Question 201

Topic: Total Hip Arthroplasty (THA)

To optimize stability and minimize impingement in THA, the concept of combined anteversion is often utilized. If the acetabular cup is placed in 15 degrees of anteversion, what is the ideal target for femoral stem anteversion to achieve a standard combined anteversion of 35-40 degrees?

. 5-10 degrees
. 10-15 degrees
. 20-25 degrees
. 30-35 degrees
. 40-45 degrees

Correct Answer & Explanation

. 20-25 degrees


Explanation

Combined anteversion is the sum of cup anteversion and femoral stem anteversion. The classic target range (Ranawat/Dorr) is 25-45 degrees. If the cup is at 15 degrees, a femoral anteversion of 20-25 degrees will achieve a combined anteversion of 35-40 degrees, providing optimal stability.

Question 202

Topic: Total Hip Arthroplasty (THA)

Which of the following surgical factors is most strongly associated with the postoperative complication of 'squeaking' in a ceramic-on-ceramic total hip arthroplasty?

. Impingement of the iliopsoas tendon
. Use of a 36-mm or larger femoral head
. Acetabular component malposition leading to edge loading
. Decreased femoral offset
. Retention of the anterior capsule

Correct Answer & Explanation

. Acetabular component malposition leading to edge loading


Explanation

Squeaking in ceramic-on-ceramic THA is most strongly associated with edge loading. This is typically caused by acetabular component malposition, such as excessive inclination or anteversion.

Question 203

Topic: Total Hip Arthroplasty (THA)

A 62-year-old male presents with new-onset cardiomyopathy, peripheral neuropathy, and visual field changes 6 years after a revision total hip arthroplasty. Aspiration of the hip yields clear fluid with a low white blood cell count. Which of the following implant configurations is most likely responsible for his systemic symptoms?

. Ceramic-on-highly crosslinked polyethylene
. Metal-on-metal surface replacement
. Dual mobility construct with a ceramic head
. Modular metal-on-polyethylene with a titanium stem and a large cobalt-chrome head
. Oxinium-on-polyethylene bearing

Correct Answer & Explanation

. Modular metal-on-polyethylene with a titanium stem and a large cobalt-chrome head


Explanation

The patient's systemic symptoms describe arthroprosthetic cobaltism. In the absence of a metal-on-metal bearing, severe trunnionosis at the head-neck junction of a cobalt-chrome head on a titanium stem is a known source of toxic cobalt levels.

Question 204

Topic: Total Hip Arthroplasty (THA)

A 55-year-old male undergoes a primary THA via a direct anterior approach. Postoperatively, he notes a burning pain and numbness isolated to the anterolateral aspect of his operative thigh. Strength is fully intact. Injury to which nerve is the most likely cause?

. Femoral nerve
. Lateral femoral cutaneous nerve
. Obturator nerve
. Superior gluteal nerve
. Sciatic nerve

Correct Answer & Explanation

. Lateral femoral cutaneous nerve


Explanation

The lateral femoral cutaneous nerve (LFCN) is a purely sensory nerve at risk during the direct anterior approach to the hip. Injury results in meralgia paresthetica, characterized by numbness and dysesthesias over the anterolateral thigh.

Question 205

Topic: Total Hip Arthroplasty (THA)

A 76-year-old female presents with a third episode of posterior dislocation following a primary THA. Radiographs show well-fixed, correctly positioned components. Intraoperatively, her abductor musculature is found to be completely avulsed, retracted, and irreparable. Which of the following is the most appropriate definitive management?

. Revision to a larger femoral head using the existing acetabular cup
. Revision of the acetabular component to increase anteversion by 10 degrees
. Placement of a constrained acetabular liner
. Revision to a dual mobility acetabular construct
. Prophylactic application of a hip abduction orthosis

Correct Answer & Explanation

. Placement of a constrained acetabular liner


Explanation

In the setting of recurrent instability with properly positioned components and a severely deficient, irreparable abductor mechanism, a constrained acetabular liner is indicated. Dual mobility components rely on a functional abductor mechanism to prevent dislocation.

Question 206

Topic: Total Hip Arthroplasty (THA)

A 68-year-old female presents with progressive groin pain 6 years after a metal-on-polyethylene THA utilizing a modular titanium stem and cobalt-chrome head. Aspiration yields fluid with 2,500 WBC/uL (80% lymphocytes) and negative cultures. Serum cobalt is markedly elevated. What is the most appropriate definitive management?

. Single-stage revision to a ceramic head with a titanium adapter sleeve
. Two-stage revision with an articulating antibiotic spacer
. Debridement, antibiotics, and implant retention (DAIR)
. Prolonged oral antibiotic suppression
. Observation and physical therapy

Correct Answer & Explanation

. Single-stage revision to a ceramic head with a titanium adapter sleeve


Explanation

This patient has trunnionosis with an adverse local tissue reaction (ALVAL), characterized by elevated cobalt and lymphocytic synovial fluid with negative cultures. Treatment involves revising the head to a ceramic head with a titanium sleeve to bypass the damaged trunnion, provided the stem is well-fixed.

Question 207

Topic: Total Hip Arthroplasty (THA)

A 45-year-old active male underwent a primary THA. Two years postoperatively, he complains of an audible squeaking noise with walking and bending, but denies pain. Radiographs show well-fixed, well-positioned components. Which bearing surface combination is most classically associated with this complication?

. Metal-on-polyethylene
. Ceramic-on-ceramic
. Ceramic-on-highly cross-linked polyethylene
. Metal-on-metal
. Oxinium-on-polyethylene

Correct Answer & Explanation

. Ceramic-on-ceramic


Explanation

Audible squeaking is a well-documented complication specific to ceramic-on-ceramic THA bearings. It is often benign but can be related to microseparation, edge loading, or impingement.

Question 208

Topic: Total Hip Arthroplasty (THA)

During preoperative templating for a total hip arthroplasty, the surgeon notes a 15 mm functional leg length discrepancy (affected side short) and a 10 mm anatomic true leg length discrepancy (affected side short). To avoid symptomatic over-lengthening postoperatively, the surgeon should aim to restore:

. Only the functional leg length discrepancy (15 mm)
. Only the anatomic true leg length discrepancy (10 mm)
. The sum of both discrepancies (25 mm)
. Global offset instead of length, ignoring the discrepancy
. A net shortening of the limb to relax the sciatic nerve

Correct Answer & Explanation

. Only the anatomic true leg length discrepancy (10 mm)


Explanation

The goal of THA is to restore true anatomic leg length. Functional leg length discrepancies are often due to pelvic obliquity from spine pathology or soft tissue contractures, which typically adapt over time; correcting to functional length leads to true anatomic over-lengthening.

Question 209

Topic: Total Hip Arthroplasty (THA)

A 74-year-old female presents for revision THA. Radiographs demonstrate an inferiorly migrated acetabular component, a transverse radiolucent line across the posterior column, and an ilioischial line break. Intraoperatively, the superior and inferior hemi-pelves move independently. Which of the following is the most appropriate reconstructive technique?

. Standard hemispherical porous-coated cup with multiple dome screws
. Impaction bone grafting and a cemented polyethylene cup
. Cup-cage construct or custom triflange acetabular component
. Bilobed (oblong) uncemented cup without supplementary fixation
. Jumbo uncemented cup alone

Correct Answer & Explanation

. Standard hemispherical porous-coated cup with multiple dome screws


Explanation

Independent movement of the superior and inferior hemi-pelves indicates pelvic discontinuity. Reconstructive options must stabilize the two halves, typically requiring a cup-cage construct, a custom triflange component, or highly porous metal augments with a distraction technique.

Question 210

Topic: Total Hip Arthroplasty (THA)

A 68-year-old female presents with recurrent episodes of anterior hip dislocation following a primary total hip arthroplasty. Her dislocations typically occur when she extends and externally rotates her hip while walking. Which of the following component malpositions is the most likely cause of this specific instability pattern?

. Retroverted acetabular component and retroverted femoral stem
. Excessively anteverted acetabular component and retroverted femoral stem
. Excessively anteverted acetabular component and excessively anteverted femoral stem
. Retroverted acetabular component and anteverted femoral stem
. Excessively abducted acetabular component with neutral femoral stem version

Correct Answer & Explanation

. Retroverted acetabular component and retroverted femoral stem


Explanation

Excessive combined anteversion of the acetabular and femoral components leads to anterior instability. This typically occurs in extension and external rotation, as the components lever against each other anteriorly.

Question 211

Topic: Total Hip Arthroplasty (THA)

A 65-year-old female undergoes primary total hip arthroplasty (THA) via a posterior approach. Post-operatively, she experiences recurrent anterior dislocations.

Which of the following component malpositions is the most likely cause of her anterior instability?

. Excessive femoral retroversion
. Excessive acetabular retroversion
. Excessive combined anteversion
. Inadequate femoral offset
. Leg length discrepancy with a short operative leg

Correct Answer & Explanation

. Excessive combined anteversion


Explanation

Anterior dislocation of a THA is most commonly caused by excessive combined anteversion (acetabular anteversion + femoral anteversion). When the combined anteversion is too high, the femoral head is driven anteriorly out of the acetabulum, especially during hip extension and external rotation. Conversely, excessive retroversion typically leads to posterior instability.

Question 212

Topic: Total Hip Arthroplasty (THA)

A 65-year-old male is 3 months post-operative from a primary total hip arthroplasty (THA) via a posterior approach. He presents with a history of three posterior dislocations. Radiographs demonstrate a well-fixed acetabular component with 45 degrees of abduction and 20 degrees of anteversion. The femoral stem is stable but appears retroverted by 10 degrees. What is the most appropriate surgical intervention to definitively address this instability?

. Prescribe a rigid hip abduction orthosis for 3 months
. Revise the acetabular component to increase anteversion to 35 degrees
. Revise the femoral component to increase anteversion
. Revise the acetabular liner to a constrained liner design
. Perform a greater trochanteric advancement

Correct Answer & Explanation

. Revise the femoral component to increase anteversion


Explanation

Combined anteversion is the critical factor in preventing THA instability. The ideal combined anteversion is typically 25 to 35 degrees for males. The acetabular component is already well-positioned (45 deg abduction, 20 deg anteversion). The femoral stem, however, is in 10 degrees of retroversion, resulting in a low combined anteversion (10 degrees total) that predisposes the patient to posterior dislocation. Revising the femoral component to appropriate anteversion (10-15 degrees) restores normal biomechanics. Over-anteverting the cup (Option B) can lead to anterior instability. A constrained liner (Option D) is a salvage option but does not address the underlying significant malposition.

Question 213

Topic: Total Hip Arthroplasty (THA)

In ceramic-on-ceramic total hip arthroplasty, which of the following component malpositions is most strongly associated with the complication of squeaking?

. Femoral stem retroversion
. Acetabular component retroversion
. Excessive acetabular inclination
. Decreased femoral offset
. Excessive acetabular anteversion

Correct Answer & Explanation

. Femoral stem retroversion


Explanation

Squeaking in ceramic-on-ceramic THA is often attributed to edge loading or microseparation. Excessive acetabular inclination (a steep cup, typically >50 degrees) or extreme versions lead to edge loading, disrupting the fluid film lubrication, causing stripe wear, and resulting in squeaking.

Question 214

Topic: Total Hip Arthroplasty (THA)
In the posterior approach to the proximal radius (proximal Thompson approach), the supinator is exposed through the interval between what two muscles?
. Aconeus and extensor carpi ulnaris
. Extensor digitorum communis and extensor carpi ulnaris
. Extensor digitorum communis and aconeus
. Extensor carpi radialis brevis and extensor digitorum communis
. Extensor carpi radialis brevis and extensor carpi ulnaris

Correct Answer & Explanation

. Extensor carpi radialis brevis and extensor digitorum communis


Explanation

The proximal exposure of the radius is most often used for internal fixation of fractures, resection of tumors, or decompression of the posterior interosseous nerve beneath the supinator muscle. The supinator muscle is exposed through the interval between the extensor carpi radialis brevis and the extensor digitorum communis muscles. This interval can be more easily palpated further distal in the forearm.

Question 215

Topic: Total Hip Arthroplasty (THA)

A 45-year-old male with a ceramic-on-ceramic total hip arthroplasty presents 3 years postoperatively complaining of a loud squeaking noise with walking. Radiographs demonstrate well-fixed components. Which of the following factors is most strongly associated with the etiology of this phenomenon?

. Acetabular component retroversion
. Increased femoral offset
. Body mass index > 35 kg/m^2
. Stripe wear from edge loading due to component malposition
. Use of a titanium femoral stem

Correct Answer & Explanation

. Stripe wear from edge loading due to component malposition


Explanation

Squeaking in ceramic-on-ceramic THA is a well-documented complication occurring in approximately 1-10% of patients. It is most strongly associated with edge loading of the ceramic bearings, which disrupts the fluid film lubrication. Edge loading typically occurs due to component malposition (e.g., excessive acetabular anteversion, high abduction angle), which leads to 'stripe wear' on the ceramic head and subsequent squeaking. While a titanium stem might be present, the edge loading mechanism is the primary cause.

Question 216

Topic: Total Hip Arthroplasty (THA)

A patient undergoes a primary total hip arthroplasty (THA) via a posterior approach. Six weeks postoperatively, the patient sustains a posterior dislocation while sitting in a low chair. Which combination of component malpositioning most commonly predisposes to this specific pattern of instability?

. Acetabular excessive anteversion and femoral retroversion
. Acetabular retroversion and femoral retroversion
. Acetabular excessive abduction and femoral excessive anteversion
. Acetabular retroversion and femoral excessive anteversion
. Acetabular excessive anteversion and femoral excessive anteversion

Correct Answer & Explanation

. Acetabular retroversion and femoral retroversion


Explanation

Posterior dislocation of a THA typically occurs with a combination of hip flexion, internal rotation, and adduction. Component malposition that decreases the anterior coverage or opens the hip up posteriorly predisposes to this. Specifically, retroversion of the acetabular component and/or retroversion of the femoral stem significantly increases the risk of posterior dislocation.

Question 217

Topic: Total Hip Arthroplasty (THA)

When comparing the direct anterior approach (DAA) to the posterior approach for primary total hip arthroplasty, which of the following complications has a definitively higher incidence in the DAA cohort based on recent large registry and meta-analysis data?

. Postoperative hip dislocation
. Sciatic nerve palsy
. Lateral femoral cutaneous nerve neurapraxia
. Deep vein thrombosis
. Symptomatic leg length discrepancy

Correct Answer & Explanation

. Lateral femoral cutaneous nerve neurapraxia


Explanation

The direct anterior approach (DAA) utilizes the internervous plane between the tensor fasciae latae and sartorius. Due to its anatomic proximity in the superficial dissection, the lateral femoral cutaneous nerve is at a significantly higher risk of injury or neurapraxia compared to the posterior approach.

Question 218

Topic: Total Hip Arthroplasty (THA)

During the posterior approach to the hip (Kocher-Langenbeck), preservation of the blood supply to the femoral head is a critical consideration if the femoral head is to be retained. The deep branch of the medial femoral circumflex artery (MFCA) is protected primarily by which of the following intact structures?

. Piriformis tendon
. Obturator internus tendon
. Obturator externus tendon
. Quadratus femoris muscle
. Superior gemellus muscle

Correct Answer & Explanation

. Obturator internus tendon


Explanation

Anatomical studies by Gautier et al. demonstrated that the deep branch of the medial femoral circumflex artery (MFCA) runs posterior to the obturator externus tendon. Preserving the obturator externus during surgical dislocation or a posterior approach provides a protective barrier for the MFCA, which supplies the majority of blood to the femoral head.

Question 219

Topic: Total Hip Arthroplasty (THA)

A patient experiences recurrent posterior dislocations following a primary total hip arthroplasty. Radiographs and a CT scan demonstrate the acetabular component is well-fixed but placed in 5 degrees of retroversion. Which of the following is the most appropriate definitive management?

. Application of a hip abduction orthosis for 6 weeks
. Cementing a constrained liner into the existing acetabular shell
. Revision of the acetabular component to increase anteversion
. Revision of the femoral component to increase femoral offset
. Advancement of the greater trochanter to increase abductor tension

Correct Answer & Explanation

. Revision of the acetabular component to increase anteversion


Explanation

Recurrent posterior instability driven by acetabular retroversion requires correction of the underlying mechanical malposition. Revision of the well-fixed but malpositioned acetabular shell to appropriate anteversion is the definitive treatment.

Question 220

Topic: Total Hip Arthroplasty (THA)

Which factor is most highly associated with the clinical phenomenon of 'squeaking' in a ceramic-on-ceramic total hip arthroplasty?

. Component malposition leading to edge loading
. Use of a short femoral neck
. Use of an extended high-offset stem
. Patient body mass index (BMI) less than 25
. Use of a 28-mm femoral head

Correct Answer & Explanation

. Component malposition leading to edge loading


Explanation

Squeaking in ceramic-on-ceramic THA is highly associated with edge loading, which typically results from component malposition (such as excessive acetabular cup inclination or anteversion outliers). Edge loading causes a loss of fluid film lubrication and stripe wear, ultimately leading to audible squeaking.