Question 1581

Topic: Biomechanics & Biomaterials

When comparing the torsional rigidity of a solid intramedullary nail to a hollow nail of the same material, the polar moment of inertia is primarily determined by which dimensional parameter?

. The length of the nail squared
. The cross-sectional area
. The inner diameter cubed
. The radius to the fourth power
. The square of the wall thickness

Correct Answer & Explanation

. The radius to the fourth power


Explanation

Torsional rigidity is governed by the polar moment of inertia. For a solid cylinder, this value is directly proportional to the radius to the fourth power (r^4), demonstrating how small increases in nail diameter drastically increase torsional stiffness.

Question 1582

Topic: Biomechanics & Biomaterials

The combination of a 316L stainless steel screw with a titanium plate in an orthopedic fracture construct significantly increases the risk of which type of corrosion?

. Crevice corrosion
. Fretting corrosion
. Galvanic corrosion
. Pitting corrosion
. Intergranular corrosion

Correct Answer & Explanation

. Galvanic corrosion


Explanation

Galvanic corrosion occurs when two dissimilar metals with different electrochemical potentials are placed in contact within a conductive fluid environment. The less noble metal acts as an anode and undergoes accelerated corrosion.

Question 1583

Topic: Biomechanics & Biomaterials

In orthopedic biomechanics, the pullout strength of a standard cortical bone screw is most significantly increased by maximizing which of the following design parameters?

. Inner (core) diameter
. Outer (thread) diameter
. Thread pitch
. Screw shaft length
. Screw head diameter

Correct Answer & Explanation

. Outer (thread) diameter


Explanation

The pullout strength of a screw is most strongly related to the outer (thread) diameter because it dictates the volume of bone engaged between the threads. Increasing the inner core diameter increases bending strength but does not primarily maximize pullout resistance.

Question 1584

Topic: Biomechanics & Biomaterials

During the biomechanical testing of an articular cartilage explant, a constant strain is applied and maintained over time. The gradual decrease in the force required to maintain this constant strain is biomechanically defined as:

. Creep
. Stress relaxation
. Hysteresis
. Fatigue failure
. Yield point

Correct Answer & Explanation

. Stress relaxation


Explanation

Stress relaxation is the time-dependent decrease in stress (force) under a constant state of strain. Creep, conversely, is the progressive deformation (strain) of a material under a constant stress (load).

Question 1585

Topic: Biomechanics & Biomaterials

A surgeon revises a nonunion treated previously with a stainless steel plate by placing a new titanium plate over the existing retained stainless steel screws. Which type of corrosion is most likely to rapidly degrade the implant interface?

. Crevice corrosion
. Fretting corrosion
. Galvanic corrosion
. Pitting corrosion
. Intergranular corrosion

Correct Answer & Explanation

. Galvanic corrosion


Explanation

Galvanic corrosion occurs when two dissimilar metals are placed in physical contact within a conducting fluid medium (like body fluids). This creates an electrochemical cell, leading to accelerated corrosion of the less noble metal.

Question 1586

Topic: Biomechanics & Biomaterials

A surgeon places a 316L stainless steel screw through a titanium alloy plate to fix a fracture. What type of corrosion is most likely to occur at the interface between the two different metals?

. Fretting corrosion
. Galvanic corrosion
. Crevice corrosion
. Pitting corrosion
. Stress corrosion cracking

Correct Answer & Explanation

. Galvanic corrosion


Explanation

Galvanic corrosion occurs when two dissimilar metals are placed in physical contact within a conductive fluid (such as serum/body fluid). The metal that is less noble (anodic) will corrode preferentially while the more noble (cathodic) metal is protected. Fretting corrosion, by contrast, involves micromotion between two surfaces.

Question 1587

Topic: Biomechanics & Biomaterials
A surgeon is evaluating implant options for a total hip arthroplasty. Which method of processing ultra-high molecular weight polyethylene (UHMWPE) maximizes its wear resistance while minimizing the risk of long-term in vivo oxidative degradation?
. Gamma irradiation in air without thermal treatment
. Ethylene oxide gas sterilization
. Gamma irradiation in a vacuum followed by remelting
. Autoclaving at 120°C
. Gamma irradiation in an inert gas followed by annealing

Correct Answer & Explanation

. Gamma irradiation in a vacuum followed by remelting


Explanation

Gamma irradiation creates cross-links in UHMWPE, which significantly improves wear resistance. However, irradiation also produces free radicals. If these free radicals are exposed to oxygen in vivo, they cause oxidative degradation, leading to embrittlement and catastrophic failure of the polyethylene. Heating the polyethylene above its melting point (remelting) after irradiation quenches these free radicals, preventing oxidation, although it slightly decreases the material's yield strength and fatigue resistance compared to annealing. Annealing (heating below the melting point) leaves residual free radicals.

Question 1588

Topic: Biomechanics & Biomaterials

When optimizing a cortical screw for fixation in diaphyseal bone, which of the following design modifications will most significantly increase its pullout strength?

. Increasing the inner (root) diameter
. Changing the material from titanium alloy to 316L stainless steel
. Decreasing the thread pitch
. Decreasing the outer (thread) diameter
. Increasing the screw head diameter

Correct Answer & Explanation

. Decreasing the thread pitch


Explanation

Screw pullout strength in bone is directly proportional to the outer (thread) diameter and the length of thread engagement, and it is inversely proportional to the thread pitch. Decreasing the thread pitch means there are more threads per unit length, which increases the amount of bone engaged and therefore increases pullout strength. Increasing the inner (root) diameter increases the screw's bending stiffness and resistance to fatigue failure, but does not increase pullout strength.

Question 1589

Topic: Biomechanics & Biomaterials

A ligament is subjected to a constant, sustained load over a period of time, resulting in a progressive increase in deformation. Which of the following biomechanical properties does this describe?

. Stress relaxation
. Creep
. Hysteresis
. Fatigue
. Isotropy

Correct Answer & Explanation

. Creep


Explanation

Creep is the progressive deformation of a viscoelastic material under a constant load over time. Stress relaxation, in contrast, is the decrease in stress over time when the material is held at a constant strain.

Question 1590

Topic: Biomechanics & Biomaterials

During the clinical management of a clubfoot, a serial casting technique is utilized. The progressive deformation of the contracted soft tissues under a constant applied load over time demonstrates which of the following viscoelastic properties?

. Hysteresis
. Stress relaxation
. Creep
. Fatigue
. Anisotropy

Correct Answer & Explanation

. Creep


Explanation

Creep is defined as the progressive deformation of a viscoelastic material when it is subjected to a constant load over time. This principle is applied in serial casting to gradually lengthen contracted ligaments and tendons.

Question 1591

Topic: Biomechanics & Biomaterials

Which of the following material properties best explains why a titanium alloy implant exhibits a lower modulus of elasticity compared to a stainless steel implant, thereby reducing stress shielding in orthopaedic fixation?

. Increased yield strength
. Lower density
. Different atomic bonding and crystalline structure
. Increased fatigue strength
. Higher ultimate tensile strength

Correct Answer & Explanation

. Different atomic bonding and crystalline structure


Explanation

The modulus of elasticity (Young's modulus) is an intrinsic material property that dictates the stiffness of an implant. It is fundamentally determined by the atomic bonding and crystalline lattice structure of the metal. Titanium alloys have a modulus of elasticity closer to that of cortical bone compared to stainless steel or cobalt-chrome, reducing the stress-shielding effect. Yield strength, density, and fatigue strength are separate properties that do not dictate the intrinsic stiffness of the material.

Question 1592

Topic: Biomechanics & Biomaterials
The introduction of highly cross-linked polyethylene (HXLPE) in total hip arthroplasty has significantly reduced volumetric wear rates compared to conventional ultra-high-molecular-weight polyethylene (UHMWPE). However, the cross-linking and subsequent thermal stabilization processes alter the material's bulk mechanical properties. Which of the following represents a known mechanical tradeoff of HXLPE?
. Increased ultimate tensile strength
. Increased fatigue crack propagation resistance
. Decreased yield strength and ductility
. Increased oxidative degradation rate in vivo
. Generation of substantially larger wear particles

Correct Answer & Explanation

. Decreased yield strength and ductility


Explanation

Irradiation to create cross-links improves wear resistance but decreases the mechanical robustness of the polyethylene. Specifically, highly cross-linked polyethylene (HXLPE) exhibits decreased yield strength, ultimate tensile strength, ductility, and resistance to fatigue crack propagation compared to conventional UHMWPE. Thermal processing (e.g., remelting) removes free radicals to prevent oxidation but further reduces these mechanical properties.

Question 1593

Topic: Biomechanics & Biomaterials

A patient with a previously implanted stainless steel plate for a femur fracture undergoes revision surgery. The surgeon inadvertently uses a titanium screw to secure an adjunctive piece of hardware in direct physical contact with the stainless steel plate. Over time, rapid corrosion and eventual failure of the less noble metal occurs.

This process is best described as:

. Fretting corrosion
. Crevice corrosion
. Galvanic corrosion
. Pitting corrosion
. Stress corrosion cracking

Correct Answer & Explanation

. Galvanic corrosion


Explanation

Galvanic corrosion occurs when two dissimilar metals are placed in direct contact with each other within an electrolytic solution (like human bodily fluids). An electrochemical cell is created, and the less noble metal (the anode) undergoes accelerated corrosive degradation. Fretting corrosion is mechanically induced by micromotion between components, whereas crevice corrosion occurs in restricted spaces with oxygen depletion. Using titanium and stainless steel in direct contact is a classic setup for galvanic corrosion.

Question 1594

Topic: Biomechanics & Biomaterials

What is the predominant mechanism of corrosion seen at the modular head-neck taper junction (trunnionosis) in modern total hip arthroplasty?

. Uniform attack corrosion
. Mechanically assisted crevice corrosion
. Pitting corrosion
. Galvanic corrosion exclusively
. Intergranular corrosion

Correct Answer & Explanation

. Mechanically assisted crevice corrosion


Explanation

The primary mechanism of corrosion at modular junctions (like the head-neck trunnion) is Mechanically Assisted Crevice Corrosion (MACC). Micromotion (fretting) at the taper junction mechanically disrupts the protective passivation layer (titanium or chromium oxide). The fluid-filled gap (crevice) between the components becomes depleted of oxygen, leading to an acidic microenvironment that prevents repassivation and accelerates active metal dissolution.

Question 1595

Topic: Biomechanics & Biomaterials

A 65-year-old man presents with an acutely swollen and painful knee. Synovial fluid aspiration yields a yellow fluid with a WBC count of 45,000 cells/uL (predominantly PMNs). Polarized light microscopy reveals weakly positively birefringent rhomboid-shaped crystals. What is the composition of these crystals?

. Monosodium urate
. Calcium pyrophosphate dihydrate
. Basic calcium phosphate
. Cholesterol
. Hydroxyapatite

Correct Answer & Explanation

. Calcium pyrophosphate dihydrate


Explanation

The classic presentation of pseudogout (calcium pyrophosphate deposition disease, CPPD) includes an acutely inflamed joint with synovial fluid showing rhomboid-shaped crystals that exhibit weak positive birefringence under polarized light. Monosodium urate crystals (gout) are needle-shaped and strongly negatively birefringent.

Question 1596

Topic: Biomechanics & Biomaterials

Polymethylmethacrylate (PMMA) bone cement is frequently used in arthroplasty for implant fixation. Which of the following additives is responsible for making PMMA radiopaque on standard plain radiographs?

. Benzoyl peroxide
. N,N-dimethyl-p-toluidine
. Hydroquinone
. Barium sulfate
. Polystyrene

Correct Answer & Explanation

. Barium sulfate


Explanation

In standard PMMA formulations, barium sulfate (BaSO4) or zirconium dioxide (ZrO2) is added as a radiopacifier so the cement mantle can be evaluated radiographically. Benzoyl peroxide is the initiator in the powder, N,N-dimethyl-p-toluidine is the accelerator in the liquid, and hydroquinone is added to the liquid as a stabilizer to prevent premature polymerization.

Question 1597

Topic: Biomechanics & Biomaterials

Among the commonly used solid orthopedic implant metals, which of the following has a modulus of elasticity (Young's modulus) closest to that of cortical bone, thereby theoretically reducing the risk of stress shielding?

. Cobalt-chromium-molybdenum alloy
. 316L Stainless steel
. Titanium alloy (Ti-6Al-4V)
. Solid Tantalum
. Alumina ceramic

Correct Answer & Explanation

. Titanium alloy (Ti-6Al-4V)


Explanation

Cortical bone has a modulus of elasticity of approximately 15-20 GPa. While all standard solid metals are significantly stiffer than bone, Titanium alloy (Ti-6Al-4V) has a modulus of ~110 GPa, which is roughly half that of 316L Stainless Steel (~200 GPa) and Cobalt-chromium (~210-240 GPa). Therefore, titanium implants cause less stress shielding compared to stainless steel or CoCr implants.

Question 1598

Topic: Biomechanics & Biomaterials

Galvanic corrosion is an electrochemical process that can lead to early failure of orthopedic implants when two dissimilar metals are in contact in an electrolyte solution (such as body fluid). Which of the following metal combinations presents the highest theoretical risk for galvanic corrosion?

. Titanium and Cobalt-Chromium
. Stainless Steel and Titanium
. Cobalt-Chromium and Cobalt-Chromium
. Titanium and Tantalum
. Stainless Steel and Cobalt-Chromium

Correct Answer & Explanation

. Stainless Steel and Titanium


Explanation

Stainless steel and Titanium have the greatest difference in their anodic index (galvanic potential difference). When placed in contact within an electrolytic environment, stainless steel acts as the anode and corrodes rapidly, while titanium acts as the cathode. Thus, mixing stainless steel and titanium implants (e.g., a stainless steel screw in a titanium plate) is contraindicated.

Question 1599

Topic: Biomechanics & Biomaterials

In biomechanics, the mechanical behavior of a material is plotted on a stress-strain curve. The total area under the stress-strain curve from the origin up to the point of ultimate failure represents which mechanical property?

. Stiffness
. Yield strength
. Ultimate tensile strength
. Toughness
. Ductility

Correct Answer & Explanation

. Toughness


Explanation

Toughness is the amount of energy a material can absorb before it fails, represented by the total area under the stress-strain curve. Stiffness is the slope of the elastic region (Young's modulus). Yield strength is the point where plastic deformation begins. Ultimate tensile strength is the peak stress on the curve. Ductility is the amount of plastic strain before failure.

Question 1600

Topic: Biomechanics & Biomaterials

What is the primary purpose of heating (melting or annealing) highly cross-linked polyethylene (HXLPE) above its melting point after gamma irradiation during manufacturing?

. To increase crystallinity and density of the polyethylene
. To eliminate free radicals and prevent long-term oxidative degradation
. To increase the ultimate tensile strength of the material
. To increase the elastic modulus to match cortical bone
. To promote further cross-linking between polymer chains

Correct Answer & Explanation

. To eliminate free radicals and prevent long-term oxidative degradation


Explanation

Irradiation of polyethylene creates cross-links that improve wear resistance, but it also leaves behind free radicals. If left untreated, these free radicals react with oxygen in vivo, leading to oxidative chain scission and degradation of the material's mechanical properties. Post-irradiation thermal treatment (either melting above the melting point or annealing just below it) quenches these free radicals, preventing oxidation.