This practice set contains high-yield board review questions covering key concepts in Biomechanics & Biomaterials. Each clinical scenario is designed to test your diagnostic and management skills relevant to this subspecialty.
Question 521
Topic: Biomechanics & Biomaterials
Which of the following combinations of orthopedic metals is most likely to result in significant galvanic corrosion when placed in direct contact within the body?
Correct Answer & Explanation
. 316L stainless steel and titanium alloy
Explanation
Galvanic corrosion occurs when two dissimilar metals are in contact in an electrolytic environment. Mixing 316L stainless steel with titanium or cobalt-chromium leads to severe galvanic corrosion of the less noble metal (stainless steel).
Question 522
Topic: Biomechanics & Biomaterials
Ligaments and tendons exhibit stress relaxation, which is a viscoelastic property defined as:
Correct Answer & Explanation
. Decreasing stress over time under a constant deformation
Explanation
Stress relaxation is the decrease in stress (internal force) required to maintain a tissue at a constant strain (length) over time. In contrast, creep is the increasing deformation (lengthening) over time under a constant load.
Question 523
Topic: Biomechanics & Biomaterials
In orthopedic biomechanics, bending stiffness of a solid circular implant is proportional to the area moment of inertia. If the radius of a solid intramedullary rod is doubled, the bending stiffness increases by a factor of:
Correct Answer & Explanation
. 16
Explanation
The area moment of inertia for a solid cylinder is proportional to the radius to the fourth power (r^4). Therefore, doubling the radius (2^4) increases the bending stiffness 16-fold.
Question 524
Topic: Biomechanics & Biomaterials
To improve the wear characteristics of ultra-high-molecular-weight polyethylene (UHMWPE) for total joint arthroplasty, it undergoes highly cross-linking via irradiation followed by remelting. What is the primary biomechanical tradeoff of the remelting process?
Correct Answer & Explanation
. Decreases crystallinity and reduces yield strength/fatigue resistance
Explanation
Remelting UHMWPE (heating above its melting point) effectively eliminates residual free radicals, significantly reducing long-term oxidation risk. However, it decreases the polymer's crystallinity, which subsequently lowers its yield strength and fatigue resistance.
Question 525
Topic: Biomechanics & Biomaterials
A surgeon plans to revise a fractured femoral plate but decides to mix a titanium locking plate with stainless steel screws. What is the primary concern regarding galvanic corrosion in this construct?
Correct Answer & Explanation
. Stainless steel acts as the anode and preferentially corrodes
Explanation
In a galvanic cell with dissimilar metals, the less noble metal acts as the anode and undergoes corrosion. Titanium is highly noble and acts as the cathode, causing the less noble stainless steel (anode) to preferentially corrode.
Question 526
Topic: Biomechanics & Biomaterials
Viscoelastic materials, such as ligaments and articular cartilage, exhibit time-dependent mechanical properties. Which of the following correctly describes 'stress relaxation'?
Correct Answer & Explanation
. A decrease in stress within a material maintained at a constant deformation over time
Explanation
Stress relaxation refers to the decrease in applied stress required to maintain a given, constant strain (deformation) over time. Creep, in contrast, is the progressive deformation under a constant load.
Question 527
Topic: Biomechanics & Biomaterials
The phenomenon where an orthopaedic biomaterial subjected to a constant load exhibits a progressive increase in strain over time is known as:
Correct Answer & Explanation
. Creep
Explanation
Creep is the time-dependent deformation of a viscoelastic material under a constant load. Stress relaxation, conversely, is the decrease in stress over time under a constant strain.
Question 528
Topic: Biomechanics & Biomaterials
Which of the following factors is the primary driving force for galvanic corrosion when two dissimilar metals are implanted in the body?
Correct Answer & Explanation
. Difference in electrochemical potential
Explanation
Galvanic corrosion occurs when two metals with different electrochemical potentials are placed in contact within a conductive fluid. The less noble metal acts as an anode and corrodes.
Question 529
Topic: Biomechanics & Biomaterials
On a stress-strain curve representing an orthopaedic biomaterial, the area under the elastic region is defined as:
Correct Answer & Explanation
. Resilience
Explanation
Resilience is the capacity of a material to absorb energy in the elastic range, represented by the area under the elastic portion of the stress-strain curve. Toughness is the total area under the entire curve.
Question 530
Topic: Biomechanics & Biomaterials
Which of the following best describes the phenomenon of "creep" in viscoelastic orthopaedic materials?
Correct Answer & Explanation
. Increasing deformation under a constant load over time
Explanation
Creep is a viscoelastic property defined as the progressive deformation of a material when subjected to a constant load over time. Stress relaxation, conversely, is the decrease in internal stress over time when a material is held at a constant deformation.
Question 531
Topic: Biomechanics & Biomaterials
Gamma irradiation of ultra-high molecular weight polyethylene (UHMWPE) in an inert environment (e.g., argon or vacuum) followed by remelting is primarily performed to achieve which of the following?
Correct Answer & Explanation
. Eliminate free radicals and reduce oxidative degradation
Explanation
Gamma irradiation facilitates cross-linking to improve wear resistance but also creates free radicals that can lead to oxidative degradation. Remelting the polyethylene eliminates these free radicals, stabilizing the material against oxidation at the cost of slight decreases in mechanical strength.
Question 532
Topic: Biomechanics & Biomaterials
Which of the following orthopaedic biomaterials has a Young's modulus of elasticity most closely matching that of human cortical bone?
Correct Answer & Explanation
. Titanium alloy
Explanation
Human cortical bone has a Young's modulus of approximately 15-20 GPa. Titanium alloys (~110 GPa) are closer to cortical bone than stainless steel (~200 GPa) or cobalt-chromium alloys (~220 GPa), resulting in less stress shielding.
Question 533
Topic: Biomechanics & Biomaterials
Figure 1 displays a stress-strain curve for a typical metallic orthopaedic implant.
The exact point on the curve where the material transitions from completely recoverable elastic deformation to non-recoverable plastic deformation is defined as the:
Correct Answer & Explanation
. Yield point
Explanation
The yield point characterizes the transition from elastic to plastic behavior on a stress-strain curve. Loading the material beyond this specific point results in permanent (plastic) deformation.
Question 534
Topic: Biomechanics & Biomaterials
Galvanic corrosion in orthopaedic implants occurs under which of the following conditions?
Correct Answer & Explanation
. Contact between two dissimilar metals in an electrolytic fluid environment
Explanation
Galvanic corrosion occurs when two electrochemically dissimilar metals are in direct physical contact within a conductive fluid environment (e.g., body fluids). The less noble metal acts as an anode and undergoes accelerated corrosion.
Question 535
Topic: Biomechanics & Biomaterials
In the context of tendon biomechanics, which of the following best describes the phenomenon of stress relaxation?
Correct Answer & Explanation
. Decrease in stress over time when a material is held at a constant length or deformation
Explanation
Stress relaxation is a viscoelastic property where the stress in a material decreases over time when it is held at a constant deformation. In contrast, creep is the continued deformation over time under a constant load.
Question 536
Topic: Biomechanics & Biomaterials
A surgeon considers using a titanium screw to secure a stainless steel plate. Which of the following best describes the primary biomechanical and biologic concern regarding this construct?
Correct Answer & Explanation
. Galvanic corrosion
Explanation
Mixing different metals, such as stainless steel and titanium, in an aqueous environment like the human body sets up an electrochemical cell. This leads to galvanic corrosion, where the less noble metal undergoes accelerated corrosion.
Question 537
Topic: Biomechanics & Biomaterials
The torsional strength of a solid intramedullary nail is proportional to its radius raised to which power?
Correct Answer & Explanation
. 5
Explanation
The polar moment of inertia for a solid cylinder, which determines torsional stiffness and strength, is proportional to the radius to the fourth power (r^4). Therefore, small increases in nail diameter exponentially increase its torsional strength.
Question 538
Topic: Biomechanics & Biomaterials
Which of the following orthopedic materials has a Young's modulus (modulus of elasticity) that most closely matches that of cortical bone?
Correct Answer & Explanation
. Titanium alloy
Explanation
Titanium alloy has a Young's modulus much closer to that of cortical bone compared to stiffer materials like stainless steel and cobalt-chromium. This closer modulus match helps reduce stress shielding around implants.
Question 539
Topic: Biomechanics & Biomaterials
Galvanic corrosion is most likely to occur in an orthopaedic implant under which of the following conditions?
Correct Answer & Explanation
. Using a titanium screw in a stainless steel plate
Explanation
Galvanic corrosion occurs when two dissimilar metals are placed in electrical contact within a conductive fluid medium, such as body fluid. The less noble metal serves as the anode and undergoes accelerated corrosion.
Question 540
Topic: Biomechanics & Biomaterials
Stress shielding around a femoral stem is most strongly influenced by which of the following material properties of the implant?
Correct Answer & Explanation
. Modulus of elasticity
Explanation
Modulus of elasticity relates to a material's stiffness. Implants with a high modulus of elasticity bear most of the load when placed in bone, causing stress shielding and subsequent resorption of the unloaded surrounding bone.