Question 721

Topic: Biomechanics & Biomaterials

The torsional rigidity of a solid intramedullary nail is proportional to its radius raised to which power?

. First power
. Second power
. Third power
. Fourth power
. Fifth power

Correct Answer & Explanation

. First power


Explanation

Torsional rigidity of a solid cylinder is determined by its polar moment of inertia, which is proportional to the radius to the fourth power (r^4).

Question 722

Topic: Biomechanics & Biomaterials

Galvanic corrosion is most likely to occur under which of the following conditions in orthopedic implants?

. A titanium plate is used with titanium screws
. A stainless steel screw is placed in contact with a titanium plate
. A fretting interface exists between a screw head and plate hole of the same material
. An implant is subjected to highly repetitive cyclic loading
. There is a lack of oxygen at the fracture site

Correct Answer & Explanation

. A titanium plate is used with titanium screws


Explanation

Galvanic corrosion occurs when two dissimilar metals (e.g., stainless steel and titanium) are in physical contact within a conductive fluid environment like human body fluid.

Question 723

Topic: Biomechanics & Biomaterials

In total joint arthroplasty, the phenomenon of "third-body wear" is best classified as which type of wear mechanism?

. Adhesive wear
. Abrasive wear
. Fatigue wear
. Corrosive wear
. Fretting wear

Correct Answer & Explanation

. Adhesive wear


Explanation

Third-body wear is a form of abrasive wear caused by hard particles (like bone, cement, or metal debris) caught between the articulating surfaces, which scratch and gouge the softer material.

Question 724

Topic: Biomechanics & Biomaterials

Which of the following biomechanical terms best describes the phenomenon where a ligament subjected to a constant, sustained load exhibits a progressive increase in deformation over time?

. Stress relaxation
. Creep
. Hysteresis
. Fatigue failure
. Strain rate dependency

Correct Answer & Explanation

. Stress relaxation


Explanation

Creep is the progressive deformation of a viscoelastic material when subjected to a constant load over time. Stress relaxation, conversely, is the decrease in stress within a material when it is held at a constant deformation.

Question 725

Topic: Biomechanics & Biomaterials

On a standard stress-strain curve for a given orthopaedic implant, the total area under the curve prior to the point of structural failure represents the material's:

. Stiffness
. Toughness
. Ductility
. Yield strength
. Fatigue limit

Correct Answer & Explanation

. Stiffness


Explanation

Toughness is defined as the ability of a material to absorb energy and deform plastically before fracturing, represented by the total area under the stress-strain curve. Stiffness is represented by the slope of the linear elastic portion (Young's modulus).

Question 726

Topic: Biomechanics & Biomaterials

The use of a stainless steel screw through a titanium plate in an internal fixation construct is generally avoided in orthopedic surgery due to the risk of which type of corrosion?

. Crevice corrosion
. Fretting corrosion
. Galvanic corrosion
. Pitting corrosion
. Intergranular corrosion

Correct Answer & Explanation

. Crevice corrosion


Explanation

Galvanic corrosion occurs when two dissimilar metals with different electrochemical potentials are placed in contact within a conductive fluid (such as bodily fluids). The less noble metal undergoes accelerated corrosion.

Question 727

Topic: Biomechanics & Biomaterials

When analyzing the biomechanical properties of a hollow intramedullary nail, its torsional rigidity (polar moment of inertia) depends most heavily on which geometric parameter?

. Inner diameter to the second power
. Outer radius to the fourth power
. Total cross-sectional area
. Length of the nail to the third power
. Wall thickness to the second power

Correct Answer & Explanation

. Inner diameter to the second power


Explanation

The polar moment of inertia for a hollow cylinder is proportional to the difference between the outer radius to the fourth power and the inner radius to the fourth power. Thus, small increases in the outer diameter exponentially increase torsional rigidity.

Question 728

Topic: Biomechanics & Biomaterials

A 45-year-old woman reports an acute pop in the back of her knee while squatting. MRI reveals a medial meniscus posterior root tear with 3 mm of extrusion. Which of the following biomechanical changes occurs in the knee as a result of this injury?

. Decreased peak contact pressure in the medial compartment
. Increased contact area in the medial compartment
. Loss of hoop stresses leading to equivalent biomechanics of a total meniscectomy
. Increased anterior tibial translation comparable to an ACL tear
. Decreased varus laxity

Correct Answer & Explanation

. Decreased peak contact pressure in the medial compartment


Explanation

A posterior root tear of the medial meniscus completely disrupts the circumferential hoop stresses. This leads to functional total meniscectomy biomechanics with dramatically increased contact pressures and decreased contact area.

Question 729

Topic: Biomechanics & Biomaterials

Which historical sterilization method for ultra-high-molecular-weight polyethylene (UHMWPE) is most associated with subsequent severe oxidation, delamination, and rapid wear in THA?

. Gamma irradiation in a vacuum
. Gamma irradiation in air
. Ethylene oxide gas
. Gas plasma sterilization
. Electron beam irradiation in an inert gas

Correct Answer & Explanation

. Gamma irradiation in a vacuum


Explanation

Gamma irradiation in air produces free radicals that react with ambient oxygen over time. This leads to severe oxidation, chain scission, and catastrophic delamination and wear of the polyethylene.

Question 730

Topic: Biomechanics & Biomaterials

Highly cross-linked polyethylene (HXLPE) is widely used in modern THA to reduce wear rates. What is the primary mechanical trade-off associated with increasing the radiation dose used to create cross-linking in polyethylene?

. Increased volumetric wear
. Decreased oxidation resistance
. Decreased fracture toughness and fatigue strength
. Increased biological reactivity of wear debris
. Increased modulus of elasticity leading to stress shielding

Correct Answer & Explanation

. Increased volumetric wear


Explanation

While high-dose radiation increases cross-linking and dramatically reduces wear, it significantly decreases the mechanical properties of the polyethylene, specifically reducing its fracture toughness, ultimate tensile strength, and fatigue strength. This raises the risk of rim fracture.

Question 731

Topic: Biomechanics & Biomaterials

A 65-year-old active man undergoes a total hip arthroplasty using a highly cross-linked polyethylene liner. Which of the following mechanical properties is most significantly decreased in highly cross-linked polyethylene compared to conventional ultra-high-molecular-weight polyethylene (UHMWPE)?

. Wear resistance
. Fatigue crack propagation resistance
. Oxidation resistance
. Elastic modulus
. Material density

Correct Answer & Explanation

. Wear resistance


Explanation

Highly cross-linked polyethylene (HXLPE) significantly improves wear resistance but decreases mechanical properties such as fatigue strength and crack propagation resistance. This makes it more susceptible to fracture, particularly if thin liners are used or impingement occurs.

Question 732

Topic: Biomechanics & Biomaterials

Trunnionosis has emerged as a significant mode of failure in modern total hip arthroplasty, particularly involving modular head-neck junctions. This process is most accurately described by which of the following mechanisms?

. Galvanic corrosion exclusively due to dissimilar metals
. Abrasive wear from third-body bone cement particles
. Mechanically assisted crevice corrosion
. Polyethylene delamination
. Adhesive wear from edge loading

Correct Answer & Explanation

. Galvanic corrosion exclusively due to dissimilar metals


Explanation

Trunnionosis refers to the wear and corrosion at the modular head-neck taper interface. It is primarily driven by mechanically assisted crevice corrosion (MACC), which is a combination of fretting wear and subsequent electrochemical corrosion.

Question 733

Topic: Biomechanics & Biomaterials

In modern cementing techniques for total hip arthroplasty (third-generation cementing), vacuum mixing of the polymethylmethacrylate (PMMA) bone cement is standard practice. What is the primary biomechanical benefit of vacuum mixing?

. It accelerates the polymerization time to allow faster stem insertion.
. It decreases the porosity of the cement, thereby increasing its fatigue strength.
. It prevents the exothermic reaction of PMMA from causing thermal necrosis of bone.
. It increases the viscosity of the cement to prevent distal migration past the restrictor.
. It chemically bonds the cement to the titanium femoral stem.

Correct Answer & Explanation

. It accelerates the polymerization time to allow faster stem insertion.


Explanation

Vacuum mixing removes air bubbles from the PMMA during the mixing process, which significantly reduces the cement's macroporosity. This reduction in porosity eliminates stress risers, markedly increasing the fatigue life and tensile strength of the cement mantle.

Question 734

Topic: Biomechanics & Biomaterials

A 35-year-old woman is 4 months postoperative from open reduction and internal fixation of an ankle fracture, which included placement of two 3.5mm trans-syndesmotic screws. She is completely asymptomatic and asks if the screws must be removed. Based on current orthopedic literature, what is the recommendation regarding routine removal of syndesmotic screws?

. They must be removed at 3 months to prevent catastrophic hardware failure.
. They must be removed prior to full weight-bearing to restore normal ankle kinematics.
. Routine removal is not required, as retained or broken screws do not significantly worsen clinical outcomes.
. They should be removed only if they were placed rigidly through 4 cortices.
. Removal is mandatory to prevent late-onset distal tibiofibular synostosis.

Correct Answer & Explanation

. They must be removed at 3 months to prevent catastrophic hardware failure.


Explanation

Current evidence demonstrates that routine removal of syndesmotic screws is unnecessary. Clinical outcomes are comparable whether the screws are removed, retained intact, or broken, and elective removal exposes the patient to unnecessary surgical risks.

Question 735

Topic: Biomechanics & Biomaterials

Which of the following manufacturing processes is most effective at reducing the wear rate of ultra-high-molecular-weight polyethylene (UHMWPE) while minimizing the risk of long-term in vivo oxidation?

. Gamma irradiation in air followed by shelf storage
. High-dose irradiation followed by thermal remelting
. High-dose irradiation followed by sub-melting annealing without subsequent sterilization
. Addition of barium sulfate prior to extrusion
. Ethylene oxide sterilization without prior irradiation

Correct Answer & Explanation

. Gamma irradiation in air followed by shelf storage


Explanation

High-dose irradiation cross-links UHMWPE, significantly improving its wear resistance. Subsequent thermal remelting extinguishes residual free radicals trapped in the crystalline regions, preventing long-term in vivo oxidation, although it slightly decreases fatigue strength.

Question 736

Topic: Biomechanics & Biomaterials

A dynamic splint is applied to a patient's elbow to treat a flexion contracture. Over time, the tension in the elastic band of the splint decreases despite the elbow remaining locked in the exact same position. This viscoelastic phenomenon is best described as:

. Creep
. Stress relaxation
. Hysteresis
. Fatigue failure
. Isotropic deformation

Correct Answer & Explanation

. Creep


Explanation

Stress relaxation is the time-dependent decrease in stress (tension) within a viscoelastic material when it is held at a constant strain (deformation). Creep, in contrast, is the progressive deformation of a material over time under a constant load.

Question 737

Topic: Biomechanics & Biomaterials

The combination of a titanium femoral stem and a cobalt-chromium femoral head in total hip arthroplasty relies on a passivation layer to resist corrosion. If this oxide layer is repeatedly breached by micromotion at the head-neck junction, the resulting degradation is most accurately termed:

. Galvanic corrosion
. Crevice corrosion
. Fretting corrosion
. Pitting corrosion
. Intergranular corrosion

Correct Answer & Explanation

. Galvanic corrosion


Explanation

Fretting corrosion occurs when micromotion between two metal surfaces mechanically strips the protective oxide (passivation) layer, leading to accelerated electrochemical degradation. This combined mechanical and chemical process is the primary mechanism of trunnionosis in modular hip arthroplasty.

Question 738

Topic: Biomechanics & Biomaterials

Which of the following orthopedic implant materials possesses a modulus of elasticity most closely resembling that of human cortical bone?

. Cobalt-chromium alloy
. Alumina ceramic
. Titanium alloy (Ti-6Al-4V)
. 316L Stainless steel
. Polymethylmethacrylate (PMMA)

Correct Answer & Explanation

. Cobalt-chromium alloy


Explanation

Titanium alloys have a modulus of elasticity (roughly 110 GPa) that is closer to cortical bone (15-20 GPa) than stainless steel (200 GPa) or cobalt-chromium (220 GPa). This closer biomechanical match helps reduce stress shielding around the implant.

Question 739

Topic: Biomechanics & Biomaterials

Which of the following alterations occurs when a radiopacifier such as barium sulfate or zirconium dioxide is added to polymethylmethacrylate (PMMA) bone cement?

. Decreased fatigue strength
. Increased compressive strength
. Decreased exothermic reaction temperature
. Increased setting time
. Decreased volumetric shrinkage

Correct Answer & Explanation

. Decreased fatigue strength


Explanation

The addition of radiopacifiers creates microscopic inclusions and stress risers within the PMMA mantle. This significantly decreases the material's fatigue strength and tensile strength, though it remains highly resistant to compressive loads.

Question 740

Topic: Biomechanics & Biomaterials

During tensile testing of a human anterior cruciate ligament, the initial non-linear 'toe region' of the load-deformation curve is primarily caused by:

. Microfailure of collagen fibrils
. Plastic deformation of elastin fibers
. Uncrimping of the naturally wavy type I collagen fibers
. Viscoelastic stress relaxation
. Fluid exudation from the proteoglycan matrix

Correct Answer & Explanation

. Microfailure of collagen fibrils


Explanation

The non-linear 'toe region' observed at low strains in the stress-strain curve of a ligament corresponds to the uncrimping and straightening of the naturally wavy collagen fibers. Once straightened, the ligament enters the steeper linear elastic region.