an anterolateral approach to the hip, the surgeon proceeds to release the reflected head of the rectus femoris to improve acetabular exposure. Ligation of which of the following vascular structures is necessary to prevent significant hemorrhage during this step?

. Medial femoral circumflex artery
. Inferior gluteal artery
. Ascending branch of the lateral femoral circumflex artery
. First perforating branch of the profunda femoris
. Obturator artery

Correct Answer & Explanation

. Medial femoral circumflex artery


Explanation

The ascending branch of the lateral femoral circumflex artery transverses the operative field between the tensor fasciae latae and the sartorius/rectus interval. It must be identified and ligated during an anterolateral or direct anterior approach to prevent severe postoperative hematoma.

Question 586

Topic: Total Hip Arthroplasty (THA)

An MRI of the hip reveals an intra-articular mass requiring a surgical dislocation via a posterior approach. To protect the deep branch of the medial circumflex femoral artery (MCFA), the tendon of which muscle must be preserved or carefully handled?

. Piriformis
. Obturator internus
. Quadratus femoris
. Superior gemellus
. Obturator externus

Correct Answer & Explanation

. Piriformis


Explanation

The profound branch of the medial circumflex femoral artery (MCFA) runs posterior to the obturator externus and anterior to the short external rotators. Preserving the obturator externus tendon helps protect the main blood supply to the femoral head.

Question 587

Topic: Total Hip Arthroplasty (THA)

During a direct lateral (Hardinge) approach to the hip, the anterior third of the gluteus medius is split to gain access to the joint. Splitting the muscle too proximally risks denervating the remaining anterior portion of the muscle. What is the accepted safe zone for splitting the gluteus medius proximal to the tip of the greater trochanter to avoid injuring the superior gluteal nerve?

. 1 to 2 cm
. 3 to 5 cm
. 6 to 8 cm
. 9 to 11 cm
. 12 to 14 cm

Correct Answer & Explanation

. 1 to 2 cm


Explanation

The superior gluteal nerve courses approximately 3 to 5 cm proximal to the tip of the greater trochanter. Limiting the proximal split of the gluteus medius to less than 3 cm minimizes the risk of iatrogenic denervation.

Question 588

Topic: Total Hip Arthroplasty (THA)

Which of the following histological findings is pathognomonic for an Aseptic Lymphocyte-Dominated Vasculitis-Associated Lesion (ALVAL) in the setting of a failed metal-on-metal THA?

. Abundant neutrophils with intracellular bacteria
. Sheets of foamy macrophages containing birefringent polyethylene debris
. Perivascular lymphocytic infiltrates with tissue necrosis
. Non-caseating granulomas with giant cells
. Extensive acute fibrinoid necrosis entirely lacking lymphocytes

Correct Answer & Explanation

. Abundant neutrophils with intracellular bacteria


Explanation

ALVAL (Aseptic Lymphocyte-Dominated Vasculitis-Associated Lesion) is a Type IV delayed hypersensitivity reaction seen in metal-on-metal hip replacements. Histologically, it is defined by a heavy perivascular infiltrate of T-lymphocytes, macrophages, and varying degrees of tissue necrosis.

Question 589

Topic: Total Hip Arthroplasty (THA)

A 55-year-old female underwent THA with a ceramic-on-ceramic bearing. At her 2-year follow-up, she complains of a loud squeaking noise from her hip when bending to tie her shoes. Which of the following factors is most strongly associated with squeaking in ceramic-on-ceramic THA?

. Use of a larger femoral head (>36 mm)
. Cup anteversion greater than 20 degrees
. Edge loading due to component malposition or impingement
. High body mass index (>35 kg/m2)
. Uncorrected limb length discrepancy

Correct Answer & Explanation

. Use of a larger femoral head (>36 mm)


Explanation

Squeaking in ceramic-on-ceramic hips is multifactorial but is most strongly correlated with edge loading and stripe wear. This frequently results from component malposition (such as excessive cup inclination) or impingement causing micro-separation.

Question 590

Topic: Total Hip Arthroplasty (THA)

A dual mobility cup is often used in revision THA to reduce the risk of dislocation. Which of the following best describes the primary biomechanical principle of a dual mobility construct?

. It relies on a constrained rigid liner that locks the femoral head in place.
. It increases the 'jumping distance' by utilizing a large effective head size.
. It features an eccentric center of rotation to lateralize the femur.
. It increases the abductor lever arm through a specialized offset liner.
. It prevents impingement by utilizing a truncated femoral neck geometry.

Correct Answer & Explanation

. It relies on a constrained rigid liner that locks the femoral head in place.


Explanation

Dual mobility constructs feature a smaller inner head articulated within a large mobile polyethylene liner, which articulates with the metal shell. This mechanism essentially creates a very large effective head size, vastly increasing the 'jumping distance' required for dislocation.

Question 591

Topic: Total Hip Arthroplasty (THA)

When planning a primary THA, the surgeon uses the AP pelvis radiograph to assess leg length discrepancy.

Which of the following radiographic landmarks is most reliable for establishing a horizontal reference line for leg length measurement?

. Lesser trochanters
. Ischial tuberosities
. Superior aspect of the obturator foramina
. Radiographic teardrops
. Anterior superior iliac spines

Correct Answer & Explanation

. Lesser trochanters


Explanation

The inter-teardrop line is the most consistent and reliable horizontal reference line on an AP pelvis radiograph. It represents the inferior margin of the true acetabulum and is much less affected by pelvic tilt and rotation than the obturator foramina or ischial tuberosities.

Question 592

Topic: Total Hip Arthroplasty (THA)

Biomechanical optimization during Total Hip Arthroplasty (THA) is crucial for longevity and function. Increasing the femoral offset during a THA has which of the following mechanical effects?

. Decreases the abductor moment arm
. Increases the overall joint reaction force
. Increases the abductor moment arm and decreases the joint reaction force
. Decreases tension on the iliotibial (IT) band
. Shifts the anatomical center of rotation superiorly

Correct Answer & Explanation

. Decreases the abductor moment arm


Explanation

Increasing the femoral offset increases the moment arm of the hip abductors. Because the abductor moment arm is longer, less force is required by the abductor muscles to maintain pelvic level during the single-leg stance phase of gait. The decreased necessary muscle force translates directly into a decreased overall joint reaction force on the hip.

Question 593

Topic: Total Hip Arthroplasty (THA)

A 78-year-old female with a history of recurrent dislocations following a primary total hip arthroplasty is scheduled for revision surgery.

The surgeon decides to use a dual mobility construct. What is the defining biomechanical feature of a dual mobility articulation?

. It utilizes a constrained liner that locks the femoral head in place with a locking ring.
. It consists of a large inner metallic head that articulates with a fixed, thin polyethylene liner.
. It utilizes a mobile polyethylene liner that articulates with both a smaller inner femoral head and a highly polished metallic outer acetabular shell.
. It prevents dislocation by increasing the offset of the femoral stem without changing the center of rotation.
. It is composed entirely of cross-linked polyethylene to allow for greater jump distance.

Correct Answer & Explanation

. It utilizes a constrained liner that locks the femoral head in place with a locking ring.


Explanation

A dual mobility construct features two distinct articulations: a smaller standard femoral head that articulates within a larger, mobile polyethylene liner. This mobile liner then articulates within a highly polished, fixed metallic acetabular shell. This design maximizes the effective head size, significantly increasing the jump distance and range of motion before impingement, thereby substantially reducing the risk of dislocation.

Question 594

Topic: Total Hip Arthroplasty (THA)

A 65-year-old female presents with recurrent posterior dislocations following a primary THA. Radiographs demonstrate adequate acetabular component position, but the femoral component has decreased offset compared to the contralateral side. Increasing the femoral offset during revision provides which of the following biomechanical advantages?

. Increases the joint reaction force across the hip
. Decreases the moment arm of the abductor musculature
. Increases tension on the abductor musculature and soft tissues
. Medializes the center of rotation of the hip joint
. Decreases the functional leg length

Correct Answer & Explanation

. Increases the joint reaction force across the hip


Explanation

Increasing femoral offset lateralizes the femur, which restores the abductor moment arm, increases abductor tension, and reduces bony impingement, thereby enhancing stability.

Question 595

Topic: Total Hip Arthroplasty (THA)

During acetabular preparation in THA, a screw is placed in the anterior-superior quadrant of the acetabulum based on the Wasielewski safe zones. Which neurovascular structure is at greatest risk of injury?

. Obturator nerve
. Obturator artery
. Sciatic nerve
. Superior gluteal nerve
. External iliac artery and vein

Correct Answer & Explanation

. Obturator nerve


Explanation

Screws placed in the anterior-superior quadrant place the external iliac artery and vein at risk. The posterior-superior and posterior-inferior quadrants are considered the safest zones.

Question 596

Topic: Total Hip Arthroplasty (THA)

A 62-year-old female is undergoing a direct anterior approach for a THA. The surgeon utilizes the interval between the tensor fasciae latae and the sartorius. Which of the following structures is at the highest risk of injury during the initial superficial dissection?

. Femoral nerve
. Sciatic nerve
. Superior gluteal nerve
. Lateral femoral cutaneous nerve
. Ascending branch of the lateral circumflex femoral artery

Correct Answer & Explanation

. Femoral nerve


Explanation

The lateral femoral cutaneous nerve crosses the superficial interval of the direct anterior approach and is at highest risk of neuropraxia or transection during the superficial dissection.

Question 597

Topic: Total Hip Arthroplasty (THA)

A 45-year-old active male underwent a total hip arthroplasty utilizing a ceramic-on-ceramic bearing. Two years later, he complains of an audible squeaking sound from the hip during deep flexion activities. Which of the following factors is most strongly associated with the development of this phenomenon?

. Use of a 28mm ceramic head
. Improper cup positioning leading to edge loading
. Use of a cemented femoral stem
. High level of activity in a low BMI patient
. Use of highly cross-linked polyethylene

Correct Answer & Explanation

. Use of a 28mm ceramic head


Explanation

Squeaking in ceramic-on-ceramic THA is highly associated with edge loading, which often results from component malposition (such as excessive cup anteversion or inclination), microseparation, and higher patient BMI.

Question 598

Topic: Total Hip Arthroplasty (THA)

During templating for a primary THA, the surgeon notes a 15 mm leg length discrepancy (affected leg is short) and a highly offset native femur. If a standard offset stem is used instead of a high offset stem, what compensatory intraoperative error is most likely to occur in an attempt to restore soft tissue tension?

. Excessive medialization of the cup
. Increasing acetabular anteversion
. Shortening the leg excessively
. Excessive lengthening of the operative leg
. Decreasing acetabular inclination

Correct Answer & Explanation

. Excessive medialization of the cup


Explanation

If a standard offset stem is used in a patient requiring high offset, the abductors will remain lax. To achieve stability and proper soft tissue tension, the surgeon may inadvertently increase the neck length, resulting in an excessively lengthened leg.

Question 599

Topic: Total Hip Arthroplasty (THA)

A 42-year-old highly active female undergoes THA using a ceramic-on-ceramic bearing. One year later, she complains of a loud, audible squeaking noise from the hip during specific movements, though she is pain-free. What is the most significant risk factor for this phenomenon?

. Use of a large diameter femoral head
. Femoral stem retroversion
. Acetabular component malposition leading to edge loading
. Body mass index less than 25
. High levels of daily physical activity

Correct Answer & Explanation

. Use of a large diameter femoral head


Explanation

Squeaking in ceramic-on-ceramic THA is most strongly associated with component malposition (such as a steeply inclined or excessively anteverted/retroverted cup). This malposition leads to edge loading, loss of fluid-film lubrication, and stripe wear.

Question 600

Topic: Total Hip Arthroplasty (THA)

A 75-year-old male with Parkinson's disease and a history of recurrent posterior hip dislocations is scheduled for revision THA. A dual mobility construct is selected. What is the primary biomechanical advantage of a dual mobility bearing that reduces dislocation risk?

. It provides a constrained locking ring that physically prevents the head from escaping the liner
. It increases the effective head diameter to increase jump distance while allowing motion at two separate articulations
. It shifts the center of rotation medially to increase the abductor moment arm
. It severely restricts total arc of motion, thereby preventing impingement
. It utilizes a completely fixed acetabular liner to eliminate sheer forces

Correct Answer & Explanation

. It provides a constrained locking ring that physically prevents the head from escaping the liner


Explanation

A dual mobility bearing features a small inner metal or ceramic head that articulates within a larger unconstrained mobile polyethylene head, which in turn articulates within the metallic acetabular shell. The primary advantage is an increased effective head diameter (the size of the large poly head), which drastically increases the jump distance required for dislocation. The dual articulation minimizes wear compared to a traditional monolithic large head design.